Is my new machine damaged?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
PEF
Posts: 423
Joined: Tue Sep 29, 2015 9:41 pm

Is my new machine damaged?

Post by PEF » Mon Aug 15, 2016 8:24 am

Hi Everyone. I just got back from a 2 week trip to the US. I have a question I hope some of you can answer.

I purchased a new Dreamstation and picked it up in the US while I was there. For traveling, I removed it from the box and carried it in it´s case. Unfortunately, because we had too much luggage to handle, I set it on a 3 foot high table. The case tends to roll easily instead of sitting solid on a surface. It rolled off the table on to a tiled floor, making a fairly loud noise. I opened the case and discovered that it had partially disconnected from it´s humidifier. It appears that none of the plastic on the outside of the machine was cracked or broken. When I plugged it in, it seemed to blow and power up fine. Does anyone know if any of the electronics might be damaged? We are swamped here and will not be able to get it set up for another week or so.

Any guidance would be greatly appreciated.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: I also use the Airfit P10 nasal pillow mask

User avatar
LSAT
Posts: 13397
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Is my new machine damaged?

Post by LSAT » Mon Aug 15, 2016 8:32 am

Without using it there is no way of knowing.....................You can't look at a unit and know if any electronics are damaged.

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Is my new machine damaged?

Post by Julie » Mon Aug 15, 2016 8:34 am

Ridiculous question - how would we know?

Like saying "a car hit me, do you think something's wrong?" without of course exam, xrays, anything.
Last edited by Julie on Mon Aug 15, 2016 9:19 am, edited 1 time in total.

User avatar
Sir NoddinOff
Posts: 4189
Joined: Mon May 14, 2012 5:30 pm
Location: California

Re: Is my new machine damaged?

Post by Sir NoddinOff » Mon Aug 15, 2016 9:18 am

Modern solid state electronics are usually pretty resistant to shocks like that unless the case is broken or something similar. Give it a try. There are a lot of feedback loops in CPAP machines and my experience is that the machines either work properly or they just don't turn on correctly (usually error codes on the screen). You won't be electrocuted so no worry there if that's a concern.

If you're using Sleepyhead or whatever, it would be a good idea to check your night's results to see if everything looks normal.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2.
Last edited by Sir NoddinOff on Mon Aug 15, 2016 1:21 pm, edited 1 time in total.
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.

Lucyhere
Posts: 1949
Joined: Sun Oct 18, 2015 2:45 pm

Re: Is my new machine damaged?

Post by Lucyhere » Mon Aug 15, 2016 11:38 am

Hi PEF... since you powered it up and say it seemed find, chances are it is fine. If it isn't, you'll know pretty quickly and at that point you'll need to have it checked out. I had a similar experience a few years back and my machine has worked fine since then. These machines don't break as easily as people think.
Resmed AirSense 10 Autoset for her w/humid air/heated Humidifier
Bleep/P10

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Is my new machine damaged?

Post by palerider » Mon Aug 15, 2016 4:16 pm

PEF wrote:It rolled off the table on to a tiled floor, making a fairly loud noise. I opened the case and discovered that it had partially disconnected from it´s humidifier. It appears that none of the plastic on the outside of the machine was cracked or broken. When I plugged it in, it seemed to blow and power up fine. Does anyone know if any of the electronics might be damaged?
it's unlikely that the electronics are damaged, but the clips that hold the two halves together may have been. that's why you should always separate the two halves when traveling. the resmed cases have separate compartments for the humidifier and blower, as does the respironics travel case (which costs extra $$).

hopefully, nothing in the mechanism that holds the parts together was broken, and if so, I'm sure some duct tape will hold them together without problem, or clear packing tape if the duct tape is too unsightly.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Pugsy
Posts: 65270
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Is my new machine damaged?

