please help

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
psantiago

please help

Post by psantiago » Fri Feb 26, 2016 9:04 pm

hi everyone ......i'm a newwbie and i just wanna ask a question.........i work nights and i have been using my machine (DreamStation Series Heated Humidifier) every morning when i get home at 7am whats happening is if i sleep from 730 am to 330 pm = 8 hours the machine is saying 3 1/2 hours of sleep time is there a time shift as in like midnight starts a new day on our clocks but the machine considers a new day at 12pm since most people sleep at night.......i dont know what to do and i dont want my medical ins or the medical company to think im not using the machine the recamended 4 hours per day......please help........ i also had this same issuse with resmed 10 and thought it was broken and sent it back

User avatar
Julie
Posts: 20049
Joined: Tue Feb 28, 2006 12:58 pm

Re: please help

Post by Julie » Fri Feb 26, 2016 9:10 pm

Someone here might know how to set the clock properly for you, but if not, maybe the DME might help... have you asked? Or called Resmed to ask?

Sleeprider
Posts: 1562
Joined: Tue May 06, 2008 5:57 pm
Location: Murrysville, PA

Re: please help

Post by Sleeprider » Fri Feb 26, 2016 9:18 pm

The therapy meter breaks at noon. Your total daily hours will be correct, but your particular sleep schedule gets broken into two days. If you are required to demonstrate compliance, be sure to explain this to your DME, but the total usage hours per day will be correct on average. As far as I know there is no way to change how the therapy session is recorded into two days instead of one, but someone may know differently.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleepyhead software. Just changed from PRS1 BiPAP Auto DS760TS

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: please help

Post by Wulfman... » Fri Feb 26, 2016 9:18 pm

psantiago wrote:hi everyone ......i'm a newwbie and i just wanna ask a question.........i work nights and i have been using my machine (DreamStation Series Heated Humidifier) every morning when i get home at 7am whats happening is if i sleep from 730 am to 330 pm = 8 hours the machine is saying 3 1/2 hours of sleep time is there a time shift as in like midnight starts a new day on our clocks but the machine considers a new day at 12pm since most people sleep at night.......i dont know what to do and i dont want my medical ins or the medical company to think im not using the machine the recamended 4 hours per day......please help........ i also had this same issuse with resmed 10 and thought it was broken and sent it back
Most all CPAP machines (by default) consider a "day" as from Noon (first day) to Noon (next day). Everything between those times counts towards the day's total compliance hours.
What software are you using? Some of them can be adjusted for different usage hours.
But, even if it's split like you mentioned, the next day will have accumulated hours that should add together for your compliance hours.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
OkyDoky
Posts: 2870
Joined: Mon Aug 25, 2014 5:18 pm

Re: please help

Post by OkyDoky » Fri Feb 26, 2016 9:25 pm

psantiago wrote:if i sleep from 730 am to 330 pm = 8 hours the machine is saying 3 1/2 hours of sleep time
Your machine will show all your hours. It will just be broken into 2 separate sessions in 24 hours. With this example you will have 3 1/2 hours one afternoon plus 4 1/2 hours the next morning equal 8 hours in 24 hours. The compliance hours will still show 8 hours in 24 hours. If you get Sleepyhead computer software then you can see that. I wouldn't worry about setting the clock different it will all add up.
This is a site to help you learn about Sleepyhead. https://sleep.tnet.com/resources/sleepyhead
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: please help

Post by palerider » Fri Feb 26, 2016 9:27 pm

Julie wrote:Someone here might know how to set the clock properly for you, but if not, maybe the DME might help... have you asked? Or called Resmed to ask?
dreamstation = respironics, not resmed, and you can't set the clock on respironics machines.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Blizardj
Posts: 3
Joined: Mon Feb 22, 2016 10:41 am

Re: please help

Post by Blizardj » Fri Feb 26, 2016 9:39 pm

True you can't set the internal clock... but the Dreamstation can be set at 30 minute increments for local time zone setting.

John

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: please help

Post by palerider » Fri Feb 26, 2016 9:58 pm

Blizardj wrote:True you can't set the internal clock... but the Dreamstation can be set at 30 minute increments for local time zone setting.
which is kinda weird... instead of letting you set the clock, which drifts, unless they're doing some correction off the cell network now...

extra credit, why can you set the time zone in half hour increments?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
OkyDoky
Posts: 2870
Joined: Mon Aug 25, 2014 5:18 pm

Re: please help

Post by OkyDoky » Fri Feb 26, 2016 10:09 pm

palerider wrote:
Blizardj wrote:True you can't set the internal clock... but the Dreamstation can be set at 30 minute increments for local time zone setting.
which is kinda weird... instead of letting you set the clock, which drifts, unless they're doing some correction off the cell network now...

extra credit, why can you set the time zone in half hour increments?
It does say this in the user manual. Seems strange it's not in 1 hr increments.

Allows you to adjust the time. The default setting is Greenwich Mean Time, but you may adjust the time in 30 minute increments to match your local time zone.
Note: This time setting is not displayed as a clock function on the device. It is only used to align your therapy data Ior your Provider's data reports.
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: please help

Post by palerider » Fri Feb 26, 2016 10:32 pm

OkyDoky wrote:
palerider wrote:
Blizardj wrote:True you can't set the internal clock... but the Dreamstation can be set at 30 minute increments for local time zone setting.
which is kinda weird... instead of letting you set the clock, which drifts, unless they're doing some correction off the cell network now...

extra credit, why can you set the time zone in half hour increments?
It does say this in the user manual. Seems strange it's not in 1 hr increments.
aye, that's the 'extra credit' bit, why half hour increments?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
poppi2
Posts: 590
Joined: Thu Nov 18, 2010 2:54 pm
Location: Houston, near JSC

Re: please help

Post by poppi2 » Fri Feb 26, 2016 10:42 pm

"North Korea, Newfoundland, India, Iran, Afghanistan, Venezuela, Burma, Sri Lanka, the Marquesas, as well as parts of Australia use half-hour deviations from standard time, and some nations, such as Nepal, and some provinces, such as the Chatham Islands, use quarter-hour deviations." Via Wikipedia

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: PAPCap, 3M Microfoam Surgical Tape, PoliGrip Strip, APAP 12.0 - 14.0 cm

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: please help

Post by palerider » Fri Feb 26, 2016 11:06 pm

poppi2 wrote:"North Korea, Newfoundland, India, Iran, Afghanistan, Venezuela, Burma, Sri Lanka, the Marquesas, as well as parts of Australia use half-hour deviations from standard time, and some nations, such as Nepal, and some provinces, such as the Chatham Islands, use quarter-hour deviations." Via Wikipedia


and, they keep dorking around with 'em, based on the frequency of timezone file updates I see going into unix

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: please help

Post by chunkyfrog » Sat Feb 27, 2016 12:40 am

All the product of somebody's consuming desire to leave their "mark" on the world--even if it is a black mark.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her