New numbers advice

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Miscanthus
Posts: 28
Joined: Fri Apr 01, 2016 2:22 pm

New numbers advice

Post by Miscanthus » Wed Apr 20, 2016 3:54 pm

I still can't get my AHI down. My doctor started me on CPAP at pressure 15 and AHI didn't improve. With advice from the board I switched my machine to auto and set my pressures at 15/17 and have used this for several nights now. I'm confused and don't know if I really should be using CPAP at a higher pressure or the machine on autoset. I do have centrals as well as obstructive apneas. Will someone look at my numbers again and help me understand what they mean and if I am on the right track or just chasing cats? Again, Thanks.
Roy

Image

User avatar
LSAT
Posts: 13397
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: New numbers advice

Post by LSAT » Wed Apr 20, 2016 4:29 pm

In looking at your charts it appears to me that you may need a bipap machine rather than a cpap. Your pressure seems to be maxing out at the 17...I would try raising the max to 20 and see if your AHI drops. The Centrals are insignificant in your case. You need help with leaks also.

lilly747
Posts: 450
Joined: Thu Sep 27, 2012 9:39 am
Location: Northern Calif, USA

Re: New numbers advice

Post by lilly747 » Wed Apr 20, 2016 6:40 pm

Miscanthus wrote:I still can't get my AHI down. My doctor started me on CPAP at pressure 15 and AHI didn't improve.
Down from what? Was 15 your Sleep Study AHI? Did you have a Sleep Study and was a pressure found that worked. I guess I am trying to ask, where did the 15 come from?

Also what does your FL and "snore" graph look like? How long are your events? Do you have any idea why your AHI is high with a pressure of 15? Medications or flopping head for example?

Sorry for the 1001 questions......

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: New numbers advice

Post by Wulfman... » Wed Apr 20, 2016 7:06 pm

Miscanthus wrote:I still can't get my AHI down. My doctor started me on CPAP at pressure 15 and AHI didn't improve. With advice from the board I switched my machine to auto and set my pressures at 15/17 and have used this for several nights now. I'm confused and don't know if I really should be using CPAP at a higher pressure or the machine on autoset. I do have centrals as well as obstructive apneas. Will someone look at my numbers again and help me understand what they mean and if I am on the right track or just chasing cats? Again, Thanks.
Roy
Go back and look at the report you posted for the 15th. Looking at that, your best pressure was 14 cm. Very little going on at that pressure (other than some leaking). I would suggest trying a straight 14 for a few nights and see how things look.
In other words, study your reports.......note where the pressure is "stable" where there are few to no events and maybe try those settings. It wasn't until your pressure shot up in that report of the 15th that the hypopneas got more evident. That may have been caused by "positional" sleeping, too.........as is the possibility that some of these in this report you posted.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

Miscanthus
Posts: 28
Joined: Fri Apr 01, 2016 2:22 pm

Re: New numbers advice

Post by Miscanthus » Wed Apr 20, 2016 7:13 pm

Image

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: New numbers advice

Post by Wulfman... » Wed Apr 20, 2016 7:25 pm

Yeah. Your Flow Limitations are driving your pressure increases and also coincide with many of your Hypopneas, which may indicate "positional" events (limiting your air flow).


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

Mudrock63
Posts: 327
Joined: Thu Oct 22, 2015 10:04 am

Re: New numbers advice

Post by Mudrock63 » Wed Apr 20, 2016 7:36 pm

I was stuck in a range with middling AHI's with an APAP setting of min 14.5 max 20. I didn't get consistently below 2.0 until I tightened the range to min 18.0 max 20.0. Of course everyone is different. I am just chiming in to say you still have room to work with.

_________________
Mask
Additional Comments: 560P

lilly747
Posts: 450
Joined: Thu Sep 27, 2012 9:39 am
Location: Northern Calif, USA

Re: New numbers advice

Post by lilly747 » Wed Apr 20, 2016 7:43 pm

I may be all "WET" here but............

You have times when your FL are all but nil and then times they are crazy high...
If your head is flopping no amount of pressure will work...

As an experiment, I would get a soft cervical collar and try it for ONE night......if it's your head flopping....this will tell you...

You have nothing to lose but your problem.....A collar is harmless...

Miscanthus
Posts: 28
Joined: Fri Apr 01, 2016 2:22 pm

Re: New numbers advice

Post by Miscanthus » Wed Apr 20, 2016 7:53 pm

Thanks guys for all your help!