Still stop breathing

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
ronMonster
Posts: 13
Joined: Wed Feb 04, 2015 12:44 am

Still stop breathing

Post by ronMonster » Mon Feb 09, 2015 11:14 pm

Tonight is night three with CPAP. I fell asleep quickly and was woken up when I stopped breathing, time after time. This is what happened before started CPAP. Does it take while to stop happening?

I set my settings to 11 like doctor said, should turn it up? I don't see the doctors for another month.

thanks
Ron

REMStarPlus M Series Model #1022334
With a useless humidifier that neither warms nor humidifies
Last edited by ronMonster on Tue Feb 10, 2015 12:24 am, edited 1 time in total.

User avatar
grayghost4
Posts: 1554
Joined: Wed Aug 20, 2014 6:52 pm
Location: Norther Illinois
Contact:

Re: Still stop breathing

Post by grayghost4 » Tue Feb 10, 2015 12:08 am

need more info ..... what cpap.....
If you're not part of the solution you're just scumming up the bottom of the beaker!

Get the Clinicians manual here : http://apneaboard.com/adjust-cpap-press ... tup-manual

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Still stop breathing

Post by Wulfman... » Tue Feb 10, 2015 12:22 am

ronMonster wrote:Tonight is night three with CPAP. I fell asleep quickly and was woken up when I stopped breathing, time after time. This is what happened before started CPAP. Does it take while to stop happening?

I set my settings to 11 like doctor said, should turn it up? I don't see the doctors for another month.

thanks
Ron
Without seeing a report, it's virtually impossible to figure out what's going on with your sleep and events.
Which machine do you have and is it fully data-capable?


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

ronMonster
Posts: 13
Joined: Wed Feb 04, 2015 12:44 am

Re: Still stop breathing

Post by ronMonster » Tue Feb 10, 2015 12:36 am

My model is
REMStarPlus M Series Model #1022334
With a useless humidifier that neither warms nor humidifies.

No data.

I thought that waking up in a panic as soon as you fall asleep was part of apnea and would go away with a clap.
But it sounds like that might another issue.
If I'm not making sense it's because I've not slept two days and kinda at the end of my rope.

sorry for the hassles

User avatar
grayghost4
Posts: 1554
Joined: Wed Aug 20, 2014 6:52 pm
Location: Norther Illinois
Contact:

Re: Still stop breathing

Post by grayghost4 » Tue Feb 10, 2015 12:41 am

Welcome to the cpaptalk board

you are making sense ... it probably is apnea .... but without data there is no way anyone can give you any suggestions to improve your therapy .
If you're not part of the solution you're just scumming up the bottom of the beaker!

Get the Clinicians manual here : http://apneaboard.com/adjust-cpap-press ... tup-manual

User avatar
grayghost4
Posts: 1554
Joined: Wed Aug 20, 2014 6:52 pm
Location: Norther Illinois
Contact:

Re: Still stop breathing

Post by grayghost4 » Tue Feb 10, 2015 12:56 am

sounds like you are having a very bad time getting started with the therapy
first machine broken ... second machine a brick (meaning no data).

what was the diagnosis of your sleep test ... do you have a copy of it ?

it is up to you to decide ... but raising the pressure is probably what you need.
If you're not part of the solution you're just scumming up the bottom of the beaker!

Get the Clinicians manual here : http://apneaboard.com/adjust-cpap-press ... tup-manual

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Still stop breathing

Post by Wulfman... » Tue Feb 10, 2015 12:57 am

ronMonster wrote:My model is
REMStarPlus M Series Model #1022334
With a useless humidifier that neither warms nor humidifies.

No data.

I thought that waking up in a panic as soon as you fall asleep was part of apnea and would go away with a clap.
But it sounds like that might another issue.
If I'm not making sense it's because I've not slept two days and kinda at the end of my rope.

sorry for the hassles
Sorry to hear you're "flying blind".
The problem with increasing pressure is that you don't know what type of events they are. If they're Central Apneas, increasing pressure MAY make them worse. But Centrals aren't that much more serious than Obstructives, so you might try increasing pressure and see what happens.
Actually, some Centrals are experienced by some users as they fall asleep and just before waking.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

ronMonster
Posts: 13
Joined: Wed Feb 04, 2015 12:44 am

Re: Still stop breathing

Post by ronMonster » Tue Feb 10, 2015 12:58 am

I know my insurance won't cover any more sleep tests. So no data there.
They just said "severe sleep apnea set your machine to 11 and you have a $1,500 deductible" Bought one on criagslsit.

