EPR on Resmed VPAP

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Bileveluser

EPR on Resmed VPAP

Post by Bileveluser » Sat Dec 06, 2014 8:24 am

Does anybody know why EPR is listed as a feature on Resmed VPAP machines?
Wouldn't the VPAP function take the place of EPR with a much larger pressure range available?
Does EPR change the wave shape of flow when operated in the VPAP mode?
Or does the EPR only work when it is set in the straight CPAP mode?
Thanks.

User avatar
Pugsy
Posts: 65255
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: EPR on Resmed VPAP

Post by Pugsy » Sat Dec 06, 2014 8:29 am

EPR is only available on the VPAP models when in CPAP mode.
Since EPR is pretty much the same thing as the difference between EPAP and IPAP in the bilevel modes it is redundant to have it in the bilevel modes. Bilevel mode is like having unofficial EPR that is not limited to 3 cm reduction.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Bileveluser

Re: EPR on Resmed VPAP

Post by Bileveluser » Sat Dec 06, 2014 8:36 am

Thanks Pugsy! That explains it.