PR System One Humidifier Popping Noise

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
filovirus
Posts: 28
Joined: Sun Apr 14, 2013 6:44 pm

PR System One Humidifier Popping Noise

Post by filovirus » Sun Aug 02, 2015 9:24 am

My PRS One is making a popping noise at the beginning of inspiration and expiration. It sounds like a diaphragm closing and seems to come from the humidifier unit. I cannot find any source and when I bypassed the humidifier the noise is gone.

Any ideas on this issue?

Thanks.

User avatar
Pugsy
Posts: 65399
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: PR System One Humidifier Popping Noise

Post by Pugsy » Sun Aug 02, 2015 9:33 am

Take the water chamber out and look in the back for the little clear silicone gasket things and see if one of them maybe has come loose.
I think there are 2 of them.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
filovirus
Posts: 28
Joined: Sun Apr 14, 2013 6:44 pm

Re: PR System One Humidifier Popping Noise

Post by filovirus » Sun Aug 02, 2015 9:37 am

I see three of them. One on top one that interfaces with the water chamber and one that interfaces with the main unit. They look worn and the top one has signs of mold. I am taking them out and washing them. Also ordering replacements from cpap.com. None of them seemed loose.

I don't see anything else that could cause the issue so maybe washing and replacing will help. I'll know more once its all back together.

User avatar
Sir NoddinOff
Posts: 4189
Joined: Mon May 14, 2012 5:30 pm
Location: California

Re: PR System One Humidifier Popping Noise

Post by Sir NoddinOff » Sun Aug 02, 2015 9:43 am

Sounds like you found a problem... hope that works. I sometimes get that, but there are typically only a few pops. First of all, I turn off the machine then I pull out the water tank to check if the lid is seated correctly. There's enough play in the 2 hinges on the back side of the tank to allow the lid to be clamped down unevenly (a misaligned lid/gasket connection, IOW) These days I don't even have to turn on the light, I can simply tell by feeling around with my fingers if the tank lid is snapped down evenly all around. Of course, if the popping persists, try what Pugsy suggested... the tank gasket and the three round silicone seals, can be pulled out. There's a latch on the bottom of the humidifier that allows the whole assembly to be removed.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2.
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.

User avatar
Wulfman...
Posts: 6689
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: PR System One Humidifier Popping Noise

Post by Wulfman... » Sun Aug 02, 2015 8:30 pm

filovirus wrote:My PRS One is making a popping noise at the beginning of inspiration and expiration. It sounds like a diaphragm closing and seems to come from the humidifier unit. I cannot find any source and when I bypassed the humidifier the noise is gone.

Any ideas on this issue?

Thanks.
Lots of times people have reported similar popping noises and then found that they had their HH setting too high and there is water vapor condensing into liquid water in a low spot the hose and the air has difficulty getting past the obstruction.

Lower HH settings or keeping the hose and machine positioned so that a low spot won't occur so any excess water runs back into the HH tank may help.