Breathing with cpap off

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
AMK
Posts: 398
Joined: Mon Feb 23, 2015 8:34 am

Breathing with cpap off

Post by AMK » Mon Mar 28, 2016 10:12 am

I can tell from my data chip that the power went out at 1:20AM but I didn't wake up till 2:30AM. Just a lot of dreams about being unable to breathe as I sucked air through the little mask vent. Didn't get power till 4:30AM. So, as with others here I'm sure, I am here now with a brain that isn't quite working combing through past threads trying to determine what to do to prevent this from happening again. A computer UPS? The "cpap backup power supply" my DME sells for $340? I understand from what I've read that my S10 with humidifier running is a power suck. Is it better to have an alarm that wakes me up and switch off the humidifier rather than have any hope of uninterrupted sleep?

User avatar
Krelvin
Posts: 1977
Joined: Tue Jun 06, 2006 5:23 pm
Location: Metro Phx Area - Dry Heat!

Re: Breathing with cpap off

Post by Krelvin » Mon Mar 28, 2016 10:19 am

I can't sleep when the machine stops, period. Wake up pretty much right away, no alarm needed.

If you search Google: site:cpaptalk.com battery
You will find a ton of links talking about battery power on this forum.

I also have a resource page about batteries include a DYI system based on info from a forum member CapnLoki
https://sleep.tnet.com/resources/batterysolutions
Current Settings PS 4.0 over 10.6-18.0 (cmH2O) - Resmed S9 VPAP Auto w/h5i Humidifier - Quattro Air FFM
TNET Sleep Resource Pages - CPAP Machine Database
Put your equip in your Signature - SleepyHead v1.0.0-beta-1
Kevin... alias Krelvin

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Breathing with cpap off

Post by chunkyfrog » Mon Mar 28, 2016 10:33 am

Bear in mind that before cpap, we all slept--just not very well.
Early on, I was so tired I slept several hours without turning the machine on!
Auto-on is now activated so that bonehead move will not be repeated.
If the power goes off for a short time, the machine turns back on when power is restored.
(Our utility is fairly reliable, and most outages are short)
I don't need to wake up to get it going again.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her
Last edited by chunkyfrog on Mon Mar 28, 2016 10:35 am, edited 1 time in total.

User avatar
AMK
Posts: 398
Joined: Mon Feb 23, 2015 8:34 am

Re: Breathing with cpap off

Post by AMK » Mon Mar 28, 2016 10:34 am

I followed the links on your page Krelvin and it looks to me like these backup batteries are not something that can be left in place in case of an unexpected power outage. That's what I need. I had no idea I would lose power last night.

User avatar
BlackSpinner
Posts: 9742
Joined: Sat Apr 25, 2009 5:44 pm
Location: Edmonton Alberta
Contact:

Re: Breathing with cpap off

Post by BlackSpinner » Mon Mar 28, 2016 10:39 am

AMK wrote:I followed the links on your page Krelvin and it looks to me like these backup batteries are not something that can be left in place in case of an unexpected power outage. That's what I need. I had no idea I would lose power last night.
You need an alarm that will go off when the power goes out. You have the batteries set up and full in an easily accessible place.

Computer UPS are not designed for the power draw of a cpap machine with humidifier and wouldn't give you more then 20-30 minutes at most but most of them beep loudly so you would wake up.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Additional Comments: Quatro mask for colds & flus S8 elite for back up
71. The lame can ride on horseback, the one-handed drive cattle. The deaf, fight and be useful. To be blind is better than to be burnt on the pyre. No one gets good from a corpse. The Havamal

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Breathing with cpap off

Post by chunkyfrog » Mon Mar 28, 2016 10:44 am

Unless there is an intermediate device, the next step up would be a whole house power backup.
Not cheap; but a reasonable investment for someone on a ventilator or an oxygen concentrator.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
LSAT
Posts: 13412
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Breathing with cpap off

Post by LSAT » Mon Mar 28, 2016 10:48 am

AMK wrote:I can tell from my data chip that the power went out at 1:20AM but I didn't wake up till 2:30AM. Just a lot of dreams about being unable to breathe as I sucked air through the little mask vent. Didn't get power till 4:30AM. So, as with others here I'm sure, I am here now with a brain that isn't quite working combing through past threads trying to determine what to do to prevent this from happening again. A computer UPS? The "cpap backup power supply" my DME sells for $340? I understand from what I've read that my S10 with humidifier running is a power suck. Is it better to have an alarm that wakes me up and switch off the humidifier rather than have any hope of uninterrupted sleep?
I think most people using a pillow or nasal mask would automatically open their mouth to breathe when the power goes off...Shouldn't be a struggle.

User avatar
AMK
Posts: 398
Joined: Mon Feb 23, 2015 8:34 am

Re: Breathing with cpap off

Post by AMK » Mon Mar 28, 2016 10:52 am

LSAT, I cannot automatically open my mouth to breathe.

