BIPAP vs CPAP
BIPAP vs CPAP
Hi everyone - so I am having trouble getting my AHI to a reasonable level with my machine. At my last sleep doctor appointment she mentioned I may have to switch to a different machine, which I would assume would be a BIPAP.
I don't know anything about them, what are the advantages to them? Are they simply for people who require a higher pressure? I am currently on min of 17 and max of 20 and am still having high numbers. Would a BIPAP be the best solution for me?
Joel
I don't know anything about them, what are the advantages to them? Are they simply for people who require a higher pressure? I am currently on min of 17 and max of 20 and am still having high numbers. Would a BIPAP be the best solution for me?
Joel
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear |
Re: BIPAP vs CPAP
Bipap machines provide a lower pressure during the exhale portion of the breathing cycle. With the higher pressures, some people find it hard to exhale against the higher pressures.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Simplus Full Face CPAP Mask with Headgear |
Re: BIPAP vs CPAP
I'm sure it would help. Not sure what the protocol is though.joeljjk11 wrote:I am currently on min of 17 and max of 20
My initial RX years ago was for 20/27. There was no question about using a Bilevel then. At the time even normal machines didn't go that high so I got cranky ST unit until I got the level reduced to 16/20 three weeks later and got a normal Bilevel which I used several years after then, which is lower than your current pressures.
Current Settings PS 4.0 over 10.6-18.0 (cmH2O) - Resmed S9 VPAP Auto w/h5i Humidifier - Quattro Air FFM
TNET Sleep Resource Pages - CPAP Machine Database
Put your equip in your Signature - SleepyHead v1.0.0-beta-1
Kevin... alias Krelvin
TNET Sleep Resource Pages - CPAP Machine Database
Put your equip in your Signature - SleepyHead v1.0.0-beta-1
Kevin... alias Krelvin
Re: BIPAP vs CPAP
With your pressure numbers, I'm surprised that they didn't start you on a BiPap right from the beginning. Straight CPAP only goes to 20 cmH2O leaving you no headroom to increase pressure. BiPap machines can go to 25 while still allowing the exhale pressure to be lower in line with your 17 from the sleep study.
I am assuming that you are not having any central apnea issues that would require a still different machine.
Get a copy of your sleep study reports and see what they say.
I am assuming that you are not having any central apnea issues that would require a still different machine.
Get a copy of your sleep study reports and see what they say.
_________________
Mask: Oracle HC452 Oral CPAP Mask |
Humidifier: DreamStation Heated Humidifier |
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV |
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
Re: BIPAP vs CPAP
My doctor prescribed me min 10 and max 19. I was not having good luck with this and my numbers were pretty high. I asked for help on the forum and recently I've been doing 17 min and max 20. Last night my AHI was 7.7, this still seems high.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear |
Re: BIPAP vs CPAP
It is. Do you or can you get your own reports to see what the makeup of the 7.7 AHI is?
_________________
Mask: Oracle HC452 Oral CPAP Mask |
Humidifier: DreamStation Heated Humidifier |
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV |
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
Re: BIPAP vs CPAP
It is always Obstructive.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear |
Re: BIPAP vs CPAP
For plain vanilla obstructive sleep apnea, 7.7 events per hour is high. It is in the mild sleep apnea range and would be justification for prescribing a CPAP machine. You already are on a machine and the implication is that your pressures are not providing you with a correct therapeutic level of treatment. The switch to a different machine would likely be to a BiPap as it can provide pressures above 20 cmH2O that is the top pressure your machine can provide. Since you are on an auto machine now, it would make sense to move to an auto BiPap machine instead of a fixed bi-level one. The reimbursement code is the same and the cost to you will be the same but the functionality will be better.
Until that happens, you might try setting your machine to a single pressure of 20 cmH2O by making the range 20/20 instead of 17/20 as it is now set. Then you can see if there is any improvement.
As always, check with you doctor before changing the pressures so the both of you are on the same page.
