mrje1
Posts: 20
Joined: Sat Nov 07, 2015 12:49 pm
Location: Chicago, Illinois

Suffering Need help.

Post by mrje1 » Sat Nov 07, 2015 1:34 pm

Hey everyone,

I am so grateful to have found this website. I am hoping you guys can give me some insight because I am sick and tired of being sick and tired. The doctor's are just not finding the answers or keep screwing me up. I have to keep waiting for weeks for my next appointment. can't take it any longer. It is totally interfering with life. I will describe my issue as best as I can and can add more as I go along with this topic or as you guys ask questions. Let me know of any information you need and I will do my best find it or answer.

I have been diagnosed with central and obstructive sleep apnea. I took the sleep study in summer of 2014 and another one in September 2015 with another doctor. I also have problems with weight. I weigh 260 pounds and currently losing weight and getting back into shape. I have been struggling with this for the past year where I keep on waking up non-refreshed and constantly feeling a non functioning fatigue and tiredness which causes me to lay down multiple times a day. Some days vary than others though where it might not be as bad etc. I have been complaining about this to the doctor over the past year, gave me some answers that I don't recall right now, but no resolve. There were spurts where I had to let it go and delt with it while I was heavily distracted by other things. Now I am really attacking this because I just can't get a quality life right now. I also can't get 7 or 8 hours of sleep. I am averaging from 4-6 hours as well.

I am now resorting to messing with the settings on my machine because I refuse to wait week after week for the doctors to come up with an answer. If anybody has any thoughts, questions etc. please let me know. I am hoping to have this resolved and looking forward to a better quality life again.

I thank you all for your help. I really do appreciate it.


HERE IS MY EQUIPMENT INFO
Machine: PR BiPAP AutoSV S1 960
Humidifier: PR System One 60 Series Heated tubing
Mask: Resmed Micro Nasal CPAP Mask

HERE IS THE LINK TO MY SLEEP STUDIES & MACHINE REPORT
http://1drv.ms/1GRnpMS

HERE IS MY CURRENT MACHINE SETTINGS. MAYBE THIS WILL BE OF USE.

Max Pressure = 25.0
EPAP Min = 6.0
EPAP Max = 12.0
PS Min = 8.0
PS Max = 14.0
BPM = Auto
Flex Type = None
Rise Time = 3
Ramp Time = 0.30 Minutes
Ramp Start = 5.0
Tubing Type Lock = Off
Tubing Type = 15h
System One Resistance = 0
Lock System One = Off
Heated Tube Humidification = On
Humidity Level = 3
Tube Temperature = 3
Auto on = On
Patient Disconnect Alarm = 0
Apnea Alarm = 0
Low Min Vent Alarm = 0
Mask Fit Check = Off
Humidifier LED Backlight = Off
Show AHI/leak/PB = Off
Setup Parameter Displayed = None
Language = EN
Last edited by mrje1 on Mon Nov 09, 2015 10:37 am, edited 3 times in total.

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Suffering Need help.

Post by Julie » Sat Nov 07, 2015 1:48 pm

Hi - what machine (fullest name & model # you have) are you using? And which mask?

User avatar
Pugsy
Posts: 65327
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Suffering Need help.

Post by Pugsy » Sat Nov 07, 2015 1:53 pm

Welcome to the forum.

So it looks like you have a PR System One model 960 machine so it will works with Sleepyhead or Encore.
The fact that you mention "system one resistance" and the min epap and max ipap means a PR S1 960 machine as it is the only one that does min and max EPAP in the Respironics line.

Can you post an image of a typical daily report?
How to post images
https://sleep.tnet.com/reference/tips/imgur
examples of what we like to see are shown in this thread
If you don't know about SleepyHead
https://sleep.tnet.com/equipment

How come you can't get more than 4 to 5 hours of sleep?
Do you take any meds of any kind for any reason?
Do you have any other health issues?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
metsfan302
Posts: 687
Joined: Fri Oct 09, 2015 9:44 pm
Location: Dewey Beach, Delaware

Re: Suffering Need help.

Post by metsfan302 » Sat Nov 07, 2015 2:02 pm

Welcome to the Forum!

You found a great group of helpful people.....

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Prior - ResMed AirCurve ASV User
This is my Orig Post "The start of my venture to find proper therapy" - http://www.cpaptalk.com/viewtopic.php?f ... &sk=t&sd=a

mrje1
Posts: 20
Joined: Sat Nov 07, 2015 12:49 pm
Location: Chicago, Illinois

Re: Suffering Need help.

Post by mrje1 » Sat Nov 07, 2015 2:17 pm

Julie wrote:Hi - what machine (fullest name & model # you have) are you using? And which mask?
Hi Julie. Thank you for your question. The details are in my profile. Are they not displayed? I am using the Respironics System One bipap autoASV with the ResMed Mirage Micro Mask.

mrje1
Posts: 20
Joined: Sat Nov 07, 2015 12:49 pm
Location: Chicago, Illinois

Re: Suffering Need help.

Post by mrje1 » Sat Nov 07, 2015 2:19 pm

metsfan302 wrote:Welcome to the Forum!

You found a great group of helpful people.....
Thank you so much metsfan302. I can use it!

mrje1
Posts: 20
Joined: Sat Nov 07, 2015 12:49 pm
Location: Chicago, Illinois

Re: Suffering Need help.

Post by mrje1 » Sat Nov 07, 2015 2:28 pm

Pugsy wrote:Welcome to the forum.