BlackSpinner, yes, that sounds like a plan; I guess there's no affordable way to have sleep be uninterrupted when this happens but at the very least I never want to spend over an hour pulling air through the mask vent again. And Chunkyfrog, you're right; if nothing else I can sleep with no cpap, though I'll be a zombie the next day. Thanks for the replies.

User avatar
Krelvin
Posts: 1977
Joined: Tue Jun 06, 2006 5:23 pm
Location: Metro Phx Area - Dry Heat!

Re: Breathing with cpap off

Post by Krelvin » Mon Mar 28, 2016 11:22 am

AMK wrote:I followed the links on your page Krelvin and it looks to me like these backup batteries are not something that can be left in place in case of an unexpected power outage. That's what I need. I had no idea I would lose power last night.
https://sleep.tnet.com/resources/batter ... collection

Not an automatic switch over system but.. If the power goes off, you just plug your CPAP in to the battery system and go back to sleep.

The battery powered system on that page provides:

35 Amp-Hour system
Charger to keep it charged.
Cables to put it all together.
Without Humidity, you could get upwards of a weeks use on many CPAP machines.
With Humidity, most likely a day or more depending on pressures and your CPAP machine.
Safe to located in your bedroom (no out gassing, and continually on battery tender)

It will last a lot longer than a UPS designed for a computer and let you use your humidifier as well.

Right now I have a generator which can power a lot more (fridge) when we have outages, but I have to go outside to start it up. But I am thinking of building this one for short outages where the generator is overkill.
AMK wrote:LSAT, I cannot automatically open my mouth to breathe.
You are a human? not a fish or something, because humans can breath through their mouths when they are not getting enough air.
Current Settings PS 4.0 over 10.6-18.0 (cmH2O) - Resmed S9 VPAP Auto w/h5i Humidifier - Quattro Air FFM
TNET Sleep Resource Pages - CPAP Machine Database
Put your equip in your Signature - SleepyHead v1.0.0-beta-1
Kevin... alias Krelvin

User avatar
AMK
Posts: 398
Joined: Mon Feb 23, 2015 8:34 am

Re: Breathing with cpap off

Post by AMK » Mon Mar 28, 2016 12:13 pm

Krelvin wrote: You are a human? not a fish or something, because humans can breath through their mouths when they are not getting enough air.
Humans cannot breathe through their mouths when their mouths are taped shut because it's the only cpap mask arrangement they have been able to tolerate. I was hoping to avoid this particular debate, but, there it is.

Thanks for that link, that looks very interesting and I appreciate that all the info is organized on the page!

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Breathing with cpap off

Post by chunkyfrog » Mon Mar 28, 2016 12:24 pm

Chin up strips allow mouth breathing, but make it awkward enough for cpap.
My pattern for my home-made version, to be cut from wide painters tape:
Image

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her
Last edited by chunkyfrog on Mon Mar 28, 2016 12:31 pm, edited 3 times in total.

User avatar
ChicagoGranny
Posts: 15535
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Breathing with cpap off

Post by ChicagoGranny » Mon Mar 28, 2016 12:27 pm

Power failure alarm - $13.69 - http://www.amazon.com/Powerout-Power-Fa ... +off+alarm
AMK wrote:if nothing else I can sleep with no cpap
If you are forced to sleep without CPAP, make sure to sleep on your sides or stomach totally avoiding backsleeping. Typically, OSA is at its worst on the back because gravity is pulling the tongue and soft palate directly into the airway.

lilly747
Posts: 450
Joined: Thu Sep 27, 2012 9:39 am
Location: Northern Calif, USA

Re: Breathing with cpap off

Post by lilly747 » Mon Mar 28, 2016 1:22 pm

ChicagoGranny wrote:Power failure alarm - $13.69 - http://www.amazon.com/Powerout-Power-Fa ... +off+alarm

AMK wrote:
This alarm works for me.

AMK, I too use a Swift FX and TAPE, and use a cervical collar (thanks CG). I do worry about breathing too much CO2 with the machine off...maybe it's OK but not something I want to be doing. I have not been able to find a FFM that works for me...face is too small.....

icipher
Posts: 306
Joined: Thu Oct 10, 2013 11:09 am

Re: Breathing with cpap off

Post by icipher » Mon Mar 28, 2016 1:45 pm

It absolutely blows my mind that our machines don't come included with some sort of battery backup unit.

User avatar
BlackSpinner
Posts: 9742
Joined: Sat Apr 25, 2009 5:44 pm
Location: Edmonton Alberta
Contact:

Re: Breathing with cpap off

Post by BlackSpinner » Mon Mar 28, 2016 1:52 pm

icipher wrote:It absolutely blows my mind that our machines don't come included with some sort of battery backup unit.
So you want to pay at least $500 more for them? For something that is very rarely used? In the almost 7 years I have been using a cpap machine I have had TWO nights where the power went out for over 1 hour - TWO!

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Additional Comments: Quatro mask for colds & flus S8 elite for back up
71. The lame can ride on horseback, the one-handed drive cattle. The deaf, fight and be useful. To be blind is better than to be burnt on the pyre. No one gets good from a corpse. The Havamal