Repairing Small Hole in Heated Tube?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Jewel
Posts: 18
Joined: Mon Apr 20, 2015 3:20 pm
Location: Ottawa, Ontario

Repairing Small Hole in Heated Tube?

Post by Jewel » Wed Jul 29, 2015 9:39 pm

I just recently purchased my machine and hose after trying things out for a bit with my RT.

I'm using an AirSense 10 AutoSet for Her with a heated tube. I had it at home for only a couple of weeks when my cat decided to see how the tube would taste. So there's a small puncture and a few other tooth marks on it. With the puncture, I can't use is as it's leaking air loudly.

Has anyone every tried to repair a very small hole in a heated tube? Any recommendations on how best to do this?

Thanks.

P.S. The cat has sworn never to do this again.

_________________
Mask: Amara View Full Face CPAP Mask with Headgear
Additional Comments: Testing the Amara View Full Face Mask and just picked up my Airsense 10 AutoSet for Her machine!

User avatar
grayghost4
Posts: 1554
Joined: Wed Aug 20, 2014 6:52 pm
Location: Norther Illinois
Contact:

Re: Repairing Small Hole in Heated Tube?

Post by grayghost4 » Wed Jul 29, 2015 10:02 pm

If you're not part of the solution you're just scumming up the bottom of the beaker!

Get the Clinicians manual here : http://apneaboard.com/adjust-cpap-press ... tup-manual

yaconsult
Posts: 1099
Joined: Fri Nov 14, 2014 10:43 pm
Location: "Silicon Valley", CA

Re: Repairing Small Hole in Heated Tube?

Post by yaconsult » Wed Jul 29, 2015 10:44 pm

Wrap it in tape to seal it. Pretty much any durable tape should do - electrical, painter's, duct, etc. Check it for leaks afterwards. Because of the spiraling, you want to make sure you get a good seal and that it doesn't leak where the tape ends.

The hose doesn't get very warm at all so the temperature shouldn't be a problem. You can decide for yourself if you want to try another solution, but this should get you back in business for tonight.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: January 2015 Sleep Study Results: Apnea/Hypopnea Index (AHI): 80.2, Sleepyhead

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Repairing Small Hole in Heated Tube?

Post by palerider » Wed Jul 29, 2015 10:51 pm

Jewel wrote:P.S. The cat has sworn never to do this again.
the cat lies. just consider that shifty look it had when it made that promise.

I'd probably use a bit of tape... and keep more handy

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Repairing Small Hole in Heated Tube?

Post by Wulfman... » Wed Jul 29, 2015 11:06 pm

Yeah, I would agree that something like "Duck Tape" (or duct tape) would get you by for the time being or longer.
Two things I always keep handy are the aforementioned tape and Krazy Glue.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
distantdreamer
Posts: 15
Joined: Mon Mar 09, 2015 10:59 pm

Re: Repairing Small Hole in Heated Tube?

Post by distantdreamer » Wed Jul 29, 2015 11:53 pm

I have two kitties, one of them demolished a few tubes on my old machine. Of course I got to a point where I didn't want to keep buying tubing so I just got some heavy duty duct tape and taped it up really good. Seemed to work.

Since I have gotten a new machine with a heated hose I have been much better about keeping it out of paws reach when not in use. I unhook the hose every morning and place it, along with my mask, in my nightstand drawer and haven't had any more little holes poked where they shouldn't be.

SewTired
Posts: 1736
Joined: Thu Apr 16, 2015 8:33 am
Location: Minneapolis area

Re: Repairing Small Hole in Heated Tube?

Post by SewTired » Thu Jul 30, 2015 7:23 am

Flexible putty followed by electrical tape is what we used on bro's hose awhile back.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleepyhead
Diabetes 2, RLS & bradycardia
Airsense For Her; Settings: range 8-12, Airfit P10 (M)

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: Repairing Small Hole in Heated Tube?

