Fix: Respironics System One (50 series) Humidifier Problems

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
breathright
Posts: 33
Joined: Tue Dec 12, 2006 3:49 pm
Location: Northern VA

Re: Fix: Respironics System One (50 series) Humidifier Problems

Post by breathright » Sun Nov 01, 2015 8:49 am

Thanks for the post! My humidifier just stopped working so I opened it up and it was exactly like your posted pictures. Took about 5 min to repair, thanks again.

_________________
MachineMask
CPAP Pressure Setting = 9 cm, C-Flex Setting =3
APAP Pressure Range = 9cm-14cm, A-flex setting = 2
User since 11/10/05

User avatar
Sir NoddinOff
Posts: 4190
Joined: Mon May 14, 2012 5:30 pm
Location: California

Re: Fix: Respironics System One (50 series) Humidifier Problems

Post by Sir NoddinOff » Sun Nov 01, 2015 9:29 am

breathright wrote:Thanks for the post! My humidifier just stopped working so I opened it up and it was exactly like your posted pictures. Took about 5 min to repair, thanks again.
Another satisfied customer <clapping> (mostly directed to the helpful CPAPtalk techies, not so much towards Phillips Respironcs )

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2.
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Fix: Respironics System One (50 series) Humidifier Problems

Post by Goofproof » Sun Nov 01, 2015 7:35 pm

css28 wrote:I opened up my wife's machine to proactively exercise connectors and apply dielectric grease to each and find the charring shown in this presentation. Her humidifier had continued to work but I wanted to check out the situation.
I found that the only force pressing the connector contact to the tang from the pcb is the tight fit of the plastic housing (the contact only presses on one side of the tang). With the plastic melted like that, the contact force relaxes, accelerating the self-destruction.

I opted to clean the tang with a file but leave the connector intact. By bending the tangs down towards the board I've got positive contact pressure now as the tangs clamp the connector against the board.
I applied dielectric grease to the connection and the humidifier continues to work a couple of months later.
I may have to solder things eventually but decided to give this a try for now.
The plastic does nothing to make the connection, except hold the wires insulated and lined up. What makes the connection is the end on the wire (female) is split and smaller than the pin. Pushing it on makes a forced fit, thats what make the connection work. If the pin isn't incerted straight on the female pin spreads, resulting in a poor connection and failure.

A proper connection has close to zero ohms resistance, resulting in no heat buildup in the wiring. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

padraignotlad

Re: Fix: Respironics System One (50 series) Humidifier Problems

Post by padraignotlad » Sat Mar 19, 2016 1:01 pm

Wow! I had the exact same problem! I was thinking about the exact same fix also! And it works again! Thanks.

However, it is a little scary this burnt connection issue. I remember waking up in the middle of the night to this odd smell coming from my CPAP. At the time, I thought maybe a bug was sucked in or something, but after I took it apart and saw the burnt plastic connector, I'm a little freaked out that I was breathing that in while it happened. Not good. There should be a recall on this if you ask me. Your picture looked just like mine.

User avatar
ColCpap
Posts: 20
Joined: Tue Mar 22, 2011 3:25 pm
Location: Little Rock, AR

Re: Fix: Respironics System One (50 series) Humidifier Problems

Post by ColCpap » Sat Apr 02, 2016 10:05 am

Perfect! When I opened mine up... the same exact situation. the burn was identical, which was cool and freaky at the same time. Thank you for the great post!!

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Fix: Respironics System One (50 series) Humidifier Problems

Post by palerider » Sat Apr 02, 2016 1:05 pm

ColCpap wrote:Perfect! When I opened mine up... the same exact situation. the burn was identical, which was cool and freaky at the same time. Thank you for the great post!!
I read something the other day about the *previous* generation of respironics machines having problems with humidifier heater connections burning....

I think their engineers need to be slapped with a basic electrics handbook, repeatedly.

I wonder if the new dreamstations are as poorly engineered.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Fix: Respironics System One (50 series) Humidifier Problems

Post by Wulfman... » Sat Apr 02, 2016 6:50 pm

palerider wrote:
ColCpap wrote:Perfect! When I opened mine up... the same exact situation. the burn was identical, which was cool and freaky at the same time. Thank you for the great post!!
I read something the other day about the *previous* generation of respironics machines having problems with humidifier heater connections burning....

I think their engineers need to be slapped with a basic electrics handbook, repeatedly.

I wonder if the new dreamstations are as poorly engineered.
I think almost EVERY generation of Respironics machines have had some types of "issues" or recalls with some of their humidifier units or tanks. There was an old recall of some of the Legacy models, the M Series had leaking problems and went through several re-design configurations and then there were the System One problems.

Since I don't use heated humidity, I've never worried about it. But, the stories and links are out there.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

BenD

Re: Fix: Respironics System One (50 series) Humidifier Problems

Post by BenD » Mon Apr 25, 2016 7:42 pm

Thanks so much....fixed mine today......

User avatar
capt_caper
Posts: 4
Joined: Thu Mar 28, 2013 2:35 pm

Re: Fix: Respironics System One (50 series) Humidifier Problems

Post by capt_caper » Fri Jan 06, 2017 11:37 am

Here's another one.

I'm a really happy camper I found this. I was about to buy a replacement humidifier. Turned out to be an easy repair.

I hope others find this when faced with a Respironics RemStar CPAP that has a blinking or flashing humidifier light.

Image

bob b

Re: Fix: Respironics System One (50 series) Humidifier Problems

Post by bob b » Thu Jul 13, 2017 11:49 am

Saw this after a search two screws and saw same thing. Fixed in 20 minute all good. you are the man.

jaychuah
Posts: 1
Joined: Mon Nov 20, 2017 7:24 pm

Re: Fix: Respironics System One (50 series) Humidifier Problems

Post by jaychuah » Mon Nov 20, 2017 7:28 pm

little burn on the connctor but didnt require soldering, just removed the connector, cleant it, ad plug it back its working!hopefully wont burn futher in future. thanks again for this post!

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Fix: Respironics System One (50 series) Humidifier Problems

Post by palerider » Tue Nov 21, 2017 1:46 am

jaychuah wrote:little burn on the connctor but didnt require soldering, just removed the connector, cleant it, ad plug it back its working!hopefully wont burn futher in future. thanks again for this post!
I'd suggest a regular reseating of the connector, to scrape off the oxidation that builds up, causing increased resistance and increasing heat.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

capon
Posts: 7
Joined: Sat May 27, 2017 9:28 am
Location: Columbus, OH

Re: Fix: Respironics System One (50 series) Humidifier Problems

Post by capon » Fri Aug 03, 2018 3:01 pm

Thanks so much for this info. I recently open up my machine and saw the exact burnt connector. I was going to send it off to a repair shop but I'm willing to give this fix a try. I ended up cutting off the previous connector, tinning the wire and corresponding connector pin, soldering the wire and protecting the joint with some heat shrink tubing I had on hand. The result? Works like a champ! I just saved several $$ plus the inconvenience of going more nights without this machine.
Looking forward to get a great night's sleep without snoring and sending my better half off to the spare room. :)
Attachments
20180803_155431.jpg
20180803_155431.jpg (318.66 KiB) Viewed 4806 times

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Fix: Respironics System One (50 series) Humidifier Problems

Post by chunkyfrog » Fri Aug 03, 2018 3:19 pm

Almost looks like a fusible link in there.
Worst thing Chrysler ever did.
Imagine being 400 miles from home and a fusible link goes out on your motorhome.
Before the age of smartphones.
Dealer wanted a pile of moolah to install one--but refused to sell just the part.
Stinkers! We found a parts store, limped there, and the bullfrog did his thing in their parking lot.
Needless to say, we NEVER owned another Chrysler product, and sold the Dodge class C.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Fix: Respironics System One (50 series) Humidifier Problems

Post by palerider » Fri Aug 03, 2018 3:37 pm

chunkyfrog wrote:
Fri Aug 03, 2018 3:19 pm
Almost looks like a fusible link in there.
I think that's the heat shrink ready to go on, but that's just a guess.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.