Which ASV machine can be changed to Auto or cpap mode?
Respironics BiPAP AutoSV Advanced
PR System One AutoSV Advanced
Resmed s9 adapt sv
Resmed AirCurve 10 A
Which ASV machine
Re: Which ASV machine
The s9 adapt 36037 (not lower ref#), if I recall correctly.
Not sure about the respironics 960, but I can check it later. Don't know about the first one and last one on your list. Not sure why the asv aircurve 10 wouldn't but I don't have one so I don't know.
Not sure about the respironics 960, but I can check it later. Don't know about the first one and last one on your list. Not sure why the asv aircurve 10 wouldn't but I don't have one so I don't know.
_________________
| Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier |
| Additional Comments: Resmed Aircurve 10 ASVAUTO Min Epap 10.4, Max Epap 11.6, PS 1.6-12.0 |
Re: Which ASV machine
These are the pertinent pressure parameters on the PR 960
Parameter________________Pressure Range
Max pressure______________4 - 25.0
EPAP min_________________4 - Max pres
EPAP max__________________EPAP min - Max Pres
PS min____________________0 - (Max Pres - Max EPAP)
PS max____________________PS min - (Max Pres - Min EPAP)
Without having set my actual machine to see if will accept the following,
it may be that to get a BiPap machine with pressures of 6/4:
set the following:
Max Pressure = 6
EPAP Min = 4
EPAP Max = 4
PS Min = 2
Max PS = 2
That should make it into a straight bi-level machine with IPAP=6 and EPAP=4
Parameter________________Pressure Range
Max pressure______________4 - 25.0
EPAP min_________________4 - Max pres
EPAP max__________________EPAP min - Max Pres
PS min____________________0 - (Max Pres - Max EPAP)
PS max____________________PS min - (Max Pres - Min EPAP)
Without having set my actual machine to see if will accept the following,
it may be that to get a BiPap machine with pressures of 6/4:
set the following:
Max Pressure = 6
EPAP Min = 4
EPAP Max = 4
PS Min = 2
Max PS = 2
That should make it into a straight bi-level machine with IPAP=6 and EPAP=4
_________________
| Mask: Oracle HC452 Oral CPAP Mask |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV |
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
Re: Which ASV machine
all resmed machines have a 'cpap' mode.danielqk wrote:Which ASV machine can be changed to Auto or cpap mode?
Respironics BiPAP AutoSV Advanced
PR System One AutoSV Advanced
Resmed s9 adapt sv
Resmed AirCurve 10 A
also, there's no such thing as a resmed s9 adapt sv. it's the resmed s9 vpap adapt, and aircurve 10 asv.