ResMed Mask Setting

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

ResMed Mask Setting

Post by Jay Aitchsee » Mon May 11, 2015 3:21 pm

From thread viewtopic/t105618/Does-the-Airfit-F10-v ... l#p1000537
palerider wrote:
RogerSC wrote:There is a setting for mask type in the S9 menus. That is supposed to help correct for a mask's normal venting. The setting that I'm using is "nasal pillows" and works fine for me, but since nasal pillows don't vent as much as a full face mask, not setting this one appropriately could be your problem. If you haven't set this already *smile*.
while others don't agree with me, I believe that this commonly held idea is wrong.

if you look at a chart of resmed mask vent rates, and add in others from other manufacturers. you'll find that they all vent at approximately the same rate. around 20lpm @4cm to around 50lpm@20cm, there are minor differences, but they are minor.

however, if you look in the manuals for the masks and look at their RESISTANCE specs, they all pretty much line up with full face having the lowest resistance, nasal having slightly higher, and pillow masks having the greatest resistance.

this data leads to the conclusion that the mask type setting in the resmeds is so that the machine can adjust it's pressures to compensate for the resistance that the different masks provide, in much the same way that the hose type setting allows it to compensate for the different types of hoses.

Code: Select all

pressure   p10     f10
4          20      22
8          29      32
12         37      41
16         43      48
20         49      54
less, yes, but not enough to trigger a leak alert.

contrast that with the resistance, with the p10, @50lpm, .4cm @100lpm, 1.4cm f10: @50, .2cm @100 1.0cm.

knowing the pressure you've set, and the type of hose and mask, the machine can raise the pressure where it's measured (by the air outlet) to compensate for the pressure drops at the mask, and thus give you the desired pressure.
Jay Aitchsee wrote:
palerider wrote:this data leads to the conclusion that the mask type setting in the resmeds is so that the machine can adjust it's pressures to compensate for the resistance that the different masks provide, in much the same way that the hose type setting allows it to compensate for the different types of hoses.
...contrast that with the resistance, with the p10, @50lpm, .4cm @100lpm, 1.4cm f10: @50, .2cm @100 1.0cm.
Interesting theory, but those pressure drops stats are given at fairly high flow rates (50 and 100lpm) which would generally correspond to static vent rates at pressures of 20cmH2O and higher. But even at 50lpm the difference in resistance for the two masks is only .2cm. Since, I think, resistance is a function of the square of flow, the absolute difference in the two at more typical values of use would be even less.

I'm just sayin'

Edit: In thinking about this a little more, it may not make any sense to talk about static vent flow rates in conjunction with Resistance. Maybe Minute Ventilation would be more appropriate? Mine is about 30Lpm max and 8 average with FFM and 15lpm max and 7 average w/ pillows. Don't know, it is interesting.
palerider wrote:
Jay Aitchsee wrote:Edit: In thinking about this a little more, it may not make any sense to talk about static vent flow rates in conjunction with Resistance. Maybe Minute Ventilation would be more appropriate? Mine is about 30Lpm max and 8 average with FFM and 15lpm max and 7 average w/ pillows. Don't know, it is interesting.
it sounds to me like you're onto something there...

to get the applicable flow rate, you'd have to look at the expected flow @presssure, say, 35lpm@10ish cm. then add in the varying flow from breathing, looking at one night of my data, I see breathing flows up to 120lpm, though usually more like 30-60.

that gives a wide range of flow from 65-155lpm on the plus side.

without the mask type compensation, your actual delivered pressure would vary more, I think that respironics resistance control is their way of copying the idea in the 60 series, but they got more specific. it makes perfect sense to me,

Image

without the added compensation for the flow resistance of the masks, the final end pressure isn't nearly as accurate.
What if it does both; i.e., set the flow rate and the resistance value? From using different masks, with different settings, I'm pretty sure the setting determines the base leak/vent threshold. For example, my Simplus, which has a slightly lower static vent rate, always shows a lower base leak line than my F10. I'm pretty sure, though not positive, that the base line shifts slightly if a different setting is used with the same mask.

Your point was the static vent rates were practically the same for all types of ResMed masks, so the vent rate wouldn't need differentiation. However, I differ with that view. From the latest ResMed Flow Rate chart: FFM @8 = 32, @20 = 54, while pillows @ 8 = 29, @20 = 49. For differences of 3 and 5 cmH2O. Those differences are great enough to shift the computed leak base line if the wrong value is used.

