Can someone explain the functionality ASV settings?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Dave5059
Posts: 15
Joined: Mon May 18, 2015 11:36 am

Can someone explain the functionality ASV settings?

Post by Dave5059 » Fri Jun 12, 2015 9:17 am

I was wondering if any can explain on a 'basic easy to understand' level about if EPAP helps with Centrals or Obstructive? If PSMin helps with Centrals or Obstructive? When would be a good time to change the BPM 'Auto' setting. For instance how would the person need to breathe in order to change the Auto setting for bpm?

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Can someone explain the functionality ASV settings?

Post by chunkyfrog » Fri Jun 12, 2015 10:58 am

Bump, or hop. Keep smiling!

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
grayghost4
Posts: 1554
Joined: Wed Aug 20, 2014 6:52 pm
Location: Norther Illinois
Contact:

Re: Can someone explain the functionality ASV settings?

Post by grayghost4 » Fri Jun 12, 2015 11:32 am

How much to Haunt a house ..... depends on how many rooms.


Image
If you're not part of the solution you're just scumming up the bottom of the beaker!

Get the Clinicians manual here : http://apneaboard.com/adjust-cpap-press ... tup-manual

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: Can someone explain the functionality ASV settings?

Post by JDS74 » Fri Jun 12, 2015 11:50 am

Dave5059 wrote:I was wondering if any can explain on a 'basic easy to understand' level about if EPAP helps with Centrals or Obstructive? If PSMin helps with Centrals or Obstructive? When would be a good time to change the BPM 'Auto' setting. For instance how would the person need to breathe in order to change the Auto setting for bpm?
Let's try to explain what happens in a Respironics 960 ASV machine. Details will be different for a ResMed equivalent.

Min EPAP is the continuous lowest pressure the machine supplies. The pressure will never go below this value. In general this helps prevent Hypopnea events.
Max EPAP is the highest pressure the machine will go to in EPAP mode.
Min Pressure Support is the continuous lowest pressure above the current EPAP the machine supplies in IPAP mode. It helps prevent obstructive events.
Max Pressure Support is the highest IPAP pressure above the current EPAP value that the machine will supply in "normal" IPAP mode.
Max Pressure is the maximum pressure the machine is allowed to provide. Many are set up with this pressure set to 25 CmH2O, the actual machine upper limit.

Compared to a BiPap machine, EPAP will be Min EPAP and IPAP will be Min EPAP + Min Pressure Support.
IPAP will never get below Min EPAP + Min Pressure support and it will never get above Min EPAP + Max Pressure Support when not in ventilator mode.

As needs change, the values for EPAP will vary between Min EPAP and Max EPAP, and IPAP will vary between Min EPAP+Min Pressure Support and Max EPAP and Max Pressure Support. Respironics algorithm allows both the current EPAP and the Current IPAP to vary independently but never letting them get closer than Min Pressure Support.

Now here comes a central event (ventilator mode).
Then EPAP will revert to Min EPAP, the upper limit for IPAP switches to Max Pressure. All else remains the same. The 960 will try to titrate up to Max Pressure in order to keep the patient breathing and assist in restoring normal breathing effort. The longest my machine kept me in this mode was a little over two hours one night but I frequently see runs of 5 to 15 minutes. BTW, I don't wake up during the periods switching between 12 CmH2O and 25 CmH2O, my current settings.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: Can someone explain the functionality ASV settings?

Post by JDS74 » Fri Jun 12, 2015 11:55 am

Dave5059 wrote:... When would be a good time to change the BPM 'Auto' setting. For instance how would the person need to breathe in order to change the Auto setting for bpm?
The time to consider changing from Auto mode to a fixed setting is if you breathing rate is really unstable and you can see by looking at the wave data that ventilator mode is triggering only because of this instability. That is, your breath rate goes from 10 BPM to 18 BPM erratically during the night but you are not having centrals. Generally, though, leaving the machine in Auto is better as it can cope more easily In this mode.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

User avatar
Pugsy
Posts: 65423
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Can someone explain the functionality ASV settings?

