Can higher pressures mask leaks?

Post by VictorC » Sat Jan 10, 2015 7:53 am

Last night was my second night with my Airfit F10. I managed to sleep pretty well, though I did have my first large leak ever. I also had to remove the mask at one point to that I could reposition it. None of that bothers me too much. I am curious though, because I see a period from about 3:15 to 4:00 where I had a cluster of events. My machine responded by increasing the pressure to the maximum level I allow. That's normal enough, but the curious bit is what happened to my leak rate. It went down to 0 until the pressure decreased. Can higher pressures sometimes mask leaks, or is this a normal phenomenon? Here is a screenshot from Sleepyhead.

Image

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Settings: APAP at 10.5-14cm H2O, AFLEX @ 2, Humidifier @ 2, Hose Temperature @ 3, No Ramp

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Can higher pressures mask leaks?

Post by Julie » Sat Jan 10, 2015 11:11 am

Victor - go to your other thread of ? yesterday - you have replies there.

User avatar
OkyDoky
Posts: 2869
Joined: Mon Aug 25, 2014 5:18 pm

Re: Can higher pressures mask leaks?

Post by OkyDoky » Sat Jan 10, 2015 11:27 am

Julie wrote:Victor - go to your other thread of ? yesterday - you have replies there.
I think what he is looking for is about the graph not showing leaks with the higher pressures brought on by the cluster of events, not about controlling leaks. Since he has been having leaks earlier but not at that time, seems he doubts the graph.
I understand your question, I think, but I don't know an answer for you. I probably would trust the graph but their may be factors that I don't know.
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Can higher pressures mask leaks?

Post by palerider » Sat Jan 10, 2015 11:56 am

VictorC wrote: Can higher pressures sometimes mask leaks, or is this a normal phenomenon?
no, higher pressures can't hide leaks.

leaks are an absolute, (at least the top line), measured by the flow sensor in the machine. it can tell exactly how much air is flowing through the hose.

you just happened to get lucky and instead of the higher pressure making the leak worse, either you moved a bit and sealed up the leak, or somehow, the pressure caused the mask to fit better

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Last edited by palerider on Sat Jan 10, 2015 1:34 pm, edited 1 time in total.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

VictorC
Posts: 62
Joined: Thu Oct 16, 2014 3:54 pm
Location: Belleville, IL

Re: Can higher pressures mask leaks?

Post by VictorC » Sat Jan 10, 2015 12:58 pm

Thanks Palerider. I thought of that, but this was the first time I had seen a leak graph like that. It was also the first time I had experienced a large leak.

@Julie, thanks, I did read the thread from the other day, and I appreciate all the suggestions made. But in this case, I was just wondering if the graph was accurate. I assumed it was, but it seemed odd to me.