Now, if the setting is used to set the resistance, then the resistance for each class of mask should be the same, or quite close. I haven't looked at that. It might be interesting to do so. I agree that the Resistance of the mask will affect the pressure delivered and at 100lpm a 1.4cmH2O pressure drop could exceed ResMed's claim of a pressure tolerance within a nominal +/- 0.5 cmH2o +/- 4% of the measured pressure. Though I think, under normal circumstances, 100lpm is quite high.

So, maybe both?

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: ResMed Mask Setting

Post by palerider » Mon May 11, 2015 3:45 pm

Jay Aitchsee wrote:Your point was the static vent rates were practically the same for all types of ResMed masks, so the vent rate wouldn't need differentiation. However, I differ with that view. From the latest ResMed Flow Rate chart: FFM @8 = 32, @20 = 54, while pillows @ 8 = 29, @20 = 49. For differences of 3 and 5 cmH2O. Those differences are great enough to shift the computed leak base line if the wrong value is used.
that's a good point, however, if you look at the published leak charts for nasal masks... it starts to look like it's not scientifically defensible, since nasal mask rates vary from less than pillows to more than FFM.
http://www.resmed.com/content/dam/resme ... er_eng.pdf
Jay Aitchsee wrote:Now, if the setting is used to set the resistance, then the resistance for each class of mask should be the same, or quite close. I haven't looked at that. It might be interesting to do so. I agree that the Resistance of the mask will affect the pressure delivered and at 100lpm a 1.4cmH2O pressure drop could exceed ResMed's claim of a pressure tolerance within a nominal +/- 0.5 cmH2o +/- 4% of the measured pressure. Though I think, under normal circumstances, 100lpm is quite high.

So, maybe both?
I have looked at that, as mentioned above: "however, if you look in the manuals for the masks and look at their RESISTANCE specs, they all pretty much line up with full face having the lowest resistance, nasal having slightly higher, and pillow masks having the greatest resistance."

if you go through the manuals and write down the published resistance for each type of mask, and then sort them, they line up pretty neatly into pillow, nasal, ffm, order. there are a couple of exceptions, like the liberty.

unfortunately, I don't still have temporary spreadsheet that I did that with.

also, I'll point out that it really makes no sense to have a preset table of leak settings. since that table won't take into account any new mask that might need to be different. AND, that you can, pretty easily, calculate a baseline leak by watching the flow rate. you can see it happen on a vpap if you watch the therapy screen when you first turn the machine on... and, it's what mark does in sleepyhead to get the lower (unintended) leak trace for respironics machines.

I don't know why everybody thinks that the mask type setting is for leak calculation, but it just doesn't make sense that it would be so.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: ResMed Mask Setting

Post by Jay Aitchsee » Mon May 11, 2015 5:31 pm

palerider wrote:I don't know why everybody thinks that the mask type setting is for leak calculation, but it just doesn't make sense that it would be so.
Well, because ResMed essentially says so in a couple of places. I know, I know, we've had this conversation before. You don't accept that's what they mean.
I'll give you that the VPAP line may treat leaks differently.
But, I don't buy it doesn't make sense. Makes sense to me. It would appear that ResMed designs most of it's masks in the same category to have the same characteristics (Vent Rate, Resistance, etc.) A new FFM? Just give it the same characteristics and use the FFM setting. Makes a lot of sense.
By the way, have you looked at some of the outliers you mentioned that have different characteristics from others in the same category? If memory serves, I think some old masks say to use a different setting than the category they fall in if they're being used with a newer machine. I don't know, I'll have to look it up.
Anyway, it's interesting.

