How Many Centrals Are Too "Many"

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
MDee
Posts: 33
Joined: Wed Jan 13, 2016 12:21 pm
Location: USA

How Many Centrals Are Too "Many"

Post by MDee » Thu Apr 28, 2016 9:08 am

I am still in my first 30 days of biPap. MY AHI is around 2-3 each night. Right now my OA/Hypo combined is around 1. The rest is Central. I figure I need at least one month (or 2) to get the settings right and my body adjusts. But when I have adjusted and if I am still not feeling rested, and if the Centrals still make up more of my AHI than OA/Hypo, do I revisit? During diagnostic testing my CA AHI was at or near 5, so an average of 2 is still much better than without any biPap. And the machine has done wonders in lowering my combined OA/Hypo down from 50 AHI to 1.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Swift™ FX Nano For Her Nasal CPAP Mask with Headgear
Additional Comments: Ipap: 10, Epap: 4, PS: 4. Set to Vauto mode. Sleepyhead Software
Machine: ResMed AirCurve 10 Vauto
Mask: F&P Eson Nasal Mask
IPAP: 15, EPAP: 11 PS: 4. Set to Vauto mode
Software: Sleepyhead

User avatar
LSAT
Posts: 13395
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: How Many Centrals Are Too "Many"

Post by LSAT » Thu Apr 28, 2016 9:18 am

I would guess that if you have more than 20-25 a night...and a high AHI...it's too many for the average user. ...
Last edited by LSAT on Thu Apr 28, 2016 9:44 am, edited 2 times in total.

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: How Many Centrals Are Too "Many"

Post by Jay Aitchsee » Thu Apr 28, 2016 9:28 am

Don't worry about a small number of centrals. Most centrals reported by machine probably wouldn't be scored under PSG conditions. "Centrals" (holding your breath) often accompany brief arousals due to disturbance, turning over in bed, waking, etc. Since the machine doesn't have any way to tell if your were awake, it scores them as "centrals". A CAI under 5 would not normally be of concern.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video

User avatar
tmoody
Posts: 30
Joined: Fri Mar 18, 2016 10:27 am
Location: Philadelphia

Re: How Many Centrals Are Too "Many"

Post by tmoody » Thu Apr 28, 2016 10:27 am

Take a look at http://www.carolinasleepsociety.org/doc ... of_asv.pdf and go to page 34, where it gives a definition of central sleep apnea:

AHI>5
Centrals >50% of apneas
Centrals or hypos >5/hour
Symptoms of disrupted sleep

Those conditions are conjunctive; they all have to be satisfied. It gives one perspective on how many centrals are too many, which is the point at which central sleep apnea is defined to be present. Below that threshold, I guess they're just considered stray centrals.

User avatar
jnk...
Posts: 2988
Joined: Fri Sep 19, 2014 12:36 pm
Location: New York State

Re: How Many Centrals Are Too "Many"

Post by jnk... » Thu Apr 28, 2016 12:03 pm

A stray one here and there ain't no biggie. But it isn't always a matter of number. It can be telling when they occur in clusters, one right after another, in a breathing pattern resembling waveform overshoot and undershoot, in close succession. (Wait, was that repetitively repetitious and redundant?) That is the sort of thing that can sometimes benefit from specialized forms of PAP to get around a sort of side-effect of PAP that some people have, if it doesn't resolve itself over time on straight CPAP.
-Jeff (AS10/P30i)

Accounts to put on the foe list: Me. I often post misleading, timewasting stuff.

SewTired
Posts: 1736
Joined: Thu Apr 16, 2015 8:33 am
Location: Minneapolis area

Re: How Many Centrals Are Too "Many"

Post by SewTired » Thu Apr 28, 2016 3:31 pm

MDee wrote:I am still in my first 30 days of biPap. MY AHI is around 2-3 each night. Right now my OA/Hypo combined is around 1. The rest is Central. I figure I need at least one month (or 2) to get the settings right and my body adjusts. But when I have adjusted and if I am still not feeling rested, and if the Centrals still make up more of my AHI than OA/Hypo, do I revisit? During diagnostic testing my CA AHI was at or near 5, so an average of 2 is still much better than without any biPap. And the machine has done wonders in lowering my combined OA/Hypo down from 50 AHI to 1.
If AHI is low and you are getting fewer than 15 centrals a night, shouldn't be a problem. Keep in mind that turning over can result in a 'central' flag. I found that the centrals decreased a lot after a month or two. If they keep increasing, then that's a flag to talk to your doc.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleepyhead
Diabetes 2, RLS & bradycardia
Airsense For Her; Settings: range 8-12, Airfit P10 (M)

MDee
Posts: 33
Joined: Wed Jan 13, 2016 12:21 pm
Location: USA

Re: How Many Centrals Are Too "Many"

Post by MDee » Fri Apr 29, 2016 8:49 am

This is all very helpful. I am getting between 15-20 per night (so far).I don't roll over in my sleep (bad shoulders). They tend to cluster. I'll keep monitoring them. Right now I am trying to recover from the "you need a wide mask" expedition. So much air leaked into my eyes a few nights ago, I am still having to add artificial tears throughout the day.