Post by Pugsy » Fri Jun 12, 2015 12:18 pm

http://www.isetonline.org/yahoo_site_ad ... 190318.pdf


EPAP is used for obstructive events...either OAs or hyponeas that are obstructive in nature.
Remember hyponeas can also be central in nature.
So usually it is EPAP for OAs and IPAP for hyponeas that are central in nature because that's where PS comes in to help with centrals.

See the above link for in depth explanation.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Sir NoddinOff
Posts: 4189
Joined: Mon May 14, 2012 5:30 pm
Location: California

Re: Can someone explain the functionality ASV settings?

Post by Sir NoddinOff » Fri Jun 12, 2015 12:20 pm

When someone refuses to list their equipment after a certain period of time, I tend to lead with a lot of cautions; especially if ASV is involved. Here’s an intensive video on ASV/ventilator principles and some cautions… If it gets too heavy going for you, then look to the right of the Youtube screen and you’ll see several thumbnails of other ASV/ventilator video tutorials of varying degrees of difficulty. Read what Pugsy linked to first:

https://www.youtube.com/watch?v=6nERo9g5sBU

Note that at about minute 24 in the video he discusses cardiac dysfunction resulting from some ASV/ventilator settings and people with heart problems - ie. closely related to the warnings popping up in May 2015 concerning ResMed's Emergency Alerts to their international medical clients, specifically regarding LVEF equaling or below 45% (Google it). This is the only really clear presentation of the material I've seen online. Earlier he touches on ASV relating to COPD issues. It's good concise stuff.

I'm not a doctor, so consult your medical team for any concerns you have.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2.
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.

StephenR
Posts: 172
Joined: Fri Jul 25, 2014 4:43 pm

Re: Can someone explain the functionality ASV settings?

Post by StephenR » Fri Jun 12, 2015 3:19 pm

Sir NoddinOff wrote:When someone refuses to list their equipment after a certain period of time, I tend to lead with a lot of cautions; especially if ASV is involved. Here’s an intensive video on ASV/ventilator principles and some cautions… If it gets too heavy going for you, then look to the right of the Youtube screen and you’ll see several thumbnails of other ASV/ventilator video tutorials of varying degrees of difficulty. Read what Pugsy linked to first:

https://www.youtube.com/watch?v=6nERo9g5sBU

Note that at about minute 24 in the video he discusses cardiac dysfunction resulting from some ASV/ventilator settings and people with heart problems - ie. closely related to the warnings popping up in May 2015 concerning ResMed's Emergency Alerts to their international medical clients, specifically regarding LVEF equaling or below 45% (Google it). This is the only really clear presentation of the material I've seen online. Earlier he touches on ASV relating to COPD issues. It's good concise stuff.

I'm not a doctor, so consult your medical team for any concerns you have.
What a great video..

Thanks for sharing that....

Stephen

User avatar
bwexler
Posts: 1591
Joined: Mon Nov 08, 2010 11:52 pm
Location: San Marcos, Ca. USA

Re: Can someone explain the functionality ASV settings?

Post by bwexler » Fri Jun 12, 2015 10:13 pm

That was a great video.
Why didn't you tell me to watch it sooner?

He didn't exactly talk about PAP machines. But he certainly did explain ASV very clearly as well as what the settings mean.

Thank yo.

_________________
Mask: SleepWeaver 3D Soft Cloth Nasal CPAP Mask with Headgear
Additional Comments: AurCurve 10 ASV Also using Sleaplyhead 1.1, ResScan 6 and CMS50i

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Can someone explain the functionality ASV settings?

Post by palerider » Fri Jun 12, 2015 10:21 pm

bwexler wrote:That was a great video.
Why didn't you tell me to watch it sooner?

He didn't exactly talk about PAP machines. But he certainly did explain ASV very clearly as well as what the settings mean.
actually, there was nothing in there about asv... that was pretty much all tightly controlled therapeutic ventilators.

the fact that there were a couple settings that you can find on an asv doesn't mean he was talking about ASVs, he wasn't.

that was more applicable to a bipap S/T or vpap ST-A.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.