SleepyHead Question
-
Guest
SleepyHead Question
Hey there:
I started using sleepyhead software and when I put SD card in my reader, I put it on write-protect as recommended, and saw data perfectly. But then, when I put it back in my machine and after I turned off write protect, it's now not reading data to the card.
Any thoughts?
Do they lock it if they know we are reading it ourselves?
Do I now need to ask them for new card?
Thanks,
Bob
I started using sleepyhead software and when I put SD card in my reader, I put it on write-protect as recommended, and saw data perfectly. But then, when I put it back in my machine and after I turned off write protect, it's now not reading data to the card.
Any thoughts?
Do they lock it if they know we are reading it ourselves?
Do I now need to ask them for new card?
Thanks,
Bob
Re: SleepyHead Question
What machine? And I am assuming you meant "...now it's not writing data to the card". Not sure which machine you are using but my PRS1 (both 460 and 560 models) never cared if the physical SD card switch was in the locked or unlocked position - the machine always wrote data to the card either way. So try re-seating the card one or two times, gently remove and re-insert once or twice and see if that helps.
As for a new card, there is nothing special about these data cards and you can pick up a small capacity generic SD card 1GB or 2GB for less than $10 as a tester. If it works you've got a spare to deliver to your DME or doctor or any provider than asks for your data.
As for a new card, there is nothing special about these data cards and you can pick up a small capacity generic SD card 1GB or 2GB for less than $10 as a tester. If it works you've got a spare to deliver to your DME or doctor or any provider than asks for your data.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Oracle HC452 Oral CPAP Mask |
-
Diamondminek
- Posts: 100
- Joined: Mon Sep 14, 2015 11:30 am
Re: SleepyHead Question
i think i am having a similar problem, while also breaking card readers at a rate of knots. used it twice fine, went to read it a third time and am only getting summary data - time used and ahi.
_________________
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Re: SleepyHead Question
If you've ever seen the innards of one of these card readers they are often constructed very very minimally and would not be too hard to damage. Try to be gentle with them and use only the minimum force required to remove and insert the cards. It helps to have at least one spare blank card around for testing - without that you really can't know if it's the card, the reader, or the operator's error
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Oracle HC452 Oral CPAP Mask |
Re: SleepyHead Question
When I was using the standard SD card and forgot to turn off the write protection my machines would pop up a card read error or something like that. Even if I remembered to turn off write protection I sometimes had to remove it and turn the switch on and off again.
I since changed to a Flashair/Flashpap combination and been much happier. Any SD card I tried in them worked but I don't know how large a card they will support. The card readers have worked for years for many members but they were not actually designed to remove as often as we tend to do which was one of my motivations for switching to the WIFI card.
I since changed to a Flashair/Flashpap combination and been much happier. Any SD card I tried in them worked but I don't know how large a card they will support. The card readers have worked for years for many members but they were not actually designed to remove as often as we tend to do which was one of my motivations for switching to the WIFI card.
_________________
| Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: PRS1 760 (BiPap Auto set up as APAP) |
Happily hosing since 4/19/2015
Do not believe a word I say until Pugsy or PaleRider has had a chance to review!
Cheers,
Paul
Do not believe a word I say until Pugsy or PaleRider has had a chance to review!
Cheers,
Paul
Re: SleepyHead Question
I've had to slide the slider a few times from lock to unlock to get it to do what I wanted.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Mirage Activa™ LT Nasal CPAP Mask with Headgear |
Sleeping better since 01/13/2016
Masks I have tried
AirFit F10-FFM, AirFit N10-Nasal, Mirage Liberty-Hybrid FFM, Mirage Quattro-FFM, Swift FX Nano-Pillow, Mirage Activa LT-Nasal, ComfortGel Blue-FFM, Amara View -FFM, Pilairo Q -Pillow, Dreamwear -Nasal
Masks I have tried
AirFit F10-FFM, AirFit N10-Nasal, Mirage Liberty-Hybrid FFM, Mirage Quattro-FFM, Swift FX Nano-Pillow, Mirage Activa LT-Nasal, ComfortGel Blue-FFM, Amara View -FFM, Pilairo Q -Pillow, Dreamwear -Nasal
Re: SleepyHead Question
Some computer operating systems will write a file to the SD card. The S9s do not like that and will require a reformat of the card.
The card readers can be delicate and easily damaged.
The card readers can be delicate and easily damaged.