So it looks like you have a PR System One model 960 machine so it will works with Sleepyhead or Encore.
The fact that you mention "system one resistance" and the min epap and max ipap means a PR S1 960 machine as it is the only one that does min and max EPAP in the Respironics line.

Can you post an image of a typical daily report?
How to post images
https://sleep.tnet.com/reference/tips/imgur
examples of what we like to see are shown in this thread
If you don't know about SleepyHead
https://sleep.tnet.com/equipment

How come you can't get more than 4 to 5 hours of sleep?
Do you take any meds of any kind for any reason?
Do you have any other health issues?
Hi Pugsy. Thank you very much for your questions etc. I posted my equipment info in my profile, but for some reason you guys can't see it. Is there away to show it? I have the Respironics System one Bipap Auto ASV.

This report, is this from the machine that the doctor's download? If so, I will look into the report you are looking for, but give me some time maybe a day. When it comes to sleep, I have no clue right now why I am only getting 4-6 hours of sleep. I trying to find the answers. The only answer I got was because I am laying down so much during the day that it is causing me to possibly get less sleep at night. I remember a few times when I didn't lay down that much for the day and I still only got 4-6 hours.

No, I don't take any meds for anything.

Health issues as of right now is obesity, sleep apnea, acid reflux, sinus issues and some anxiety. Anything else I forget I will let you know.

User avatar
Pugsy
Posts: 65327
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Suffering Need help.

Post by Pugsy » Sat Nov 07, 2015 2:35 pm

Nothing at all is showing in your profile..now I know the machine link is broken but the humidifier and mask should show up. Did you click submit after making your choices?
If it won't show up...just manually add to the comments section.
You can keep it brief...it's PR S1 960 ASV model.

I don't know what report your doctor gave you so I don't know if that's what we would want to see or not.
Your machine is full data machine and you can generative your own reports.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

mrje1
Posts: 20
Joined: Sat Nov 07, 2015 12:49 pm
Location: Chicago, Illinois

Re: Suffering Need help.

Post by mrje1 » Sat Nov 07, 2015 3:01 pm

I put my equipment info inside my topic post.

mrje1
Posts: 20
Joined: Sat Nov 07, 2015 12:49 pm
Location: Chicago, Illinois

Re: Suffering Need help.

Post by mrje1 » Sat Nov 07, 2015 3:04 pm

Pugsy wrote:Nothing at all is showing in your profile..now I know the machine link is broken but the humidifier and mask should show up. Did you click submit after making your choices?
If it won't show up...just manually add to the comments section.
You can keep it brief...it's PR S1 960 ASV model.

I don't know what report your doctor gave you so I don't know if that's what we would want to see or not.
Your machine is full data machine and you can generative your own reports.
I thought I did hit submit. I will do it again. Where would that information show up? I placed the information in my topic post anyway.

Aweseome! I will learn how to get those reports. Thank you!

User avatar
OkyDoky
Posts: 2869
Joined: Mon Aug 25, 2014 5:18 pm

Re: Suffering Need help.

Post by OkyDoky » Sat Nov 07, 2015 3:05 pm

Pugsy, I'm looking and it looks like no one's equipment profile is showing up for some reason. Must be a site issue.
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Suffering Need help.

Post by chunkyfrog » Sat Nov 07, 2015 3:10 pm

Oh, rats!

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Pugsy
Posts: 65327
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Suffering Need help.

Post by Pugsy » Sat Nov 07, 2015 4:21 pm

OkyDoky wrote:Pugsy, I'm looking and it looks like no one's equipment profile is showing up for some reason. Must be a site issue.
Rats...didn't even notice. I think the website has had some issues today and yesterday and in the past when that happens everyone's profile seems to go AWOL. Probably will get sorted out next week when the IT guys get back to work. May take a little while. Eventually they will show up but the machine link will still be broken.
That's okay...OP has a 960 I am certain because of the settings available in the menu that was quoted.
No other ASV has those setting options.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Pugsy
Posts: 65327
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Suffering Need help.

Post by Pugsy » Sat Nov 07, 2015 4:23 pm

Until you have the software up and running how about posting the limited data on the machine's LCD screen?

If you don't know about the data on the LCD screen or it isn't accessible (has to be set to "Show AHI" for you to see it...check out this link and the video.
http://www.apneaboard.com/pr-system-one ... structions

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

mrje1
Posts: 20
Joined: Sat Nov 07, 2015 12:49 pm
Location: Chicago, Illinois

Re: Suffering Need help.

Post by mrje1 » Sat Nov 07, 2015 7:14 pm

Pugsy wrote:Until you have the software up and running how about posting the limited data on the machine's LCD screen?

If you don't know about the data on the LCD screen or it isn't accessible (has to be set to "Show AHI" for you to see it...check out this link and the video.
http://www.apneaboard.com/pr-system-one ... structions
Ok. Besides the data I posted in my original post, here is additional settings. If I am missing anything let me know. Thank you!

HUMIDIFIER
80W
22
3

AVERAGE PS
1 DAY: 9.7
7 DAY: 11.3
30 DAY: 9.7

90% EPAP
1 DAY: 6.0
7 DAY: 7.1
30 DAY: 8.1

PERIODIC BREATHING
1 DAY, 7 DAY, 30 DAY: 0%

AHI
1 DAY 0.0
7 DAY: 3.8
30 DAY: 5.3

LARGE LEAK
1 DAY & 7 DAY: 0%
30 DAY: 5%

DAYS >4
1 DAY: 0
7 DAY: 7
30 DAY: 30