Cannot Control AHI Consistently

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
ronlsujones
Posts: 4
Joined: Fri Jul 20, 2012 7:21 am
Location: Friendswood, Texas

Cannot Control AHI Consistently

Post by ronlsujones » Mon Dec 08, 2014 9:43 pm

During the spring of 2012, I participated in 4 sleep studies over a 2 month period. The data resulted in a prescribed pressure of 26 mm water. Yes, that is correct 26 mm water. As I read most posting on this site and listen to people complaining about "high pressure" prescriptions of 10 or maybe even 15 cm water. Strap on 26 cm water for a while and you will understand high pressure. Just achieving that pressure is an amazing feat. Initially, it was next to impossible but I finally mastered using full face masks until they wreck the bridge of my nose then switching to nasal masks for a few days.

The pathetic thing about my situations is even with the compliance with prescribed pressure, AHI sucks most of the time. During a single week it varied from 3.2 to 35.8 and I don't have a clue about the causes of either the low or high results. I will have a night where I actually sleep continually and think that the index will be low but it isn't. I take a lot of medication including blood pressure drugs, blood thinners and a large amount of pain medication daily. Treatment for pain includes three 20 mg extended release Oxycodone pills along with four 15 mg Oxycodone pills per day because of a staph infection of the spine experienced a few years ago.

There are two things that cause me to be totally puzzled. The first is after a few months of therapy my BP medication requirement decreased to the minimum allowed then gradually increased back to maximum. I do not understand what caused it to drop but surely would like to know. The other confusing situation involves the dramatic variations of AHI. If I could understand what causes the AHI of 3.2 and learn how to maintain that performance, obviously my problems would be solved.

I just switched from the initial sleep doctor who seemed to know less about apnea than I to a highly recommended doctor who really seems to care. The problem I now have is after hoping that the new physician would have an answer to my questions I learned that he does now. When I switched, I fully anticipated being subjected to numerous sleep studies again but that did not happen. He spend significant time studying my initial studies and data I provided on performance during the last few months, said something about "being realistic with expectations" and suggested seeing him again in 3 months.

Hopefully, someone else had problems similar to mine and found a solution? I review my performance data frequently but there just doesn't seem to be any logic to what I observe. While I didn't spend a lot of time bitching about it, yes I fell like hell every day. On weekends, I often take a nap shortly after noon and during the week if I don't sleep for an hour at 5 or even 6 PM I am falling asleep by 8 PM. As mentioned BP medication is at maximum and obviously the doctors don't know what to do. There is little doubt in my mind that the problems are caused by apnea but I can't figure out a solution to the problem. Any help will be much appreciated.

Ron

User avatar
kteague
Posts: 7788
Joined: Tue May 16, 2006 8:30 pm
Location: West and Midwest

Re: Cannot Control AHI Consistently

Post by kteague » Mon Dec 08, 2014 11:30 pm

I'm sure someone will come who who can talk with you about the specifics of yours treatment. I'm wondering why 4 studies in 2 months - did you end up having to get titrated on an ASV machine? I have a friend whose prescribed pressure is 25, and that high pressure triggered centrals and so he uses an ASV. Does your data show your events are obstructives or centrals? The suggestions you get will be more targeted and helpful with more information. Good luck pinpointing ways to make your treatment better for you.

_________________
Mask: TAP PAP Nasal Pillow CPAP Mask with Improved Stability Mouthpiece
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Bleep/DreamPort for full nights, Tap Pap for shorter sessions

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Cannot Control AHI Consistently

Post by palerider » Mon Dec 08, 2014 11:39 pm

ronlsujones wrote:During the spring of 2012, I participated in 4 sleep studies over a 2 month period. The data resulted in a prescribed pressure of 26 mm water. Yes, that is correct 26 mm water.
hmm, my machine starts out at 100mm of water... and goes up to as much as 250mm h20.

are you SURE you're correct?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Tatooed Lady
Posts: 984
Joined: Sat Jun 07, 2014 6:18 pm
Location: Central Wisconsin

Re: Cannot Control AHI Consistently

Post by Tatooed Lady » Tue Dec 09, 2014 6:19 am

I think palerider is pointing out that these use cm of pressure, not mm. He's a real wisecracker. generally harmless, though.

_________________
Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Precious and POW are very very good to me.
As Bette Davis famously said, “Old age ain’t for sissies.”
I'm with the band.
So.Many.TOYS!