Post by JDS74 » Thu Jul 30, 2015 11:04 am

One other suggestion is to go the pharmacy and get a small syringe and needle. Put some silicone sealant in the syringe, and inject just a tiny bit into the holes with just a little bit on the outside as well. When the sealant hardens, it will form a plug with a supporting head inside and out that will keep it from falling off. Clear silicone will be almost invisible and won't interfere with the hose flexibility.

Just a thought.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

User avatar
Pugsy
Posts: 65395
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Repairing Small Hole in Heated Tube?

Post by Pugsy » Thu Jul 30, 2015 11:12 am

Someone once told me that they repaired a tiny pin hole sized hole in their hose with fingernail polish.
I don't know that I would try that but thought I would mention it.
For myself if my cats from hell put a hole in my hose I would probably just duct tape over the hole as long as it still seemed to heat up like it is supposed to.
Fortunately my cats from hell don't seem to care about cpap stuff...they like power cords and rubber modem stuff instead...which unfortunately costs a lot more than a heated hose.
So far no fried kitties but I keep telling them to leave the power cords alone or they will end up fried kitty.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

HoseCrusher
Posts: 2744
Joined: Tue Oct 12, 2010 6:42 pm

Re: Repairing Small Hole in Heated Tube?

Post by HoseCrusher » Thu Jul 30, 2015 12:31 pm

Having extensive experience repairing a wide variety of hose leaks I find that Rescue Tape works very well. It is easy to apply and lasts a long time.

http://www.amazon.com/Rescue-Tape-RT100 ... escue+tape

_________________
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier.
SpO2 96+% and holding...

ClayL
Posts: 369
Joined: Mon Feb 12, 2007 8:45 am
Location: Palisade, CO now, was Full time in a motor home for 12 years- Anywhere USA

Re: Repairing Small Hole in Heated Tube?

Post by ClayL » Thu Jul 30, 2015 2:07 pm

HoseCrusher wrote:Having extensive experience repairing a wide variety of hose leaks I find that Rescue Tape works very well. It is easy to apply and lasts a long time.

http://www.amazon.com/Rescue-Tape-RT100 ... escue+tape
That is what I would use. You can also get it at Home Depot or Lowes. It is a stretchy self amalgamating tape that sticks only to itself. Works on radiator hoses and all sorts of stuff.
Respironics DreamStation
Pressure = 11.5 min 14.5 max
C-Flex = 1

Dreamwear FF

User avatar
teachcsg
Posts: 189
Joined: Sun Mar 11, 2012 2:39 pm

Re: Repairing Small Hole in Heated Tube?

Post by teachcsg » Thu Jul 30, 2015 2:16 pm

I see a lot of patients using duct tape.
Cesar Garza, RRT
Registred Respiratory Therapist
http://www.lastminuteceus.com

Jewel
Posts: 18
Joined: Mon Apr 20, 2015 3:20 pm
Location: Ottawa, Ontario

Re: Repairing Small Hole in Heated Tube?

Post by Jewel » Thu Jul 30, 2015 2:50 pm

Thank you to everyone for your advice. It's very much appreciated.

Can I use duct tape on the cat?

_________________
Mask: Amara View Full Face CPAP Mask with Headgear
Additional Comments: Testing the Amara View Full Face Mask and just picked up my Airsense 10 AutoSet for Her machine!

library lady
Posts: 1116
Joined: Wed Feb 05, 2014 11:10 am
Location: Rochester, MN

Re: Repairing Small Hole in Heated Tube?

Post by library lady » Thu Jul 30, 2015 3:31 pm

If the cat is really naughty, you might also get a hose cover... go here: http://www.padacheek.com . Then it will be harder for the cat to the hose, and it's easier to repair a hole in fabric than in the hose.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments:  Sleepyhead
Now using AirFit F10 mask; Quattro Air is backup mask. RemZzzz mask liners with both.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Repairing Small Hole in Heated Tube?

Post by palerider » Thu Jul 30, 2015 6:19 pm

Jewel wrote:Thank you to everyone for your advice. It's very much appreciated.

Can I use duct tape on the cat?