What is the highest rated pressure CPAP machine?

Post by SW 2.0 » Sun Dec 18, 2016 7:08 pm

I'm in need of the highest I can find for a project I'm working on (not for CPAP), so I need to know what used machine I should look for. Is 25 psi the highest? CPAP machines are the quietest I can find for what I'm trying to do.

Greatly appreciate any insight you can share.

Thanks!

_________________
MachineMask

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: What is the highest rated pressure CPAP machine?

Post by Wulfman... » Sun Dec 18, 2016 7:12 pm

SW 2.0 wrote:I'm in need of the highest I can find for a project I'm working on (not for CPAP), so I need to know what used machine I should look for. Is 25 psi the highest? CPAP machines are the quietest I can find for what I'm trying to do.

Greatly appreciate any insight you can share.

Thanks!
First of all, the pressure isn't measured in PSI. It's measured in CWP (Centimeters of Water Pressure).

CPAP machines only go to 20 cm.
BiLevel machines can go to 25 cm. and some can go to 30 cm.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: What is the highest rated pressure CPAP machine?

Post by JDS74 » Sun Dec 18, 2016 7:35 pm

If you really need a pressure like 25 PSI, then the pressure from a CPAP machine will only provide a really small fraction of that number. A CPAP at 25 Cm H2O provides a pressure of about 161 grams per square inch. That is the equivalent of about 0.35 PSI if my calculations are close.

Putting it another way, 25 PSI = 11,340 grams per square inch or 1,734 cm H2O.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

User avatar
Goofproof
Posts: 16082
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: What is the highest rated pressure CPAP machine?

Post by Goofproof » Sun Dec 18, 2016 8:22 pm

In other words, you need a air compressure, not a fan. Jim