2jnk: to see the breathing pattern, where do I look on the Sleepyhead graph? Insp/Ex Time?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Swift™ FX Nano For Her Nasal CPAP Mask with Headgear
Additional Comments: Ipap: 10, Epap: 4, PS: 4. Set to Vauto mode. Sleepyhead Software
Machine: ResMed AirCurve 10 Vauto
Mask: F&P Eson Nasal Mask
IPAP: 15, EPAP: 11 PS: 4. Set to Vauto mode
Software: Sleepyhead

User avatar
jnk...
Posts: 2988
Joined: Fri Sep 19, 2014 12:36 pm
Location: New York State

Re: How Many Centrals Are Too "Many"

Post by jnk... » Fri Apr 29, 2016 9:02 am

I am not a SleepyHead user, but my understanding is that graphs can be zoomed, as shown in this thread:

viewtopic.php?f=1&t=108899&p=1038813#p1038628
-Jeff (AS10/P30i)

Accounts to put on the foe list: Me. I often post misleading, timewasting stuff.

lilly747
Posts: 450
Joined: Thu Sep 27, 2012 9:39 am
Location: Northern Calif, USA

Re: How Many Centrals Are Too "Many"

Post by lilly747 » Fri Apr 29, 2016 10:25 am

MDee wrote: Right now my OA/Hypo combined is around 1. The rest is Central. I figure I need at least one month (or 2) to get the settings right and my body adjusts. But when I have adjusted and if I am still not feeling rested, and if the Centrals still make up more of my AHI than OA/Hypo, do I revisit? During diagnostic testing my CA AHI was at or near 5, so an average of 2 is still much better than without any biPap.


Hi DMee,

I had the same experience. But, after 2 or 3 months I did go from 17.0 CAs per Hour to about only 0.4 CAs per Hour.

THEN, a month ago I decided I needed more PS just because it felt good...big mistake. I ONLY raised my PS 0.4cm...my CAs went from 3 a night to 75 one night and avg of 25. I put my PS back 2 weeks ago but my CAs have not come back down yet.

This morning I had an AHI of 3, ALL CAs. All I can suggest is watch yours over time....I sure am watching mine right now....good luck...

MDee
Posts: 33
Joined: Wed Jan 13, 2016 12:21 pm
Location: USA

Re: How Many Centrals Are Too "Many"

Post by MDee » Fri Apr 29, 2016 10:28 am

@jnk, that did the trick (also, the post you linked to was interesting as I too have a connective tissue disorder that can (rarely) cause Central Apnea).

Anyhow, here is what a sample cluster of CAs look like. No idea what I am seeing, but the average duration is ...15 seconds? I think that is what the number inside the parens is (15), (17), (24) etc.

Image

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Swift™ FX Nano For Her Nasal CPAP Mask with Headgear
Additional Comments: Ipap: 10, Epap: 4, PS: 4. Set to Vauto mode. Sleepyhead Software
Machine: ResMed AirCurve 10 Vauto
Mask: F&P Eson Nasal Mask
IPAP: 15, EPAP: 11 PS: 4. Set to Vauto mode
Software: Sleepyhead

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: How Many Centrals Are Too "Many"

Post by robysue » Fri Apr 29, 2016 10:48 am

MDee wrote: They tend to cluster.
Do they tend to cluster during periods when you know you are likely awake or semi-awake? If so, they may be nothing more than normal sleep transition centrals that would not be scored on an in-lab PSG.
Right now I am trying to recover from the "you need a wide mask" expedition. So much air leaked into my eyes a few nights ago, I am still having to add artificial tears throughout the day.
Is the eye misery making you more restless at night?
MDee wrote:@jnk, that did the trick (also, the post you linked to was interesting as I too have a connective tissue disorder that can (rarely) cause Central Apnea).
Is the sleep doc aware of the connective tissue disorder? Is the doc who treats you for the connective tissue disorder aware that you are now on PAP therapy for OSA? If not, you do need to make sure both docs know what's going on with both conditions.
Anyhow, here is what a sample cluster of CAs look like. No idea what I am seeing, but the average duration is ...15 seconds? I think that is what the number inside the parens is (15), (17), (24) etc.
Yep, the number in the parenthesis is the estimated time of the duration of the event. It comes from the machine's data.