REMstar Plus M series

I knew I would have to face my claustrophobia, I did not expect such a severe reaction with my sinuses, it feels like I'm in kicked in the throat, and a headache that won't go away.
All of these are going to get a lot worse once I lie down.

User avatar
grayghost4
Posts: 1554
Joined: Wed Aug 20, 2014 6:52 pm
Location: Norther Illinois
Contact:

Re: Still stop breathing

Post by grayghost4 » Tue Feb 10, 2015 1:06 am

I hate to be the one to tell you this .... you are probably gonig to need a different cpap ... that is data capable.
That is the only way you are going to be able to self titrate your pressure.
If you're not part of the solution you're just scumming up the bottom of the beaker!

Get the Clinicians manual here : http://apneaboard.com/adjust-cpap-press ... tup-manual

ronMonster
Posts: 13
Joined: Wed Feb 04, 2015 12:44 am

Re: Still stop breathing

Post by ronMonster » Tue Feb 10, 2015 1:07 am

I think the Doctor said it was obstruction, I've had 4 surgery were they basically Roto-Rootered out my throat. It worked for about 10 minutes.
The best sleep is just a humidifier. Sooner of later I'll need to get some sleep I might take a break from CPAP to get some sleep. I commute 4 hours a day and without sleep it's a scary ride.

User avatar
grayghost4
Posts: 1554
Joined: Wed Aug 20, 2014 6:52 pm
Location: Norther Illinois
Contact:

Re: Still stop breathing

Post by grayghost4 » Tue Feb 10, 2015 1:11 am

you are driving 4 hrs a day without sleep .... tell me where you are driving ... so I can make sure I am not near to your route
If you're not part of the solution you're just scumming up the bottom of the beaker!

Get the Clinicians manual here : http://apneaboard.com/adjust-cpap-press ... tup-manual

User avatar
kteague
Posts: 7788
Joined: Tue May 16, 2006 8:30 pm
Location: West and Midwest

Re: Still stop breathing

Post by kteague » Tue Feb 10, 2015 1:48 am

If your machine has a ramp feature, I suggest you turn it off. That feature is commonly used, and commonly the problem when people describe having events as soon as they doze off. Are you using the ramp?

_________________
Mask: TAP PAP Nasal Pillow CPAP Mask with Improved Stability Mouthpiece
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Bleep/DreamPort for full nights, Tap Pap for shorter sessions

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Still stop breathing

Post by Julie » Tue Feb 10, 2015 6:12 am

It is also just possible that you are one of those people who don't do well on autopap, but just need a fixed Cpap pressure as the autos actually wake you up as they try to deal with individual apneas. Can you turn your machine to straight Cpap mode (at what you feel might be the pressure you most likely stay at most nights) for 1-2 nights and see if you feel any better?

User avatar
Pugsy
Posts: 65388
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Still stop breathing

Post by Pugsy » Tue Feb 10, 2015 8:24 am

Julie the Remstar Plus M series is a fixed cpap machine...it doesn't have auto adjusting apap mode available.
ronMonster wrote: I did not expect such a severe reaction with my sinuses,
Very possibly simply from the non functioning humidifier if it isn't working.
While there are some people who can get by just fine without any or minimal added humidity, there are a lot of people who simply can't do that because their nasal mucosa and sinuses need more moisture.
You can get a working humidifier for the M series machine from secondwindcpap.com http://www.secondwindcpap.com/Used_CPAP.html for $70 but if at all possible it would be better to put that money towards a full data model machine with a working humidifier.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

ronMonster
Posts: 13
Joined: Wed Feb 04, 2015 12:44 am

Re: Still stop breathing

Post by ronMonster » Tue Feb 10, 2015 11:09 pm

Julie wrote:It is also just possible that you are one of those people who don't do well on autopap, but just need a fixed Cpap pressure as the autos actually wake you up as they try to deal with individual apneas. Can you turn your machine to straight Cpap mode (at what you feel might be the pressure you most likely stay at most nights) for 1-2 nights and see if you feel any better?
Thank you I will try.