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Cannot Control AHI Consistently

Post by palerider » Tue Dec 09, 2014 9:42 am

Tatooed Lady wrote:I think palerider is pointing out that these use cm of pressure, not mm. He's a real wisecracker. generally harmless, though.
there's also the fact that there's a vanishingly small number of machines that'll go over 25cm, though it is possible. (I know of one in the resmed line) if he is using an avaps, then that is likely one of the few that will go up to 30cm

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Last edited by palerider on Tue Dec 09, 2014 1:24 pm, edited 1 time in total.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Pugsy
Posts: 65381
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Cannot Control AHI Consistently

Post by Pugsy » Tue Dec 09, 2014 10:12 am

viewtopic.php?f=1&t=86579&p=789058#p789058
Supposedly using a Respironics ASV type of machine...humidifier points to a PR S1 machine if that is correct...so with the 2013 date..probably a model 950 maybe...I never could find where model was confirmed.

Would need to see some reports to offer any ideas why the wide range of AHI numbers but I suspect they are related to the opiate meds. Either centrals from suppressed respiration or obstructives due to opiates causing the airway tissues to really relax.
This much opiate meds is going to play hell with both centrals and obstructives.

Need to see some detailed reports from a "good" night with low AHI and a "bad" night with high AHI.
ronlsujones wrote:I take a lot of medication including blood pressure drugs, blood thinners and a large amount of pain medication daily. Treatment for pain includes three 20 mg extended release Oxycodone pills along with four 15 mg Oxycodone pills per day because of a staph infection of the spine experienced a few years ago.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
SGearhart
Posts: 434
Joined: Fri Nov 07, 2014 8:54 am
Location: Georgia

Re: Cannot Control AHI Consistently

Post by SGearhart » Tue Dec 09, 2014 10:37 am

Certain pain abatement drugs can act as a stimulant. At the same time, certain blood pressure medicines can act as a sedative, Are you taking these drugs at the same time of the day? If so, how long before you go to bed? Analyze your AHI's and when you took each drug for that day. You might find a correlation there. Keep posting your findings.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS-50F & 50IW - SleepyHead 1.0.0-Beta

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Cannot Control AHI Consistently

Post by Julie » Tue Dec 09, 2014 10:42 am

Yes, I believe the oxy's are probably having a huge effect and you might want to work on getting them down (with your MD's help) because I think you're fighting the machine, your body, the meds, etc. etc. I don't think you'll get anywhere reasonable until you address the meds and whatever else might be complicating things. Cpap is not a magic bullet and can only achieve so much (tho' it can certainly lower your BP all things being equal).

User avatar
Ogeo
Posts: 186
Joined: Tue Aug 05, 2014 5:12 pm
Location: Idaho

Re: Cannot Control AHI Consistently

Post by Ogeo » Tue Dec 09, 2014 12:21 pm

ronlsujones wrote: The data resulted in a prescribed pressure of 26 mm water. Yes, that is correct 26 mm water. As I read most posting on this site and listen to people complaining about "high pressure" prescriptions of 10 or maybe even 15 cm water. Strap on 26 cm water for a while and you will understand high pressure. Just achieving that pressure is an amazing feat. Initially, it was next to impossible but I finally mastered using full face masks until they wreck the bridge of my nose then switching to nasal masks for a few days.
I can relate to you and the "high" pressure bit. Mine starts at 21 and goes up to 25. I have to just laugh when people report that their "super high" pressure is 10. I actually like the force of the air blowing at me because I felt like I could breathe for the first time in my life. My biggest problem adjusting to my ASV was finding a mask that would stay on my face. I went through 8 before finding the one I'm using now. And believe it or not, it is a pillow mask. The FFM would keep me up all night with loud farting noises. If I tightened it, I was too sore to wear it or it put pressure on my eyeballs. I believe you are the first person I noticed on these forums that has higher pressure set than me.

With one mask, I actually added another strap around it to help hold it to my face so I didn't have to overtighten it. If you try something like that, then make sure the vent on the mask is left open and not covered with anything. You don't want to re-breathe your expelled air.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments:  Max PS - 11; Max EPAP - 18; min PS - 7; Min EPAP - 14; rate auto; rise x2; pressure averages between 21 and 24.9
Hubby's equipment: PR System One REMstar Auto A-Flex model 560P. Doctor's orders left it open 5-20. We changed it to 8-20 after a couple nights. Start Date was 1/30/15. Oh, mask is the new nasal pico.

Discovering the world one geocache at a time!