AHI Question

Post by roadkingrick » Wed Dec 10, 2014 1:33 pm

What exactly, Does 0.3 AHI mean? I understand the whole numbers representing the 10 second pauses, but whats the .3 mean 3/10ths ? I have an S9

_________________
Humidifier
Additional Comments: currently usin PR system one bipap
Never underestimate the restorative effects of a good nights sleep

User avatar
Pugsy
Posts: 65400
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: AHI Question

Post by Pugsy » Wed Dec 10, 2014 1:42 pm

AHI is a per hour average of all the events over the entire night.
So yes, three tenths per hour average. You probably had only 1 or 2 events over the entire night and diving that 1 or 2 by the number of hours when the machine was on will give you the per hour average.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

roadkingrick
Posts: 61
Joined: Sun Aug 22, 2010 12:24 am

Re: AHI Question

Post by roadkingrick » Wed Dec 10, 2014 1:58 pm

well that was easy, thanks so 5hrs with a ahi of 5.0 is one an hour. Hence 1 in 5 hours would be 0.2 ?

_________________
Humidifier
Additional Comments: currently usin PR system one bipap
Never underestimate the restorative effects of a good nights sleep

User avatar
Pugsy
Posts: 65400
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: AHI Question

Post by Pugsy » Wed Dec 10, 2014 2:06 pm

I think the machines just do tenths (one decimal point) on the LCD screen. Sleepyhead software goes to hundredths or 2 decimal points.

Less than 1.0 AHI usually means next to nothing in terms of number of events if there were any hours of use at all.
Pretty darn good and hard to beat.

AHI is an average per hour over the entire sleep session. The apnea events may have all happened in one hour or spread out over entire sleep session.
AHI of 5.0 means an average of 5 per hour so over 5 hours the total number of events would be 25.
AHI of 1.0 means an average of 1 per hour over 5 hours and the total number of events would be 5.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
Last edited by Pugsy on Wed Dec 10, 2014 6:17 pm, edited 1 time in total.
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: AHI Question

Post by palerider » Wed Dec 10, 2014 4:17 pm

roadkingrick wrote:well that was easy, thanks so 5hrs with a ahi of 5.0 is one an hour. Hence 1 in 5 hours would be 0.2 ?
no an ahi of 5 is 5 per hour, on average, so for 5 hours, 25 incidents.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: AHI Question

Post by Julie » Wed Dec 10, 2014 4:21 pm

Agree with PR.

User avatar
Tatooed Lady
Posts: 984
Joined: Sat Jun 07, 2014 6:18 pm
Location: Central Wisconsin

Re: AHI Question

Post by Tatooed Lady » Wed Dec 10, 2014 6:16 pm

palerider wrote:
roadkingrick wrote:well that was easy, thanks so 5hrs with a ahi of 5.0 is one an hour. Hence 1 in 5 hours would be 0.2 ?
no an ahi of 5 is 5 per hour, on average, so for 5 hours, 25 incidents.
I thought my brain had shorted out again. Thanks for clearing the fog.

_________________
Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Precious and POW are very very good to me.
As Bette Davis famously said, “Old age ain’t for sissies.”
I'm with the band.
So.Many.TOYS!

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: AHI Question

Post by palerider » Wed Dec 10, 2014 6:57 pm

Tatooed Lady wrote:
palerider wrote:
roadkingrick wrote:well that was easy, thanks so 5hrs with a ahi of 5.0 is one an hour. Hence 1 in 5 hours would be 0.2 ?
no an ahi of 5 is 5 per hour, on average, so for 5 hours, 25 incidents.
I thought my brain had shorted out again. Thanks for clearing the fog.
confusing a custodian of such dangerous machinery as you operate is not something I'd like to see happen!

you might be musing over that puzzling bit of misinformation while driving and... well, let's just not.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Gparr
Posts: 125
Joined: Thu May 23, 2013 4:38 pm

Re: AHI Question

Post by Gparr » Thu Dec 11, 2014 5:56 am

Not to hijack this thread...but Ido have an AHI question.

I am just passing the 1.5 years of therapy. My AHI has been sitting on .3 to .8 for quite awhile. Should I (or are there any ways to) try and get it to 0?

_________________
MachineMask
Murphy's New Law

If it doesn't fit, force it...if it breaks, it was meant too!

User avatar
Pugsy
Posts: 65400
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: AHI Question

Post by Pugsy » Thu Dec 11, 2014 10:07 am

Gparr wrote: My AHI has been sitting on .3 to .8 for quite awhile. Should I (or are there any ways to) try and get it to 0?
I wouldn't try if it were me. Less than 1.0 AHI means maybe one or two or three events during the night and depending on what they are there might not be anything that can be done anyway. Like if the flagged event was a central..it's normal to have an occasional central anyway and we can't fix them with cpap machine anyway.
Plus it's possible that the machine has maybe flagged awake breathing SWJ irregularities as some sort of event and it really isn't worthy of a flag.

AHI 0.0 doesn't guarantee anything except seeing a nice math score. Doesn't guarantee the "miracle" or really much of any change in how a person might feel.

I remember a long time ago a newbie who was lucky enough to get an AHI of 1.4 right from the start but was so intent on getting the 0.0 thing that he/she kept raising the pressure in an effort to kill all events and all that happened was a really bad case of aerophagia. The AHI never really changed but the belly pain got to be really bad.
AHI 0.0 is nice to see when it happens but it is an unrealistic goal to strive for.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: AHI Question

Post by palerider » Thu Dec 11, 2014 10:33 am

Gparr wrote:Not to hijack this thread...but Ido have an AHI question.

I am just passing the 1.5 years of therapy. My AHI has been sitting on .3 to .8 for quite awhile. Should I (or are there any ways to) try and get it to 0?
the relentless quest for 'perfection' will quite possibly result in worse sleep.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
cathyf
Posts: 515
Joined: Mon Sep 01, 2014 11:56 am

Re: AHI Question

Post by cathyf » Thu Dec 11, 2014 11:50 am

One thing to remember about AHI is that it's an average, and so it could possibly be hiding something serious. If you have short events that don't wake you up, scattered throughout the night, then you are getting plenty of good sleep. But if the events are long or clustered in such a way that your oxygen levels are getting compromised, then that IS serious. Remember that AHI is a count of events, and if you stop breathing long enough to cause brain damage or death it only takes one!

That would be pretty unusual -- usually if events are big they are also going to be frequent. But this is why you shouldn't just focus on one number, you need to put your numbers in context. And this is what is scary about using a brick -- your first sign that your (low) AHI isn't the whole story could be waking up dead...

bv1800
Posts: 118
Joined: Tue May 20, 2014 6:30 pm

Re: AHI Question

Post by bv1800 » Thu Dec 11, 2014 11:57 am

roadkingrick wrote:What exactly, Does 0.3 AHI mean? I understand the whole numbers representing the 10 second pauses, but whats the .3 mean 3/10ths ? I have an S9
3/10 is right, but not 3/10ths.... 3 per 10 hours, is the way to interpret it (it's an average rate per hour so .3 in 1 hour becomes 3 in 10 hours). You probably didn't sleep for 10 hours, so it's likely that you had 2 in roughly 6 hrs 40 min (give or take 10 min or so) or 1 in 3 hrs 20 min (give or take 5 min or so).