How is the SH bar graph for Total Time in Apnea calculated?
Last night I recorded 0 OAs, 2 CAs of 32 seconds duration, and 8 hypopneas of 155 seconds duration. My Total Time in Apnea was recorded as 00:03:07. I assume that is 3 minutes, seven seconds of what my Dream Station calculated as apnea.

I would assume the calculation is the total duration of OAs, and hypopneas. Does the Dream Station include CAs in that calculation? Are there other events that factor in as well?
Any help would be appreciated.
augustsilverbear

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Amara View Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Pico Nasal Mask with headgear,Mirage Quattro FF with headgear, ResMed chin strap, Pressure: fixed at 11, w/ EPAP relief (Flex)

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: Total Time in Apnea - SH Bar graph

Post by robysue » Sun Apr 03, 2016 12:24 pm

augustsilverbear wrote:How is the SH bar graph for Total Time in Apnea calculated?
Last night I recorded 0 OAs, 2 CAs of 32 seconds duration, and 8 hypopneas of 155 seconds duration. My Total Time in Apnea was recorded as 00:03:07. I assume that is 3 minutes, seven seconds of what my Dream Station calculated as apnea.

I would assume the calculation is the total duration of OAs, and hypopneas. Does the Dream Station include CAs in that calculation? Are there other events that factor in as well?
Any help would be appreciated.
augustsilverbear
Yes, the total time spent in CAs counts in the SH computation of "time in apnea" as does the total time in OAs and the total time in Hs. I don't think anything else counts.

In your case:

3:07 is indeed 3 minutes and 7 seconds, which equals 187 seconds. And 32+155= 187 seconds, so the CA+H time si all that went into the computation on the night in question since there were no OAs.

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

augustsilverbear
Posts: 19
Joined: Sat Feb 27, 2016 3:51 pm
Location: AR

Re: Total Time in Apnea - SH Bar graph

Post by augustsilverbear » Sun Apr 03, 2016 12:35 pm

Thanks.
Curious that CAs count since there is no obstruction there.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Amara View Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Pico Nasal Mask with headgear,Mirage Quattro FF with headgear, ResMed chin strap, Pressure: fixed at 11, w/ EPAP relief (Flex)

User avatar
palerider
Posts: 32306
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Total Time in Apnea - SH Bar graph

Post by palerider » Sun Apr 03, 2016 3:43 pm

augustsilverbear wrote:Thanks.
Curious that CAs count since there is no obstruction there.
why wouldn't they count, aren't they an apnea?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

augustsilverbear
Posts: 19
Joined: Sat Feb 27, 2016 3:51 pm
Location: AR

Re: Total Time in Apnea - SH Bar graph

Post by augustsilverbear » Sun Apr 03, 2016 6:01 pm

Of course they are! My comment was not thought through. I was thinking constriction, not the absence of air flow.