Edit: I looked at some of the nasal masks that don't seem to fall in line and it seems to be the Mirage line. I looked at a couple of those and they say to select "standard" if available. That setting no longer exists, so they are old and apparently designed for older machines.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: ResMed Mask Setting

Post by palerider » Mon May 11, 2015 6:24 pm

Jay Aitchsee wrote:
palerider wrote:I don't know why everybody thinks that the mask type setting is for leak calculation, but it just doesn't make sense that it would be so.
Well, because ResMed essentially says so in a couple of places. I know, I know, we've had this conversation before. You don't accept that's what they mean.
I've never seen that from resmed, (I'm sure there's a lot of stuff I haven't seen, I'd be interested in seeing any references.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: ResMed Mask Setting

Post by Jay Aitchsee » Tue May 12, 2015 4:31 am

palerider wrote:I've never seen that from resmed, (I'm sure there's a lot of stuff I haven't seen, I'd be interested in seeing any references.
Here's a couple:
ResScan Report Interpretation Guide, page 32 wrote: DEFINITIONS
Leak: This refers to unintentional leak,
which is the leak value after deduction
of the intentional mask leak.
ResScan Report Interpretation Guide, page 56 wrote: ResMed devices calculate unintentional leak:
Unintentional leak = Total leak – Intentional leak from mask CO2 washout
You have to set the mask type. emphasis added
http://www.resmed.eu/us/assets/html/niv/leak-management.html?nc=dealers wrote: Interface selection
ResMed offers an exceptional selection of vented masks designed to ensure adequate washout of carbon dioxide while minimizing leak and maintaining patient comfort. All of these benefits are extremely important for patients utilizing NIV therapy.

Many ResMed devices offer an interface selection menu. Selecting the suitable interface enables the device to factor in the intentional leak emitting from the mask vent, which allows for more accurate unintentional leak reporting and alarm accuracy
. emphasis added
Additionally, here's a paper you may find interesting:

http://scholar.google.com/scholar_url?u ... i=scholarr

Monitoring of Noninvasive Ventilation by
Built-in Software of Home Bilevel Ventilators
A Bench Study
In this study,we first realized that data provided by ventilator software were not measured in the same way in all devices; this is not always explicitly stated by the manufacturers. One device estimated leaks only under expiratory positive airway pressure (device E) (these results are thus misleading and cannot be compared with those of other devices); furthermore, some devices subtract from their estimation of leaks the intentional leaks expected for a given type of mask at a given pressure setting (devices F and G), while others report the sum of intentional and unintentional leaks. Once this was taken into account, the reliability of leak assessment was shown to be highly variable from one device to another. emphasis added
Devices F & G are ResMed which were found to be among the most accurate in leak reporting.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: ResMed Mask Setting

Post by palerider » Tue May 12, 2015 10:43 am

Jay Aitchsee wrote:
palerider wrote:I've never seen that from resmed, (I'm sure there's a lot of stuff I haven't seen, I'd be interested in seeing any references.
Here's a couple:
thank you, I've never seen those before.

time to revise some theories

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: ResMed Mask Setting

Post by Jay Aitchsee » Tue May 12, 2015 12:42 pm

palerider wrote:time to revise some theories
Really? Wow!

Well, they still could also use the mask setting to fine tune for Resistance as well.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video

User avatar
Pugsy
Posts: 65139
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ResMed Mask Setting

Post by Pugsy » Tue May 12, 2015 1:07 pm

palerider wrote: time to revise some theories
Do you see me "snickering"? I thought you had seen all that documentation or I would have shown it to you a long time ago when you would challenge me about the mask setting being use for leak determination and I gave up swinging the 2 X 4 because I couldn't reach you and it wasn't worth the fight. I always thought you had seen it and still felt the way you did so I didn't bother trying to change your mind.

Though like Jay says...maybe the resistance thing is also a factor that just isn't documented where we have access to the documentation.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: ResMed Mask Setting

Post by palerider » Tue May 12, 2015 2:05 pm

Jay Aitchsee wrote:
palerider wrote:time to revise some theories
Really? Wow!

Well, they still could also use the mask setting to fine tune for Resistance as well.
hey, I always said it was my theory, based on the data I've been able to figure out.

it *still* doesn't make *sense* to me that it would be used in the leak calcs, just because of the differences in the nasal masks, AND because you can calculate a baseline by watching the variation of the flow with breathing and math (that's beyond me).

but, I'm not so stupid (I may be stupid, but i'm not THAT stupid) as to look at published docs and say "that doesn't fit my idea, so it's nonsense"...

i haven't had time this morning to dig up the first one you quoted, or read the one you linked to, but I will, because I'm curious.

