AHI creep

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
simanco
Posts: 5
Joined: Mon Apr 27, 2015 4:20 pm
Contact:

AHI creep

Post by simanco » Tue Mar 29, 2016 9:36 pm

Howdy all,

I started pap therapy a little less than a year ago. We started out with an auto cpap and self titration because I didn't sleep well at the sleep lab.

That was very helpful but something still wasn't right and I "broke down" and went for my titration night. I was prescribed a bipap with 16/20 pressures. My RT programmed in a pressure support of 2 on my Aircurve. It was wonderful. My AHI dropped to under 1 and stayed there from the first night of bipap in September until sometime in December. I had stopped pulling the card and checking things out, so it wasn't until a very sleepy afternoon in January that I really noticed. Apparently I notice daytime sleepiness if my AHI is greater than 1 on a consistent basis. Last night it was 5.

My RT changed my inhalation pressure such that it ranges up to 25. That was very slightly helpful. It has only approached 25 once; it usually tops out around 23 or so.

I upped my exhalation pressure to 17 and am considering raising it some more.

She also got my PCP to order an overnight oximetry and it was normal.

I'll be getting in touch with the RT again and see if she has any further thoughts. The lab has no sleep doc and my RT is at the DME.

In the meantime, does anyone here have any thoughts?

Thank you,
Jim

_________________
MachineMask

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: AHI creep

Post by palerider » Tue Mar 29, 2016 9:44 pm

simanco wrote:In the meantime, does anyone here have any thoughts?
considering everything you've said....

get sleepyhead, post some screenshots

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

simanco
Posts: 5
Joined: Mon Apr 27, 2015 4:20 pm
Contact:

Re: AHI creep

Post by simanco » Tue Mar 29, 2016 10:00 pm

Image

_________________
MachineMask

simanco
Posts: 5
Joined: Mon Apr 27, 2015 4:20 pm
Contact:

Re: AHI creep

Post by simanco » Tue Mar 29, 2016 10:03 pm

Image

_________________
MachineMask

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: AHI creep

Post by palerider » Tue Mar 29, 2016 10:19 pm

zoom in, look at what the lower pressure is when you start to have those clusters of events, and raise the lower pressure (min-epap) to just above that.looks like it might be 18, maybe a bit more, a bit less, hard to tell

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
ChicagoGranny
Posts: 15473
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: AHI creep

Post by ChicagoGranny » Wed Mar 30, 2016 7:30 am

I saw this on the index page and assumed it was about PailRider. Sorry.

User avatar
LSAT
Posts: 13397
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: AHI creep

Post by LSAT » Wed Mar 30, 2016 8:37 am

--1