OK I'm learning to look at my data and am wondering at how the machine decides to increase pressure. I'm looking at the graphs where I see it increase the pressure and don't see anything that indicates WHY it chooses to increase when it does. What am I missing?
- Wulfman...
- Posts: 6688
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: APAP what causes increase in pressure
The two main things ALL APAPs/Autos use to trigger pressure increases are Flow Limitations and/or Snores.plainhat wrote:OK I'm learning to look at my data and am wondering at how the machine decides to increase pressure. I'm looking at the graphs where I see it increase the pressure and don't see anything that indicates WHY it chooses to increase when it does. What am I missing?
Those two things are programmed into the machines' algorithms with the belief that they should precede Apneas.
Some people have them and some people don't.......preceding apneas. If they don't, then the apneas will occur without a pressure increase to prevent them. Even with FLs and Snores, if the starting pressure is too low and it takes too long for the pressure to increase to get to a point to where it needs to be to prevent them, then the apneas will also occur without the machine being able to reach an appropriate pressure.
Virtually NO APAP/Auto in a range of pressures will attempt to stop an apnea in process. For one thing, it takes at least 10 seconds for the machine to recognize whether the "event" (lack of breath) is actually an apnea. Once that 10 seconds elapses, the "event" will be recorded as an apnea.
Any more questions?
Den
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Re: APAP what causes increase in pressure
As a person’s airway begins to reduce in size a characteristic flattening of the smooth sinusoidal curve of normal breathing appears before clinical signs are apparent. CPAPs developers came up with algorithms that compared the shape of the airflow curve of a current breath with the average of previous breaths. At the commencement of flattening, air pressure is gradually increased to a level at which the shape returns to that normal breathing. The pressure then drops to a lower maintenance level.
_________________
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6 |
- ChicagoGranny
- Posts: 15516
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: APAP what causes increase in pressure
For a more accurate description, read through this - http://www.resmed.com/int/clinicians/co ... nc=dealersplainhat wrote:how the machine decides to increase pressure
- Wulfman...
- Posts: 6688
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: APAP what causes increase in pressure
What you're stating is under the presumption that it can be done or certain characteristics exist.avi123 wrote:As a person’s airway begins to reduce in size a characteristic flattening of the smooth sinusoidal curve of normal breathing appears before clinical signs are apparent. CPAPs developers came up with algorithms that compared the shape of the airflow curve of a current breath with the average of previous breaths. At the commencement of flattening, air pressure is gradually increased to a level at which the shape returns to that normal breathing. The pressure then drops to a lower maintenance level.
Some people just have nasal obstructions or breathing issues like enlarged turbinates or deviated septums that make the "curve" flattened to begin with. For some people, increasing pressures may not only NOT restore the shape of the air curve, but may also make it worse. Consequently, the pressure keeps increasing and the Flow Limitations get worse. And, those breathing anomalies may not precede apneas either.
Den
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
- ButtermilkBuoy
- Posts: 147
- Joined: Mon Nov 19, 2012 9:13 am
Re: APAP what causes increase in pressure
The algorithms recognize persistent patterns of flow limitations (PFLs) and consider them as normal. The algorithms do not raise the pressure in response to PFLs.Wulfman... wrote:Some people just have nasal obstructions or breathing issues like enlarged turbinates or deviated septums that make the "curve" flattened to begin with. For some people, increasing pressures may not only NOT restore the shape of the air curve, but may also make it worse. Consequently, the pressure keeps increasing and the Flow Limitations get worse.
- Wulfman...
- Posts: 6688
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: APAP what causes increase in pressure
I don't know what the criteria is for which you're referring, but I've seen many cases of pressures maxing out on flow limitations. And, the increased pressures didn't reduce the flow limitations......and in many cases got worse.ButtermilkBuoy wrote:The algorithms recognize persistent patterns of flow limitations (PFLs) and consider them as normal. The algorithms do not raise the pressure in response to PFLs.Wulfman... wrote:Some people just have nasal obstructions or breathing issues like enlarged turbinates or deviated septums that make the "curve" flattened to begin with. For some people, increasing pressures may not only NOT restore the shape of the air curve, but may also make it worse. Consequently, the pressure keeps increasing and the Flow Limitations get worse.
In some cases, Flow Limitations can result from a reflex reaction from the increasing pressures.
Den
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Re: APAP what causes increase in pressure
you really should read the link that Granny posted and update what you say to be more accurate.Wulfman... wrote:The two main things ALL APAPs/Autos use to trigger pressure increases are Flow Limitations and/or Snores.plainhat wrote:OK I'm learning to look at my data and am wondering at how the machine decides to increase pressure. I'm looking at the graphs where I see it increase the pressure and don't see anything that indicates WHY it chooses to increase when it does. What am I missing?
Those two things are programmed into the machines' algorithms with the belief that they should precede Apneas.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: APAP what causes increase in pressure
here are links gathered by Robysue that will more than answer your questions:plainhat wrote:OK I'm learning to look at my data and am wondering at how the machine decides to increase pressure. I'm looking at the graphs where I see it increase the pressure and don't see anything that indicates WHY it chooses to increase when it does. What am I missing?
viewtopic.php?f=1&t=111653&p=1072250
Last edited by palerider on Thu May 19, 2016 12:12 am, edited 1 time in total.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- Wulfman...
- Posts: 6688
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: APAP what causes increase in pressure
I did and I have and I don't see anything that particularly contradicts anything I've said.palerider wrote:you really should read the link that Granny posted and update what you say to be more accurate.Wulfman... wrote:The two main things ALL APAPs/Autos use to trigger pressure increases are Flow Limitations and/or Snores.plainhat wrote:OK I'm learning to look at my data and am wondering at how the machine decides to increase pressure. I'm looking at the graphs where I see it increase the pressure and don't see anything that indicates WHY it chooses to increase when it does. What am I missing?
