Sore nose!

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Nardella
Posts: 22
Joined: Sun Jun 14, 2015 1:33 pm

Sore nose!

Post by Nardella » Sun Jun 28, 2015 6:09 am

So after almost a month of on and off attempts I fell asleep! Now I woke up in the middle of the night with a very sore nose. Do I need a larger mask? What can I do until I get a larger mask?
Machine REF 561CA Philips Respironics REMstar Auto A-Flex
Humidifier REF IN6TFLG

User avatar
OSAHell
Posts: 143
Joined: Wed Jun 17, 2015 8:56 am

Re: Sore nose!

Post by OSAHell » Sun Jun 28, 2015 6:51 am

Hi again Nardella,

I just saw your update on the other thread, congrats! As for your sore nose, I had the same problem when I start using pillow after a few days. To be more precise, my problem was irritation on the edge of my nostrils where the pillow come in contact with it. My sleep tech told me about a cream to put on the sore part (which I really don't remember the name). I used that for about a week and problem solved! That should be a lot more easier to fix that than your "airflow timing" issue! Maybe you can go to a pharmacist, there might be OTC cream that'll do the trick, told him it because of CPAP mask. Or you can wait and call your sleep clinic to see if they can recommand you something for this. My sleep tech said that it was a common problem with pillow and she even had a sample tube of the cream.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: SleepyHead on Mac OSX, Resmed S9 VPAP Adapt (36037), EPAP 5 fixed, PS 4-10
Last edited by OSAHell on Sun Jun 28, 2015 8:25 am, edited 2 times in total.

cpierc01
Posts: 66
Joined: Sun Jun 16, 2013 5:54 pm

Re: Sore nose!

Post by cpierc01 » Sun Jun 28, 2015 7:18 am

Go to the Baby isle and get laniosh cream(not sure of the spelling) You will be good in no time

User avatar
JonSnow
Posts: 20
Joined: Tue May 26, 2015 4:34 pm

Re: Sore nose!

Post by JonSnow » Sun Jun 28, 2015 8:09 am

I too was suffering from this just a couple weeks ago when I got the machine. I just used plain old neosporin because it was all I had. It cleared up in a couple days and has not been back yet. I know mine was bad and had it in both nostrils. dont loose hope and it will get better.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: No machine at the moment

User avatar
Sir NoddinOff
Posts: 4190
Joined: Mon May 14, 2012 5:30 pm
Location: California

Re: Sore nose!

Post by Sir NoddinOff » Sun Jun 28, 2015 10:30 am

https://www.lansinoh.com/products/hpa-lanolin

Above is Lansinoh's webspage. Note the spelling.

It's available at Target and at Walgreen in the nursing mother's section. Some call it nipple cream. Of course it's available online. A little bit goes a long way so don't slather it on. There's another cheaper brand out there but it smells too much like a wet wool sweater.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2.
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.

SewTired
Posts: 1737
Joined: Thu Apr 16, 2015 8:33 am
Location: Minneapolis area

Re: Sore nose!

Post by SewTired » Sun Jun 28, 2015 10:42 am

I'm having a similar problem with my new Airfit P10. I just end up switching to the Wisp when I get up to use the bathroom. The Airfit stays in place better, but it just a tad too tight. I figure after a week of stretching, it will be fine, but my poor nose couldn't take it all night.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleepyhead
Diabetes 2, RLS & bradycardia
Airsense For Her; Settings: range 8-12, Airfit P10 (M)

Nardella
Posts: 22
Joined: Sun Jun 14, 2015 1:33 pm

Re: Sore nose!

Post by Nardella » Sun Jun 28, 2015 10:44 am

Just to be clear, I am talking about soreness on the exterior of my nose, not my nostrils.
Machine REF 561CA Philips Respironics REMstar Auto A-Flex
Humidifier REF IN6TFLG

User avatar
corvallisgrace75
Posts: 110
Joined: Sun Jun 07, 2015 7:29 pm

Re: Sore nose!

Post by corvallisgrace75 » Sun Jun 28, 2015 10:48 am

Nardella wrote:Just to be clear, I am talking about soreness on the exterior of my nose, not my nostrils.
Hi Nardella! I was advised by someone here to get Lanolin cream (ointment?) from Lansinoh. I just bought some today. I found it at my local CVS drug store in the baby products section. It's advertised for women who have sore/cracked nipples from breast feeding. I'm hoping it works! Good luck to you!
Mask: AIRFIT P-10 4 HER/AIRFTI MAS FOR HER NASAL PILLOW
ResMed/S10 ELITE

User avatar
corvallisgrace75
Posts: 110
Joined: Sun Jun 07, 2015 7:29 pm

Re: Sore nose!