_________________
| Mask: SleepWeaver 3D Soft Cloth Nasal CPAP Mask with Headgear |
| Additional Comments: AurCurve 10 ASV Also using Sleaplyhead 1.1, ResScan 6 and CMS50i |
Re: SleepyHead Question
If you are getting summary data, then the machine is indeed writing to the SD card.Diamondminek wrote:i think i am having a similar problem, while also breaking card readers at a rate of knots. used it twice fine, went to read it a third time and am only getting summary data - time used and ahi.
It sounds almost as if you didn't fully insert the card into your machine before using it. And then at some point, noticed the card was not fully inserted and pushed it in all the way. At that point the CPAP would have written the summary data to the SD card.
In other words: Machines that record flow rate data record that data directly to the SD card. And Resmed machines record all detailed data directly to the SD card. If the card is not in the machine or if it is not fully inserted into the machine, the CPAP can't record the detailed data to the card and that data is lost.
_________________
| Machine: DreamStation BiPAP® Auto Machine |
| Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5 |
Re: SleepyHead Question
no they don't, they can't know you're reading it.Guest wrote:Do they lock it if they know we are reading it ourselves?
Do I now need to ask them for new card?
unless your card has gone bad, just reformat it and then let the cpap do what it wants to with the card.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: SleepyHead Question
sounds like you forgot to put the card back in the machine properly, data is written directly to the card, and if you forget to push it in till it clicks, then the detailed data can't be written, but the next time you do push it in good, the compliance data will be written from the machines memory (that's all the machine stores).Diamondminek wrote:i think i am having a similar problem, while also breaking card readers at a rate of knots. used it twice fine, went to read it a third time and am only getting summary data - time used and ahi.
like Robysue said.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: SleepyHead Question
that doesn't do anything. the sider is just a piece of plastic, like the old 'write protect' sliders on 3.5" floppies. there's no mechanical, or electrical connection in the card, all it does is trip a switch in the reader itself.Mr Nifty wrote:I've had to slide the slider a few times from lock to unlock to get it to do what I wanted.
just taking the card out and putting it back in is the only thing that actually did anything.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: SleepyHead Question
One must be somewhat careful with the SD card. The memory in the SD card is fairly immune to damage but you should keep your fingers off of the metal contact on the card. Walking on a wool carpet can generate thousands of volts of electricity that can destroy the SD card. Electrostatic discharge (the spark that shocked somone) can occur when you insert the card in a card reader but is less problematic. I try to make sure that I touch a bare metal part of the computer to discharge any potential charge. Damage is most often done by touching the metal contacts.
I am fimiliar with this problem because I am an electronic engineer and also use SD cards in photograpy and have "sucessfully" destroyed a few due to improper handling. Not often but it does occur. If card was damaged you would get nothing written in the CPAP and would not be capable of reading the card on your computer. I am not sure how your CPAP reacts to a damaged card. The Airsense 10 shows "preparing card..." after SD has been removed and reinserted. I do not know if it would flag a bad card or not.
I am fimiliar with this problem because I am an electronic engineer and also use SD cards in photograpy and have "sucessfully" destroyed a few due to improper handling. Not often but it does occur. If card was damaged you would get nothing written in the CPAP and would not be capable of reading the card on your computer. I am not sure how your CPAP reacts to a damaged card. The Airsense 10 shows "preparing card..." after SD has been removed and reinserted. I do not know if it would flag a bad card or not.
_________________
| Machine: AirSense 11 Autoset |
| Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
| Additional Comments: Jan 17, 2016 Split Sleep study results w/o CPAP AHI=83, W/CPAP AHI=29. Pressure in APAP 15-20 cm |
The only way to keep your health is to eat what you don't want, drink what you don't like, and do what you'd rather not. Mark Twain
Was Mark Twain a CPAP user?
Was Mark Twain a CPAP user?
Re: SleepyHead Question
Also worth noting that a static discharge that is too small for one to detect with tactile sensation or visually is sufficient to damage many electronic components. And if damaged in this way components may not fail immediately or in any predictable way. Always ground yourself, esp in winter or other low humidity environments, prior to getting anywhere near these things. If unsure touch a metal wall plate to ground yourself, or the metal screw holding a wall plate on to an outlet or switch (assuming your electrical system is properly grounded). And as stated above never touch the metal contactsEddy wrote:One must be somewhat careful with the SD card. The memory in the SD card is fairly immune to damage but you should keep your fingers off of the metal contact on the card. ...
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Oracle HC452 Oral CPAP Mask |