Changing masks makes a huge difference

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
andy2016
Posts: 46
Joined: Thu Jun 02, 2016 5:03 pm
Location: Los Angeles, CA

Changing masks makes a huge difference

Post by andy2016 » Fri Jun 24, 2016 9:50 am

So initially I was using the nasal pillow and I just couldn't breath. I gave it a shot for a week and although my AHI went down, I didn't feel refreshed. Yesterday I switched to the WISP nasal mask for the first time and sleep 8hours waking up just once and feeling better and bit more refreshed. My AHI was 3 but its a vast improvement from 67 per hour. My hypopnea index was 3 per hour. How do I decrease that further? Thanks

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack
Humidifier: DreamStation Heated Humidifier
Additional Comments: Started CPAP therapy June 13, 2016.......Sleep study showed AHI of 67.5. I use sleepyhead software. Pressure setting 11 to 14.5.

CPAP_DAD6
Posts: 51
Joined: Fri Jun 03, 2016 2:59 pm

Re: Changing masks makes a huge difference

Post by CPAP_DAD6 » Fri Jun 24, 2016 10:36 am

andy2016 wrote:So initially I was using the nasal pillow and I just couldn't breath. I gave it a shot for a week and although my AHI went down, I didn't feel refreshed. Yesterday I switched to the WISP nasal mask for the first time and sleep 8hours waking up just once and feeling better and bit more refreshed. My AHI was 3 but its a vast improvement from 67 per hour. My hypopnea index was 3 per hour. How do I decrease that further? Thanks



I think your doing a good job. Most ppl could only dream of having that low of numbers. Pun intended.

_________________
MachineMask

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Changing masks makes a huge difference

Post by palerider » Fri Jun 24, 2016 2:06 pm

andy2016 wrote: My hypopnea index was 3 per hour. How do I decrease that further? Thanks
hypopneas normally respond to higher pressure.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

andy2016
Posts: 46
Joined: Thu Jun 02, 2016 5:03 pm
Location: Los Angeles, CA

Re: Changing masks makes a huge difference

Post by andy2016 » Fri Jun 24, 2016 2:18 pm

palerider wrote:
andy2016 wrote: My hypopnea index was 3 per hour. How do I decrease that further? Thanks
hypopneas normally respond to higher pressure.
Yeah, you might be right. As you can see below, set pressures are 8 to 15 max. It is actually 9 to 15 , not sure why it says 8. Anyway, I keep hitting the max and my avg. is closer 12.5. So is it long enough to request an increase or should I give it more time?

Image

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack
Humidifier: DreamStation Heated Humidifier
Additional Comments: Started CPAP therapy June 13, 2016.......Sleep study showed AHI of 67.5. I use sleepyhead software. Pressure setting 11 to 14.5.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Changing masks makes a huge difference

Post by palerider » Fri Jun 24, 2016 2:38 pm

andy2016 wrote: Anyway, I keep hitting the max and my avg. is closer 12.5. So is it long enough to request an increase or should I give it more time?
the statistics page tells me very little.

I like seeing nightly pages with the charts, showing how things progressed through the night. trying to summarize the whole nights activity in one or two numbers leaves too many questions unanswered and offers too many chances for bad guesses.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

andy2016
Posts: 46
Joined: Thu Jun 02, 2016 5:03 pm
Location: Los Angeles, CA

Re: Changing masks makes a huge difference

Post by andy2016 » Fri Jun 24, 2016 5:24 pm

palerider wrote:
andy2016 wrote: Anyway, I keep hitting the max and my avg. is closer 12.5. So is it long enough to request an increase or should I give it more time?
the statistics page tells me very little.

I like seeing nightly pages with the charts, showing how things progressed through the night. trying to summarize the whole nights activity in one or two numbers leaves too many questions unanswered and offers too many chances for bad guesses.
OK. Not sure which graphs you were referring to so I just pasted all of them for last nite. Thanks.

Image

Image

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack
Humidifier: DreamStation Heated Humidifier
Additional Comments: Started CPAP therapy June 13, 2016.......Sleep study showed AHI of 67.5. I use sleepyhead software. Pressure setting 11 to 14.5.

User avatar
LSAT
Posts: 13342
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Changing masks makes a huge difference

Post by LSAT » Fri Jun 24, 2016 6:11 pm

Change your min pressure to 10 and max to 17.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Changing masks makes a huge difference

Post by palerider » Fri Jun 24, 2016 6:47 pm

LSAT wrote:Change your min pressure to 10 and max to 17.
at least, if not one or two more.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Changing masks makes a huge difference

Post by palerider » Fri Jun 24, 2016 6:48 pm

andy2016 wrote:OK. Not sure which graphs you were referring to so I just pasted all of them for last nite. Thanks.
this:https://sleep.tnet.com/resources/sleepyhead/shorganize

though the time at pressure graph would be nice, if the scale hadn't been cut off.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

andy2016
Posts: 46
Joined: Thu Jun 02, 2016 5:03 pm
Location: Los Angeles, CA

Re: Changing masks makes a huge difference

Post by andy2016 » Fri Jun 24, 2016 7:28 pm

palerider wrote:
andy2016 wrote:OK. Not sure which graphs you were referring to so I just pasted all of them for last nite. Thanks.
this:https://sleep.tnet.com/resources/sleepyhead/shorganize

though the time at pressure graph would be nice, if the scale hadn't been cut off.
Think I got it.

Image

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack
Humidifier: DreamStation Heated Humidifier
Additional Comments: Started CPAP therapy June 13, 2016.......Sleep study showed AHI of 67.5. I use sleepyhead software. Pressure setting 11 to 14.5.

andy2016
Posts: 46
Joined: Thu Jun 02, 2016 5:03 pm
Location: Los Angeles, CA

Re: Changing masks makes a huge difference

Post by andy2016 » Fri Jun 24, 2016 7:35 pm

How do I get it from cutting off on the right side?

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack
Humidifier: DreamStation Heated Humidifier
Additional Comments: Started CPAP therapy June 13, 2016.......Sleep study showed AHI of 67.5. I use sleepyhead software. Pressure setting 11 to 14.5.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Changing masks makes a huge difference

Post by palerider » Fri Jun 24, 2016 10:07 pm

andy2016 wrote:How do I get it from cutting off on the right side?
the screenshot you posted looks good. that right panel is useless for figuring out what's going on, same with the calendar, which is why hitting F12 turns them off before it takes the screenshot.

I was about to say "great" about the screenshot, then I saw your comment.

I'd say set min to 11, and max to 17, or more (I usually just leave max at default, 20 because for most people, the pressure won't go much higher than needed... though there are exceptions... however respironics machines are slower to raise pressure anyway, so that's less of a concern with them.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.