Urgent mask prob, help needed! Please, can you help?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
fremlin89@Gmail.com

Urgent mask prob, help needed! Please, can you help?

Post by fremlin89@Gmail.com » Fri Apr 22, 2016 5:32 am

Hi, can anyone help me urgently?

I have an F&P Simplus f/f mask. I have just discovered that there is a tear on the flimsy nose-bridge part of the cushion. It's pretty bad and i don't know if it will hold on the night. My biggest problem is its 9:30 on Friday night of a holiday weekend (Monday is the holiday) and EVERYWHERE is closed until Tuesday.

Does anyone know if there's a way i can patch up the tear until Tuesday when i can get a new cushion? Or if there's anyone in hills district of Sydney that knows of somewhere open tomorrow that i can get a new cushion?

Thank you, thank you, thank you!

Diamondminek
Posts: 100
Joined: Mon Sep 14, 2015 11:30 am

Re: Urgent mask prob, help needed! Please, can you help?

Post by Diamondminek » Fri Apr 22, 2016 6:57 am

I don't know if that's something eBay or amazon can help with?
I think I read about someone else having this problem, but the mask didn't leak and was still use able. Although I imagine patching it up with some tape might be worth a shot?

User avatar
richart
Posts: 40
Joined: Thu Nov 10, 2011 6:43 pm

Re: Urgent mask prob, help needed! Please, can you help?

Post by richart » Fri Apr 22, 2016 8:02 pm

I think I would be tempted to try a small piece of duct tape...

yaconsult
Posts: 1099
Joined: Fri Nov 14, 2014 10:43 pm
Location: "Silicon Valley", CA

Re: Urgent mask prob, help needed! Please, can you help?

Post by yaconsult » Fri Apr 22, 2016 10:20 pm

Things like this are why most of the experienced cpapers keep at LEAST one spare mask on hand at all times. Things can fail at the worst possible time, in the middle of the night, on a holiday weekend. That's why many of us have a backup machine, spare hoses, spare masks, battery backup, etc. I can not sleep at ALL without cpap and I know it.

It seems that your only option is what you can find in the house now or at the hardware store tomorrow. Something like duct tape or medical tape, like 3M Micropore might work for a temporary repair. Some packaging tape might work as well. I hope you are able to jury rig something to get you through the night. Some stores, like walmart, might be open late if you need it. Good luck!

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: January 2015 Sleep Study Results: Apnea/Hypopnea Index (AHI): 80.2, Sleepyhead

User avatar
Sir NoddinOff
Posts: 4189
Joined: Mon May 14, 2012 5:30 pm
Location: California

Re: Urgent mask prob, help needed! Please, can you help?

Post by Sir NoddinOff » Sat Apr 23, 2016 4:18 pm

I found that it's difficult to get duct tape or whatever to stick to smooth silicone ... so what I'd do FIRST is try do rough-up the area around the tear with some sandpaper, a nail file, anything gritty etc. You could gently rub the area on concrete or tile grout if it comes down to that Hopefully you can find an answer at the hardware store or in a drug store household's aisle. I don't know if you've tried to patch a bicycle tube but 'roughing up' the patch area is basic to that process, not that I'm suggesting a bike tube patch .

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2.
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.

SleepyToo2
Posts: 1005
Joined: Sun Sep 11, 2011 7:55 am
Location: North of Philadelphia, PA

Re: Urgent mask prob, help needed! Please, can you help?

Post by SleepyToo2 » Sat Apr 23, 2016 6:26 pm

Sir NoddinOff wrote:I found that it's difficult to get duct tape or whatever to stick to smooth silicone ... so what I'd do FIRST is try do rough-up the area around the tear with some sandpaper, a nail file, anything gritty etc. You could gently rub the area on concrete or tile grout if it comes down to that Hopefully you can find an answer at the hardware store or in a drug store household's aisle. I don't know if you've tried to patch a bicycle tube but 'roughing up' the patch area is basic to that process, not that I'm suggesting a bike tube patch .
But usually a bike tire repair kit will contain something for roughing up the tire so the patch will stick. That might help with the mask fix.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Use SleepyHead software.
Not a medical professional - just a patient who has done a lot of reading

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Urgent mask prob, help needed! Please, can you help?

Post by Wulfman... » Sat Apr 23, 2016 7:40 pm

fremlin89@Gmail.com wrote:Hi, can anyone help me urgently?

