PR ASV auto "OFF" feature?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
photonic
Posts: 261
Joined: Fri Jun 27, 2014 7:42 pm

PR ASV auto "OFF" feature?

Post by photonic » Wed Feb 25, 2015 9:42 pm

Does the PR System One REMStar 60 Series BiPAP autoSV Advanced have an auto "OFF" feature. I cannot find one anywhere but thought I would make a last check on this here in case I have overlooked something.

I use the auto "ON" feature since I have fallen asleep with mask on a few times before I have pushed the button.

Thanks!
ResMed AirCurve 10 ASV
8 years with F&P Oracle mouth mask now changing to
F&P Evora Full with S-M size
Hozer hose management device (a miracle!)
OSCAR software

User avatar
Krelvin
Posts: 1977
Joined: Tue Jun 06, 2006 5:23 pm
Location: Metro Phx Area - Dry Heat!

Re: PR ASV auto "OFF" feature?

Post by Krelvin » Wed Feb 25, 2015 10:02 pm

Can't say for the ASV as I don't have a manual but the one I do have a manual for the BiPap Auto with Bi-Flex shows in the clinicians settings both an auto on and auto off setting. I would guess the ASV has that as well if it has the Auto On. Hopefully someone has the manual to tell for sure.
Current Settings PS 4.0 over 10.6-18.0 (cmH2O) - Resmed S9 VPAP Auto w/h5i Humidifier - Quattro Air FFM
TNET Sleep Resource Pages - CPAP Machine Database
Put your equip in your Signature - SleepyHead v1.0.0-beta-1
Kevin... alias Krelvin

User avatar
Pugsy
Posts: 65339
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: PR ASV auto "OFF" feature?

Post by Pugsy » Wed Feb 25, 2015 10:14 pm

There is no Auto Off available in the 960 model as far as I can tell.
The manual I have shows Auto On but does not show Auto Off and I think that this was discussed here a few months ago.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
adipasqu
Posts: 128
Joined: Thu Apr 21, 2011 11:24 am
Location: San Francisco Bay Area, CA

Re: PR ASV auto "OFF" feature?

Post by adipasqu » Wed Feb 25, 2015 10:15 pm

No auto off on the 950 or 960 ASV's. Why not? Well, one of the main functions of the ASV is to trigger breaths when you aren't spontaneously breathing. Auto Off would most likely interfere with this important function or cause more machine complexity than it is worth. If you aren't breathing, the last thing you want is your ASV to shut off.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Now using a ResMed S9 VPAP Adapt (36037) with H5i...SleepyHead software
I am a doctor...just not that kind of doctor.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: PR ASV auto "OFF" feature?

Post by palerider » Wed Feb 25, 2015 10:30 pm

adipasqu wrote:No auto off on the 950 or 960 ASV's. Why not? Well, one of the main functions of the ASV is to trigger breaths when you aren't spontaneously breathing. Auto Off would most likely interfere with this important function or cause more machine complexity than it is worth. If you aren't breathing, the last thing you want is your ASV to shut off.
it's more likely that they omitted auto off in favor of the optional alarms, the resmed disables auto off if the leak alert feature is turned on, however, turn that alarm off, and auto on/off is available. (another little nitpicky preference for resmed, in my book)

the machine can *easily* tell the difference between not breathing and maskoff, simply due to the pressure in the line.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
photonic
Posts: 261
Joined: Fri Jun 27, 2014 7:42 pm

Re: PR ASV auto "OFF" feature?

Post by photonic » Wed Feb 25, 2015 11:21 pm

Thanks everyone! Makes sense. I will stop thinking about it!!!
ResMed AirCurve 10 ASV
8 years with F&P Oracle mouth mask now changing to
F&P Evora Full with S-M size
Hozer hose management device (a miracle!)
OSCAR software