I do know, first hand, that the setting is used for more than just the leak, though, because when I first got my p10 mask, I popped it on the hose and stuck it on my face, the machine started up.... I breathed for it a bit to see how it felt, and when I took it off, the smartstart did NOT turn off the machine... I let it run for a minute or two, thinking "well, poo, I really like that feature" and turned it off.... then I remembered the mask type setting, changed it, and just tried the mask again... and it was back to working just like it was supposed to, start on first breath, stop a few seconds after removing the mask.

so, like you said, maybe for both, they just haven't deigned to document it anywhere. I wonder if it's in any of the patents... *ponder*

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: ResMed Mask Setting

Post by palerider » Tue May 12, 2015 2:07 pm

Pugsy wrote:
palerider wrote: time to revise some theories
Do you see me "snickering"? I thought you had seen all that documentation or I would have shown it to you a long time ago when you would challenge me about the mask setting being use for leak determination and I gave up swinging the 2 X 4 because I couldn't reach you and it wasn't worth the fight. I always thought you had seen it and still felt the way you did so I didn't bother trying to change your mind.

Though like Jay says...maybe the resistance thing is also a factor that just isn't documented where we have access to the documentation.
you're funny... and, yes, I can almost HEAR you snickering.

I'm sure there's lots of docs I haven't read, it's not like they make it all that easy to find some of the stuff *lol*.

I don't remember seeing that part though... so, I do thank Jay for bringing it to me... it's more data

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: ResMed Mask Setting

Post by Jay Aitchsee » Tue May 12, 2015 2:55 pm

palerider wrote:it *still* doesn't make *sense* to me that it would be used in the leak calcs, just because of the differences in the nasal masks,
Yeah, the Mirage line kind of makes you wonder, but like I said earlier those masks are old some of them have been around since at least the S6 (http://web.medtecshop.com/Files/Equipme ... r1_620.pdf) and they did have different settings then. I'm not sure when leak reporting started - maybe with ResScan and the S7? Maybe ResMed figured nobody's using those masks anymore and, if they are, they don't care about leaks anyway.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: ResMed Mask Setting

Post by Jay Aitchsee » Wed May 13, 2015 11:50 am

PR, I thought you'd enjoy this thread from 2009: viewtopic/t44802/ResMed-Machines-quotSt ... Rates.html. It seems those nasal mask flow rates were problematic even then.

And here's the earliest I turned up - from 2004: viewtopic.php?f=1&t=29&p=82&hilit=+mask+setting#p82

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: ResMed Mask Setting

Post by palerider » Wed May 13, 2015 1:14 pm

Jay Aitchsee wrote:PR, I thought you'd enjoy this thread from 2009:
cool, thanks!

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: ResMed Mask Setting

Post by Jay Aitchsee » Wed May 13, 2015 3:09 pm

And here's another paper that is interesting. It's a study that compares EPR with C-flex and devotes some space to the differences in leak reporting between the two brands as well as to the differences in pressure reduction methodology.
http://www.sleepdt.com/comparative-effi ... n-m-mutlu/

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video

User avatar
RogerSC
Posts: 1912
Joined: Sat Jul 21, 2012 1:11 pm

Re: ResMed Mask Setting

Post by RogerSC » Wed May 13, 2015 3:22 pm

palerider wrote:
Pugsy wrote:
palerider wrote: time to revise some theories :D
Do you see me "snickering"? :lol: :lol: I thought you had seen all that documentation or I would have shown it to you a long time ago when you would challenge me about the mask setting being use for leak determination and I gave up swinging the 2 X 4 because I couldn't reach you and it wasn't worth the fight. I always thought you had seen it and still felt the way you did so I didn't bother trying to change your mind.

Though like Jay says...maybe the resistance thing is also a factor that just isn't documented where we have access to the documentation.
you're funny... and, yes, I can almost HEAR you snickering.

I'm sure there's lots of docs I haven't read, it's not like they make it all that easy to find some of the stuff *lol*.

I don't remember seeing that part though... so, I do thank Jay for bringing it to me... it's more data :)
Since I'm the person quoted at the top, I should say that when Mr. Rider reacted to my posting, I also got a little ruffled based on what I've read. But decided not to argue about it, since my peace of mind matters more than being right, in general. There are times when I won't let things go by, but this wasn't one. Glad to see it taken care of so nicely by others, though. Thanks!