Hoses

Post by NewToBiPap » Thu Mar 12, 2015 3:05 pm

What's the difference in using a regular hose and a heated hose?

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: I am using the VPAP ST . . . that wasnt a choice in machines.

User avatar
LSAT
Posts: 13403
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Hoses

Post by LSAT » Thu Mar 12, 2015 3:41 pm

Duh....The heated hose heats the moist air. A regular hose doesn't.

User avatar
SGearhart
Posts: 434
Joined: Fri Nov 07, 2014 8:54 am
Location: Georgia

Re: Hoses

Post by SGearhart » Thu Mar 12, 2015 3:43 pm

A heated hose reduces the chances for rainout. It also allows for transmission of higher humidity levels than what can be obtained from a regular hose. A heated hose also allows for heating the air you breathe so you won't get the cold air feeling.

I hope this helps.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS-50F & 50IW - SleepyHead 1.0.0-Beta

User avatar
TangledHose
Posts: 862
Joined: Wed Mar 13, 2013 11:13 am
Location: Centennial, CO

Re: Hoses

Post by TangledHose » Thu Mar 12, 2015 3:44 pm

Heated hose can reduce or eliminate moisture condensation within the hose (rainout).....also many people including myself prefer to have warmer feeling air coming in the mask......I hate a cold nose! But primary purpose is to eliminate rainout. When enough moisture condenses in the hose it collects into a puddle and starts gurgling inside the hose......the gurgling is annoying and will usually wake you up, at which point you have to disconnect the hose and dump the water out.....big PIA

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Simplus Full Face CPAP Mask with Headgear
Additional Comments: Also own F&P Icon Auto for backup and travel; and a Transcend when packing space is limited
Link to Sleepyhead:
http://sleepfiles.com/SH2/

User avatar
TangledHose
Posts: 862
Joined: Wed Mar 13, 2013 11:13 am
Location: Centennial, CO

Re: Hoses

Post by TangledHose » Thu Mar 12, 2015 3:54 pm

LSAT wrote:Duh....The heated hose heats the moist air. A regular hose doesn't.

I must say I do like LSAT's answer the best, nice and concise

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Simplus Full Face CPAP Mask with Headgear
Additional Comments: Also own F&P Icon Auto for backup and travel; and a Transcend when packing space is limited
Link to Sleepyhead:
http://sleepfiles.com/SH2/

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Hoses

Post by palerider » Thu Mar 12, 2015 4:27 pm

TangledHose wrote:
LSAT wrote:Duh....The heated hose heats the moist air. A regular hose doesn't.

I must say I do like LSAT's answer the best, nice and concise
and, wrong.

a heated hose could heat the air a little if you crank it way up, but it's purpose is to keep the inner surface of the hose above the dewpoint of the humidified air, so that it doesn't condense on the tube walls, in the same way that moisture condenses on the outside of a cold glass.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

NewToBiPap
Posts: 38
Joined: Wed Jan 07, 2015 10:18 pm

Re: Hoses

Post by NewToBiPap » Thu Mar 12, 2015 5:12 pm

Thank you for the explanation everyone. I was curious because I noticed both options and didn't know if it was necessary to buy the heated hose.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: I am using the VPAP ST . . . that wasnt a choice in machines.

Oughtsix
Posts: 92
Joined: Tue Jun 21, 2011 3:34 pm

Re: Hoses

Post by Oughtsix » Thu Mar 12, 2015 5:32 pm

I would recommend going with the heated hose if you can swing it. It solves the problem of "rain out" (moisture condensing in the hose) which gets very bothersome with non-heated hoses.

_________________
MachineMask

User avatar
Pugsy
Posts: 65327
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Hoses

Post by Pugsy » Thu Mar 12, 2015 5:34 pm

Using the heated hose can also help when a person wants to maintain the same level of humidity delivery but they wish to change the air temp going to the mask.
Like in the winter my bedroom is cold so I like to warm the air up that is going to my nose....but in the summer I turn the temp down because I don't want the extra warmth. Using the heated hose lets me do that without ever changing my humidity setting.

_________________