Post by corvallisgrace75 » Sun Jun 28, 2015 10:52 am

corvallisgrace75 wrote:
Nardella wrote:Just to be clear, I am talking about soreness on the exterior of my nose, not my nostrils.
Hi Nardella! I was advised by someone here to get Lanolin cream (ointment?) from Lansinoh. I just bought some today. I found it at my local CVS drug store in the baby products section. It's advertised for women who have sore/cracked nipples from breast feeding. I'm hoping it works! Good luck to you!
Yeah, duh. (Sir Noddinoff) Looks like someone else already included the link. But I mentioned the same problem. The outside of my nose is tender/sore from use of the nasal pillows.
Mask: AIRFIT P-10 4 HER/AIRFTI MAS FOR HER NASAL PILLOW
ResMed/S10 ELITE

User avatar
OSAHell
Posts: 143
Joined: Wed Jun 17, 2015 8:56 am

Re: Sore nose!

Post by OSAHell » Sun Jun 28, 2015 10:54 am

If it's not like I described in my previous post I can't really tell. But if your mask have any adjustment with velcro, it might be just to tight. Or pillow size maybe, other than that I have no idea, sorry.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: SleepyHead on Mac OSX, Resmed S9 VPAP Adapt (36037), EPAP 5 fixed, PS 4-10

User avatar
bwexler
Posts: 1565
Joined: Mon Nov 08, 2010 11:52 pm
Location: San Marcos, Ca. USA

Re: Sore nose!

Post by bwexler » Sun Jun 28, 2015 2:09 pm

The Wisp offers 4 different cushion sizes. You did not mention which one you used. You also did not mention if you have a tiny button nose or a Jimmy Durante Shnoze.
Most likely you are tightening the straps too tight. You need to allow the cushion to inflate to create the seal. Adjusting both the upper and lower straps is necessary. I also found when using the Wisp it would occasionally slide up toward my eyes causing the lower part to rub on my nose. Adjusting the uper straps seems to be the remedy.

_________________
Mask: SleepWeaver 3D Soft Cloth Nasal CPAP Mask with Headgear
Additional Comments: AurCurve 10 ASV Also using Sleaplyhead 1.1, ResScan 6 and CMS50i

sleep_interrupted7
Posts: 2
Joined: Fri Apr 03, 2015 7:59 pm
Location: Pittsburgh, PA

Re: Sore nose!

Post by sleep_interrupted7 » Sun Jun 28, 2015 3:27 pm

I'm glad I found this post. I'm a newbie here but I've been using my cpap machine with the nasal pillows for a while. although the thread talks about the exterior of the nose, I sometimes have soreness and irritation on the inside of my nose. I use the medium pillows Resmed Airfit P10. i'll sometimes use a q-tip to put Vaseline around the inside of my nostrils but you think I could use that lanolin around the inside of my nostril instead?

Thanks for all the advice already posted here. it's extremely helpful.

Traci

_________________
Mask

User avatar
poppi2
Posts: 590
Joined: Thu Nov 18, 2010 2:54 pm
Location: Houston, near JSC

Re: Sore nose!

Post by poppi2 » Sun Jun 28, 2015 4:07 pm

Yes, use the Lansinoh lanolin. Vaseline and the silicon parts of the mask aren't compatible.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: PAPCap, 3M Microfoam Surgical Tape, PoliGrip Strip, APAP 12.0 - 14.0 cm

User avatar
Sir NoddinOff
Posts: 4190
Joined: Mon May 14, 2012 5:30 pm
Location: California

Re: Sore nose!

Post by Sir NoddinOff » Sun Jun 28, 2015 4:35 pm

sleep_interrupted7 wrote: i'll sometimes use a q-tip to put Vaseline around the inside of my nostrils but you think I could use that lanolin around the inside of my nostril instead? Thanks for all the advice already posted here. it's extremely helpful.

Traci
According to some promotional literature I read regarding Lansinoh it is as close to human 'grease' as possible (it is refined from sheep's wool lanolin)... for all intents and purposes it is safe just about anywhere on the human body. I don't think I would shove it way up my nose but around the inside nares (nostril flares) would be fine I'm sure. I've done it with no problems.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2.
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.

Nardella
Posts: 22
Joined: Sun Jun 14, 2015 1:33 pm

Re: Sore nose!

Post by Nardella » Mon Jun 29, 2015 6:38 am

I have a big nose and am using xl size, I will make the strap less tight tonight, last night I didn't use the humidifier and used neosporin with minimal problems.
Machine REF 561CA Philips Respironics REMstar Auto A-Flex
Humidifier REF IN6TFLG