For what it's worth, those centrals do not look like the beginning of an overshoot and undershoot cycle. They may be sleep transition central associated with an arousal of some sort. Or not.

But overall, the number of them per night is not (yet) large enough to be a serious concern.

Finally, as lilly747 observed, some people are very sensitive to the PS number. PS = IPAP - EPAP, and your PS = 4 is not particularly large, but if you can exhale comfortably with your current settings, you might try reducing PS = 3.5 or PS = 3 and see if there are fewer CAs scored.

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

MDee
Posts: 33
Joined: Wed Jan 13, 2016 12:21 pm
Location: USA

Re: How Many Centrals Are Too "Many"

Post by MDee » Fri Apr 29, 2016 11:26 am

robysue wrote:
MDee wrote: They tend to cluster.
Do they tend to cluster during periods when you know you are likely awake or semi-awake? If so, they may be nothing more than normal sleep transition centrals that would not be scored on an in-lab PSG.
I tend to discount the ones at the start and at the end, focusing on the ones when I know I am asleep. So for the example I used, it brings it down from 21 to 15.
MDee wrote: Right now I am trying to recover from the "you need a wide mask" expedition. So much air leaked into my eyes a few nights ago, I am still having to add artificial tears throughout the day.
robysue wrote:Is the eye misery making you more restless at night?


I excluded info from the night with the wide mask.
MDee wrote:@jnk, that did the trick (also, the post you linked to was interesting as I too have a connective tissue disorder that can (rarely) cause Central Apnea).
robysue wrote:Is the sleep doc aware of the connective tissue disorder? Is the doc who treats you for the connective tissue disorder aware that you are now on PAP therapy for OSA? If not, you do need to make sure both docs know what's going on with both conditions.
Most doctors do not understand the genetic condition and tend to discount it or ignore it. It is an uphill battle to get them to take it into consideration. There are no docs who treat the connective tissue disorder so you have to work with existing medical professionals who maybe heard about it once in passing in medical school. My primary MD is aware but has no experience in handling it or sleep apnea (she at least listens). The sleep doctor is aware, but tends to ignore it.
MDee wrote: Anyhow, here is what a sample cluster of CAs look like. No idea what I am seeing, but the average duration is ...15 seconds? I think that is what the number inside the parens is (15), (17), (24) etc.
robysue wrote:Yep, the number in the parenthesis is the estimated time of the duration of the event. It comes from the machine's data.

For what it's worth, those centrals do not look like the beginning of an overshoot and undershoot cycle. They may be sleep transition central associated with an arousal of some sort. Or not.

But overall, the number of them per night is not (yet) large enough to be a serious concern.

Finally, as lilly747 observed, some people are very sensitive to the PS number. PS = IPAP - EPAP, and your PS = 4 is not particularly large, but if you can exhale comfortably with your current settings, you might try reducing PS = 3.5 or PS = 3 and see if there are fewer CAs scored.
This is good to know and I will keep working on getting more comfortable with the biPap. I had planned to increase the IPA slowly to deal with aerophagia first and then tackle this next. The mask fitting testing took up most of last week.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Swift™ FX Nano For Her Nasal CPAP Mask with Headgear
Additional Comments: Ipap: 10, Epap: 4, PS: 4. Set to Vauto mode. Sleepyhead Software
Machine: ResMed AirCurve 10 Vauto
Mask: F&P Eson Nasal Mask
IPAP: 15, EPAP: 11 PS: 4. Set to Vauto mode
Software: Sleepyhead

User avatar
jnk...
Posts: 2988
Joined: Fri Sep 19, 2014 12:36 pm
Location: New York State

Re: How Many Centrals Are Too "Many"

Post by jnk... » Fri Apr 29, 2016 12:08 pm

@ Robysue: Thanks.
-Jeff (AS10/P30i)

Accounts to put on the foe list: Me. I often post misleading, timewasting stuff.

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: How Many Centrals Are Too "Many"

Post by Goofproof » Fri Apr 29, 2016 12:45 pm

One that lasts over 6 minutes, it only takes one of those. Time spent in Apnea is seldom mention but very important. Sleep Apnea generally kills you slowly, time spent in Apnea tills me how fast it's killing me. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire