Daily "Plots Disabled" ??? SleepyHead

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
CPAPER Paul
Posts: 58
Joined: Sun Aug 31, 2014 3:09 am
Location: Atlanta, GA

Daily "Plots Disabled" ??? SleepyHead

Post by CPAPER Paul » Thu Nov 27, 2014 8:57 am

Using .9.8-1 successfully for several weeks, but when I open it up now and import my data, the "Daily" page graphs are blank and display "Plots Disabled" in each graph! The graphs are "turned on" and each parameter is "green" in the lower right side. What am I doing wrong?

Thanks,

Pau

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: SleepMapper, SleepyHead Software

Guest

Re: Daily "Plots Disabled" ??? SleepyHead

Post by Guest » Tue Feb 17, 2015 11:25 am

I'm having the same issue, only it's my first day with my remstar system one auto and with sleepyhead.

did you ever get an answer to your question about why it says "plots disabled"?

User avatar
SGearhart
Posts: 434
Joined: Fri Nov 07, 2014 8:54 am
Location: Georgia

Re: Daily "Plots Disabled" ??? SleepyHead

Post by SGearhart » Tue Feb 17, 2015 12:08 pm

Go to File->Preferences->Waveforms . Select the "Rest Defaults" in the lower left. Select OK and see if that works. If not try restarting the program.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS-50F & 50IW - SleepyHead 1.0.0-Beta

perasso
Posts: 13
Joined: Tue May 06, 2008 10:18 am
Location: Seattle WA

Re: Daily "Plots Disabled" ??? SleepyHead

Post by perasso » Tue Feb 17, 2015 2:28 pm

I finally found the problem with my setup.

My ~/Documents directory was write protected (read only). As a result, sleepyhead was unable to create it's data directory. once I write enabled ~/Documents and reinstalled and restarted sleepyhead, everything, including the plots, works.

_________________
Machine: Airsense 10 Card to Cloud
Additional Comments: I use Oscar, but can't seem to select it.

menaralf
Posts: 2
Joined: Sun Jul 31, 2016 10:39 am

Re: Daily "Plots Disabled" ??? SleepyHead

Post by menaralf » Sun Jul 31, 2016 10:42 am

Hi I have the same problem! Which directory do you mean exactly?

User avatar
Julie
Posts: 20025
Joined: Tue Feb 28, 2006 12:58 pm

Re: Daily "Plots Disabled" ??? SleepyHead

Post by Julie » Sun Jul 31, 2016 11:02 am

You're replying to an old post - doubtful if they're around to answer.

Guest

Re: Daily "Plots Disabled" ??? SleepyHead

Post by Guest » Sun Jul 31, 2016 1:02 pm

menaralf wrote:...........Which directory do you mean exactly?
The default is usually
C:\Users\YourName\Documents If you don't find it, post your operating system. Or better yet, in SleepyHead, under the File menu, click on "Change Data Folder." Windows explorer should open pointed to the SH data folder, whose path you can read at the top. Be sure to cancel out of this window, so as not to change the path.

-Ron