Overseas travel

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
barrygodofsky

Overseas travel

Post by barrygodofsky » Mon Sep 26, 2016 10:38 am

I have a ResMed TAngo C-Series. My question is: Will it automatically convert to 220 AC (like an iPhone for example)?
Or do I need to bring over a 110 to 220 converter in addition to an electrical adapter? P
Thanks

User avatar
Julie
Posts: 20056
Joined: Tue Feb 28, 2006 12:58 pm

Re: Overseas travel

Post by Julie » Mon Sep 26, 2016 11:01 am

Where are you going?

User avatar
Pugsy
Posts: 65119
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Overseas travel

Post by Pugsy » Mon Sep 26, 2016 11:17 am

See product spec page for international voltage

http://www.cpapxchange.com/cpap-machine ... tsheet.pdf

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
Last edited by Pugsy on Mon Sep 26, 2016 7:58 pm, edited 1 time in total.
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Overseas travel

Post by palerider » Mon Sep 26, 2016 11:19 am

barrygodofsky wrote:I have a ResMed TAngo C-Series. My question is: Will it automatically convert to 220 AC (like an iPhone for example)?
Or do I need to bring over a 110 to 220 converter in addition to an electrical adapter? P
Thanks
why ask us, look at the power supply for the thing, or the thing itself, it'll specify what type of power it accepts.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

D.H.
Posts: 3532
Joined: Sun Sep 04, 2016 7:07 pm

Re: Overseas travel

Post by D.H. » Mon Sep 26, 2016 6:42 pm

barrygodofsky wrote:I have a ResMed TAngo C-Series. My question is: Will it automatically convert to 220 AC (like an iPhone for example)?
Or do I need to bring over a 110 to 220 converter in addition to an electrical adapter? P
Thanks
I looked at link provided by "Pugsy." It seems to contain contradictory information on this subject. As far as I could tell, there is no "brick" (transformer) on the power cord for this model. Assuming that's right, the voltage information will be on the bottom or side of the machine. If there is a "brick," you will find it there.

Note that the socket or "power point" configuration will be different than it is in the U.S. and Canada, so you'll need at least an adapter.

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

D.H.
Posts: 3532
Joined: Sun Sep 04, 2016 7:07 pm

"barrygodofsky" please register

Post by D.H. » Mon Sep 26, 2016 6:44 pm

That would be appreciated.

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

Guest

Re: "this is a waste of breath II"

Post by Guest » Mon Sep 26, 2016 6:55 pm

D H wrote:That would be appreciated.
i know this is a waste of breath but people are not likely to see a msg in the subject after the initial post
k i tried

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Overseas travel

Post by palerider » Mon Sep 26, 2016 7:01 pm

D.H. wrote:
barrygodofsky wrote:I have a ResMed TAngo C-Series. My question is: Will it automatically convert to 220 AC (like an iPhone for example)?
Or do I need to bring over a 110 to 220 converter in addition to an electrical adapter? P
Thanks
I looked at link provided by "Pugsy." It seems to contain contradictory information on this subject. .
why don't you leave reading fact sheets to people more competent to read them?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
paprt
Posts: 17
Joined: Fri Sep 23, 2016 5:41 am
Location: USA

Re: Overseas travel

Post by paprt » Mon Sep 26, 2016 7:39 pm

According to Pugsy's very helpful link the pap unit itself has a power supply input range of 110V to 240V AC and therefore should be safe for the 220V you inquired about. The humidifier, however, appears only good for a power supply input of 110V AC.

User avatar
Uncle_Bob
Posts: 2777
Joined: Tue Feb 24, 2009 12:10 pm
Location: Arizona

Re: Overseas travel

Post by Uncle_Bob » Mon Sep 26, 2016 8:40 pm

Check your user guide on your unit. It should tell you what it supports. It may say something like :

AC Input range: 110–120V and 220–240V; 50–60Hz; 60VA

This works in both in the US (110-120V AC) and in most of Europe (220-240V). It supports both 50 and 60Hz frequencies. All that you need to do is to pick the correct plug adapter (not a voltage converter).

Also depending on where you are going surge protector might be of some use.

Enjoy your trip

D.H.
Posts: 3532
Joined: Sun Sep 04, 2016 7:07 pm

Re: Overseas travel

Post by D.H. » Tue Sep 27, 2016 10:47 am

Uncle_Bob wrote:Check your user guide on your unit. It should tell you what it supports. It may say something like :

AC Input range: 110–120V and 220–240V; 50–60Hz; 60VA

This works in both in the US (110-120V AC) and in most of Europe (220-240V). It supports both 50 and 60Hz frequencies. All that you need to do is to pick the correct plug adapter (not a voltage converter).

Also depending on where you are going surge protector might be of some use.

Enjoy your trip
Beware that a standard surge suppressor that you buy in the U.S. or Canada will not wourk in Europe and vice-versa.

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

RobertS975
Posts: 211
Joined: Tue Feb 11, 2014 6:03 pm
Location: MA

Re: Overseas travel

Post by RobertS975 » Tue Sep 27, 2016 2:36 pm

barrygodofsky wrote:I have a ResMed TAngo C-Series. My question is: Will it automatically convert to 220 AC (like an iPhone for example)?
Or do I need to bring over a 110 to 220 converter in addition to an electrical adapter? P
Thanks
The quick answer is yes! Just make sure that you have the correct adaptor that will plug into the wall or end table electrical outlet for whatever country you will be in.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Overseas travel

Post by palerider » Tue Sep 27, 2016 2:51 pm

RobertS975 wrote:
barrygodofsky wrote:I have a ResMed TAngo C-Series. My question is: Will it automatically convert to 220 AC (like an iPhone for example)?
Or do I need to bring over a 110 to 220 converter in addition to an electrical adapter? P
Thanks
The quick answer is yes! Just make sure that you have the correct adaptor that will plug into the wall or end table electrical outlet for whatever country you will be in.
and don't take the humidifier.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

D.H.
Posts: 3532
Joined: Sun Sep 04, 2016 7:07 pm

Re: Overseas travel

Post by D.H. » Tue Sep 27, 2016 3:37 pm

palerider wrote:
RobertS975 wrote:
barrygodofsky wrote:I have a ResMed TAngo C-Series. My question is: Will it automatically convert to 220 AC (like an iPhone for example)?
Or do I need to bring over a 110 to 220 converter in addition to an electrical adapter? P
Thanks
The quick answer is yes! Just make sure that you have the correct adaptor that will plug into the wall or end table electrical outlet for whatever country you will be in.
and don't take the humidifier.
In this case, please try using the machine at home without the humidifier first, and make sure you're comfortable with it.

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Overseas travel

Post by palerider » Tue Sep 27, 2016 4:40 pm

D.H. wrote:
palerider wrote:
RobertS975 wrote:
barrygodofsky wrote:I have a ResMed TAngo C-Series. My question is: Will it automatically convert to 220 AC (like an iPhone for example)?
Or do I need to bring over a 110 to 220 converter in addition to an electrical adapter? P
Thanks
The quick answer is yes! Just make sure that you have the correct adaptor that will plug into the wall or end table electrical outlet for whatever country you will be in.
and don't take the humidifier.
In this case, please try using the machine at home without the humidifier first, and make sure you're comfortable with it.
another unnecessary post.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.