Post by Pugsy » Mon Aug 15, 2016 4:20 pm

You won't know for sure until you actually use it but chances are the electronics are just fine.
These machines are fairly tough and unless they take a really high fall they are pretty durable.
I have knocked mine off the night table with water in the chamber and nothing bad happened. Got lucky with the water part...usually if the electronics get water on them they are toast.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
PEF
Posts: 423
Joined: Tue Sep 29, 2015 9:41 pm

Re: Is my new machine damaged?

Post by PEF » Mon Aug 15, 2016 6:23 pm

Oh thanks so much everyone. Thanks, palerider for the tip about separating the unit from the humidifer when traveling. Fortunately the clip seems to work ok, just goes in a little looser than before.

I figured others on this forum would have had some similar experiences. I am going to set this up tomorrow, so I should know by then.

Someone mentioned Sleepyhead and I have a question about that. I want to also use Sleepyhead for this Dreamstation. How do I load Sleepyhead for 2 different machines on the same computer? Right now, Sleepyhead on my computer recognizes my Resmed S9. Can you set up Sleepyhead for 2 different machines or do I need a separate install, maybe on a different computer?

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: I also use the Airfit P10 nasal pillow mask

User avatar
Pugsy
Posts: 65270
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Is my new machine damaged?

Post by Pugsy » Mon Aug 15, 2016 6:34 pm

PEF wrote:How do I load Sleepyhead for 2 different machines on the same computer?
You don't have to do anything special at all unless you just want to keep the data from the 2 machines separate.
SleepyHead won't care if you are using 2 machines under one profile.
The only thing it can't handle is if you use 2 machines within a 24 hour period...it won't know which result to post.

So you can just use the same user profile if you wish. I think I had at one time 5 different machines under one profile in SleepyHead.
If you want to keep the data totally separate all you have to do is create another user profile.
Just name it something so you can see it is for the special machine. You can have as many user profiles as you want to mess with in SleepyHead. No need for another install or different computer or anything like that. Super easy to handle multiple machines either in one profile or multiple profiles if you wish.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Uncle_Bob
Posts: 2777
Joined: Tue Feb 24, 2009 12:10 pm
Location: Arizona

Re: Is my new machine damaged?

Post by Uncle_Bob » Mon Aug 15, 2016 6:41 pm

Almost all devices, especially medical devices will perform some diagnostics at startup and through up an error code if its having a barney.

As long as your machine starts up OK, its blowing and your results are looking OK then you should be good.
You could also take it back to your sleep lab and they will likely perform a complementary calibration test to make sure its still blowing at the right pressure.

~UB

User avatar
PEF
Posts: 423
Joined: Tue Sep 29, 2015 9:41 pm

Re: Is my new machine damaged?

Post by PEF » Mon Aug 15, 2016 6:45 pm

Oh thanks Pugsy. It will be much easier than I thought to switch over.

I mainly wanted the Dreamstation because it attempts to record RERA´s which my Resmed S9 does not do.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: I also use the Airfit P10 nasal pillow mask

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Is my new machine damaged?

Post by palerider » Mon Aug 15, 2016 7:26 pm

PEF wrote:Oh thanks Pugsy. It will be much easier than I thought to switch over.

I mainly wanted the Dreamstation because it attempts to record RERA´s which my Resmed S9 does not do.
I prefer a single profile, like Pugsy suggests, that keeps all the records in one easy place so you can see night to night, without having to switch to different profiles, programs, etc.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
PEF
Posts: 423
Joined: Tue Sep 29, 2015 9:41 pm

Re: Is my new machine damaged?

Post by PEF » Tue Aug 16, 2016 7:25 am

Yes, I will do the single profile, much simpler!

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: I also use the Airfit P10 nasal pillow mask

User avatar
PEF
Posts: 423
Joined: Tue Sep 29, 2015 9:41 pm

Re: Is my new machine damaged?

Post by PEF » Thu Aug 18, 2016 1:23 pm

I have my new Dreamstation all set up and it is fine. Thanks to everyone for the advice.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: I also use the Airfit P10 nasal pillow mask

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Is my new machine damaged?

Post by chunkyfrog » Thu Aug 18, 2016 1:26 pm

Yay! No sparks or smoke!

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her