It will also be helpful if you could use the User Control Panel to post your equipment profile (using text mode). That way, folks who are using exactly the same equipment can chime in.
Until that happens, you might try setting your machine to a single pressure of 20 cmH2O by making the range 20/20 instead of 17/20 as it is now set. Then you can see if there is any improvement.
As always, check with you doctor before changing the pressures so the both of you are on the same page.
It will also be helpful if you could use the User Control Panel to post your equipment profile (using text mode). That way, folks who are using exactly the same equipment can chime in.
_________________
Mask: Oracle HC452 Oral CPAP Mask |
Humidifier: DreamStation Heated Humidifier |
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV |
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
Re: BIPAP vs CPAP
Thanks. I will try this.
Last night was a bad night as my AHI was 10.9. I will try straight 20/20 tonight.
Joel
Last night was a bad night as my AHI was 10.9. I will try straight 20/20 tonight.
Joel
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear |
Re: BIPAP vs CPAP
I don't think his obstructive sleep apnea should be called "plain vanilla".JDS74 wrote:For plain vanilla obstructive sleep apnea
His apnea is caused by a drug he is taking - see his earlier thread.
Re: BIPAP vs CPAP
Susie Kay,
I just reviewed his prior thread about untreatable OSA.
It's clear that his issues are complicated.
That said, I did not say anything in my post about the nature of his OSA. Rather, the statement you quoted was a generalization about the reported AHI level that was described as all obstructive.
It is still true that for anyone with "plain vanilla" OSA, an AHI level of 7.7 untreated qualifies for a cpap machine.
For the OP, having this level with treatment calls for changing his therapy in some manner. His doctor seems to be suggesting a BiPap machine which may allow a higher pressure to deal with whatever is the underlying problem that may actually be the medication Saphis.
I just reviewed his prior thread about untreatable OSA.
It's clear that his issues are complicated.
That said, I did not say anything in my post about the nature of his OSA. Rather, the statement you quoted was a generalization about the reported AHI level that was described as all obstructive.
It is still true that for anyone with "plain vanilla" OSA, an AHI level of 7.7 untreated qualifies for a cpap machine.
For the OP, having this level with treatment calls for changing his therapy in some manner. His doctor seems to be suggesting a BiPap machine which may allow a higher pressure to deal with whatever is the underlying problem that may actually be the medication Saphis.
_________________
Mask: Oracle HC452 Oral CPAP Mask |
Humidifier: DreamStation Heated Humidifier |
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV |
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
Re: BIPAP vs CPAP
Good news! I switched my pressure to 20/20 and my AHI was 4.6 last night. This is the lowest I have ever had in my treatment and I am pretty excited about this! The only thing is that the mask leakage was very high, I will have to tighten my F10 a bit more I think.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear |
- ChicagoGranny
- Posts: 15288
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: BIPAP vs CPAP
Numbers (l/min) please!joeljjk11 wrote:the mask leakage was very high
"It's not the number of breaths we take, it's the number of moments that take our breath away."
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Re: BIPAP vs CPAP
According to the MyAir site it was about 17 or 18.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear |
- ChicagoGranny
- Posts: 15288
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: BIPAP vs CPAP
If that is all it was, it is not high - many think anything below 25 and the machine is still able to prevent events.joeljjk11 wrote:According to the MyAir site it was about 17 or 18.
The problem with myAir data is that is an average. So it could have been, for example, you ran at zero leak much of the night and say 75 (High!) for part of the night.
Have you considered using Sleepyhead software so that you can see the detail of what is happening at night? (See Pugsy's tutorial at top of the index page.)
That mask is designed to be worn loosely and "float" on the face. Be careful about overtightening.joeljjk11 wrote: I will have to tighten my F10 a bit more I think.
Several members have reported improved results/less leaks when using a mask liner with that mask - http://www.padacheek.com/PACAirFitSeries.html
"It's not the number of breaths we take, it's the number of moments that take our breath away."
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.