I have an F&P Simplus f/f mask. I have just discovered that there is a tear on the flimsy nose-bridge part of the cushion. It's pretty bad and i don't know if it will hold on the night. My biggest problem is its 9:30 on Friday night of a holiday weekend (Monday is the holiday) and EVERYWHERE is closed until Tuesday.

Does anyone know if there's a way i can patch up the tear until Tuesday when i can get a new cushion? Or if there's anyone in hills district of Sydney that knows of somewhere open tomorrow that i can get a new cushion?

Thank you, thank you, thank you!
Sydney........WHERE?

There are 2 places named Sydney in America.

There are 2 places named Sydney in Canada.

There is one place named Sydney in South Africa.

There is one place named Sydney in Vanuatu.

There is one place named Sydney in Australia.

http://au.geotargit.com/called.php?qcity=Sydney

As has been said, this is why we have backup and spare masks (and everything else).


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

amenite
Posts: 509
Joined: Sat Jan 31, 2015 7:02 pm

Re: Urgent mask prob, help needed! Please, can you help?

Post by amenite » Sun Apr 24, 2016 12:07 am

Wulfman... wrote:
fremlin89@Gmail.com wrote:Hi, can anyone help me urgently?

I have an F&P Simplus f/f mask. I have just discovered that there is a tear on the flimsy nose-bridge part of the cushion. It's pretty bad and i don't know if it will hold on the night. My biggest problem is its 9:30 on Friday night of a holiday weekend (Monday is the holiday) and EVERYWHERE is closed until Tuesday.

Does anyone know if there's a way i can patch up the tear until Tuesday when i can get a new cushion? Or if there's anyone in hills district of Sydney that knows of somewhere open tomorrow that i can get a new cushion?

Thank you, thank you, thank you!
Sydney........WHERE?

There are 2 places named Sydney in America.

There are 2 places named Sydney in Canada.

There is one place named Sydney in South Africa.

There is one place named Sydney in Vanuatu.

There is one place named Sydney in Australia.

http://au.geotargit.com/called.php?qcity=Sydney

As has been said, this is why we have backup and spare masks (and everything else).


Den

.
Looks like there's only one "hills district in Sydney"
https://www.google.com/search?q=hills+d ... 8&oe=utf-8

User avatar
Sir NoddinOff
Posts: 4189
Joined: Mon May 14, 2012 5:30 pm
Location: California

Re: Urgent mask prob, help needed! Please, can you help?

Post by Sir NoddinOff » Sun Apr 24, 2016 9:39 am

I was thinking 'hills district' near Sidney too, but was too lazy to go searching ... good job, mystery solved. Now what holiday would that be

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2.
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.

amenite
Posts: 509
Joined: Sat Jan 31, 2015 7:02 pm

Re: Urgent mask prob, help needed! Please, can you help?

Post by amenite » Sun Apr 24, 2016 9:53 am

For the lazy amongst us ... my inquisitive nature could not be stymied this day ... but you'll need to click at least once on your own:

https://en.wikipedia.org/wiki/Public_ho ... _Australia

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Urgent mask prob, help needed! Please, can you help?

Post by Wulfman... » Sun Apr 24, 2016 2:47 pm

Sir NoddinOff wrote:I was thinking 'hills district' near Sidney too, but was too lazy to go searching ... good job, mystery solved. Now what holiday would that be
But, there are also towns named/spelled Sidney........like Sidney, Nebraska.

http://www.getzipcode.us/en/city-sidney.php


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

amenite
Posts: 509
Joined: Sat Jan 31, 2015 7:02 pm

Re: Urgent mask prob, help needed! Please, can you help?

Post by amenite » Sun Apr 24, 2016 5:47 pm

Wulfman... wrote:
Sir NoddinOff wrote:I was thinking 'hills district' near Sidney too, but was too lazy to go searching ... good job, mystery solved. Now what holiday would that be
But, there are also towns named/spelled Sidney........like Sidney, Nebraska.

http://www.getzipcode.us/en/city-sidney.php
Google seems to have solved this problem too. Note the first result. I think we've pretty well established the location regardless of misspellings or other ambiguity. Still wondering if the duct tape did (or didn't) work out.

http://lmgtfy.com/?q=%27hills+district%27+near+Sidney

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Urgent mask prob, help needed! Please, can you help?

Post by chunkyfrog » Sun Apr 24, 2016 6:31 pm

Sidney, Nebraska is on the edge of Wildcat Hills.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her