Those two things are programmed into the machines' algorithms with the belief that they should precede Apneas.
I'm even copying and highlighting the information from that website link to emphasize the points. And, I wasn't just talking about ResMed Autos because many users may not have ResMed APAPs or their more recent machines with newer algorithms.
Den
.
Lower average treatment pressures
The clinically validated1 AutoSet™ algorithm allows the delivery of lower average pressures1, without compromising efficacy
The pre-emptive response focuses on keeping the upper airway open
This can mean lower treatment pressures, increased comfort for your patients and fewer complaints of pressure-related side effects
AutoSet device's Three Lines of Defence
Click on the question to expand and reveal more information...
1. Flow Limitation
What is flow limitation?
A physiological change in the status of the upper airway
Causes a spectrum of closure, from subtle narrowing to partial collapse
What is the significance of flow limitation?
It usually precedes snoring and apnea. It is therefore, usually, the earliest sign of impending airway collapse3
Airway changes limit flow despite respiratory effort and in fact, cause increased respiratory effort. This may lead to arousal
How do I observe flow limitation?
The inspiratory flow time curve represents inspiratory flow limitation
The shape of the curve indicates the status of the upper airway. A normal upper airway is observed as a bell-shaped curve. The curve flattens with increasing flow limitation and airway narrowing/ closure
How does a ResMed AutoSet device measure flow limitation?
It calculates flow limitation on a breath-by-breath basis, detecting and responding to subtle changes
How does a ResMed AutoSet device respond to flow limitation?
If flow limitation is detected, AutoSet Spirit responds by gradually increasing pressure to bring the airway back to normal. Typically, this helps prevent snoring and apneas4
If no further events occur, AutoSet Spirit gently decreases the pressure, towards the minimum set pressure
2. Snore
What is snore?
Snoring occurs when the walls of the upper airway narrow and vibrate
What is the significance of snore?
It may precede obstruction of the airway and apneic events
How do I observe snore?
During snore, the inspiratory flow time curve flattens and the snore oscillation is super-imposed
How does a ResMed AutoSet device measure snore?
It has a pressure transducer that measures vibrations of the airflow
How does a ResMed AutoSet device respond to snore?
Increases pressure based upon the severity of the snore
If no further events occur, AutoSet Spirit gently decreases the pressure, towards the minimum set pressure
3. Apnea
What is apnea?
A condition where the upper airway becomes completely obstructed or severely limited
What is the significance of apnea?
Apnea involves sleep fragmentation, oxygen desaturation, and arousals during the night as well as several daytime symptoms5
Sleep Disordered Breathing (SDB) has been shown to be a significant contributing factor for hypertension,6,7,8 one of the leading risk factors for stroke and Congestive Heart Failure (CHF)
How do I observe apnea?
During apnea, the inspiratory flow time curve disappears, leaving a straight time line, since there is no flow
How does a ResMed AutoSet device measure apnea?
As the apneic cessation of airflow or breathing for more than 10 seconds
How does a ResMed AutoSet device respond to apnea?
Typically prevents apneas, by treating flow limitation and snore. These usually act as early signals for more severe apneic events
In the event of a sudden apnea that occurs without any preceding flow limitation or snore, AutoSet Spirit increases the pressure after the apnea has ceased to prevent further apneic events
Pressure response is based on the duration of the apnea
If no further events occur, AutoSet Spirit gently decreases the pressure, towards the minimum set pressure
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Re: APAP what causes increase in pressure
it's not so much contradictions of what you said, more the things you left out.Wulfman... wrote:I did and I have and I don't see anything that particularly contradicts anything I've said..palerider wrote:you really should read the link that Granny posted and update what you say to be more accurate.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- Wulfman...
- Posts: 6688
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: APAP what causes increase in pressure
< Yaaaaaawwwwwwnnnnnnnnn >palerider wrote:it's not so much contradictions of what you said, more the things you left out.Wulfman... wrote:I did and I have and I don't see anything that particularly contradicts anything I've said..palerider wrote:you really should read the link that Granny posted and update what you say to be more accurate.
Den
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Re: APAP what causes increase in pressure
yeah, I know, giving people good and accurate information is boring... give the half of the story that interests you, that's good enough, eh?Wulfman... wrote:< Yaaaaaawwwwwwnnnnnnnnn >palerider wrote:it's not so much contradictions of what you said, more the things you left out.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- Wulfman...
- Posts: 6688
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: APAP what causes increase in pressure
Well, there was almost an hour from the time that the OP originally posted........and YOU didn't post any information, so I took a stab at it. I also hate to post loooooong posts that put people to sleep.......so I try to keep my responses somewhat brief and to the point. And, there already is an abundance of information on this forum and the manufacturers sites.palerider wrote:yeah, I know, giving people good and accurate information is boring... give the half of the story that interests you, that's good enough, eh?Wulfman... wrote:< Yaaaaaawwwwwwnnnnnnnnn >palerider wrote:it's not so much contradictions of what you said, more the things you left out.
From now on, I'll just wait for you to post first. And then critique your posts.
Den
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Re: APAP what causes increase in pressure
well, if you hadn't beaten the "APAPs raise pressure because of snores and flow limitations" incomplete bit into the dirt already, I might buy that...Wulfman... wrote:Well, there was almost an hour from the time that the OP originally posted........and YOU didn't post any information, so I took a stab at it. I also hate to post loooooong posts that put people to sleep.......so I try to keep my responses somewhat brief and to the point. And, there already is an abundance of information on this forum and the manufacturers sites.