Can Heat/Humidity Be Too High?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Paralel
Posts: 216
Joined: Thu Aug 14, 2014 7:49 pm

Can Heat/Humidity Be Too High?

Post by Paralel » Mon Oct 13, 2014 1:54 pm

Is it possible to set the heat/humidity so high on a heated hose/heated humidifier machine that it can actually increase your AHI?

User avatar
BlackSpinner
Posts: 9742
Joined: Sat Apr 25, 2009 5:44 pm
Location: Edmonton Alberta
Contact:

Re: Can Heat/Humidity Be Too High?

Post by BlackSpinner » Mon Oct 13, 2014 2:07 pm

If it cause inflammation it is possible.
Everybody is different - some need it high and others need it low and it also depends on the season, temperatures, room humidity and the colour of the sheets.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Additional Comments: Quatro mask for colds & flus S8 elite for back up
71. The lame can ride on horseback, the one-handed drive cattle. The deaf, fight and be useful. To be blind is better than to be burnt on the pyre. No one gets good from a corpse. The Havamal

User avatar
Pugsy
Posts: 65414
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Can Heat/Humidity Be Too High?

Post by Pugsy » Mon Oct 13, 2014 2:07 pm

Unlikely at least directly.

If too much causes congestion the nose may stop up and maybe..big maybe...fool the machine into thinking nasal congestion is airway congestion and get you some flags.
Remember where the airway collapses for it to be sleep apnea...it isn't in the nose.

Or maybe too much causes discomfort and restlessness and a person has a lot of wake/semi awake/tossing and turning time on the machine and stuff gets flagged by mistake. So maybe indirectly elevating the AHI but awake events don't count anyway.

Will extra moisture actually cause the airway tissues down by the tongue and below to actually collapse more...I have never heard of it or any situation where it happened. I have never heard of moisture causing the airway tissues to get more floppy/saggy.
Sleeping position ..yes
REM sleep ...yes
Extra moisture...never heard of it and I can't think of any situation where moisture makes the airway tissue collapse more.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Paralel
Posts: 216
Joined: Thu Aug 14, 2014 7:49 pm

Re: Can Heat/Humidity Be Too High?

Post by Paralel » Mon Oct 13, 2014 2:13 pm

Sigh. I guess I'm just grasping at straws to try and get my numbers back to where they were before.

I guess I just have to accept that my numbers will probably not be as good as they were, possibly ever again, now that I have partial paralysis of my laryngeal and pharyngeal muscles...

User avatar
Pugsy
Posts: 65414
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Can Heat/Humidity Be Too High?

Post by Pugsy » Mon Oct 13, 2014 2:21 pm

Refresh my memory...what were they before and what are they now and which category is worse.
I can't remember and too lazy to go looking through your past threads.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Wulfman...
Posts: 6694
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Can Heat/Humidity Be Too High?

Post by Wulfman... » Mon Oct 13, 2014 2:32 pm

Paralel wrote:Is it possible to set the heat/humidity so high on a heated hose/heated humidifier machine that it can actually increase your AHI?
I can remember posts from years past where users remarked that their AHI decreased when they reduced their humidifier settings. They felt that the heated humidity caused more nasal congestion......and if you can't breathe through your nose, you have to try to breathe through your mouth......and then your tongue can fall back in your throat and result in apneas.

In my own case, I tried heated humidity only about once or twice when I started therapy.......before I disconnected the cord from the heated humidifier plate to my machine......I didn't want to accidentally turn it on. I like breathing the cool air.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Can Heat/Humidity Be Too High?

Post by archangle » Mon Oct 13, 2014 3:02 pm

I think this is one of the many things that are just trial and error.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
Captain_Midnight
Posts: 761
Joined: Sat Jan 28, 2006 8:10 pm
Location: The Great State of Idaho

Re: Can Heat/Humidity Be Too High?

Post by Captain_Midnight » Mon Oct 13, 2014 5:32 pm

Can heat/humidity be too high?

Yes.

If I use heated humidity, my AHI goes up, and varies directly with the amount of heat. (Meaning the effect is related to the temp and moisture.)

I have done the experiment many, many times (sometimes inadvertently) , and it's typically the same result.

Some folks have the opposite experience; and, that's fine.

What do I do for minimum AHI and max apap benefit? Either no humidification, or humidification in passover (no heat) mode.

What might account for the differences?

I'm not sure. But just like Wulfman Den mentioned, the direct correlation between AHI and heated humidification does exist in some apnea patients, and has been mentioned hereabouts several times. (Occasionally by me.)

.

.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: ComfortGel Blue Nasal CPAP Mask with Headgear
Additional Comments: APAP range = 10 - 12.5 In H20

Paralel
Posts: 216
Joined: Thu Aug 14, 2014 7:49 pm

Re: Can Heat/Humidity Be Too High?

Post by Paralel » Mon Oct 13, 2014 5:53 pm

I guess its worth playing with the settings a little more then and see how it goes.

User avatar
tortoisegirl
Posts: 167
Joined: Thu Jan 16, 2014 6:52 am
Location: WA

Re: Can Heat/Humidity Be Too High?

Post by tortoisegirl » Mon Oct 13, 2014 8:39 pm

Agreed to mess with the settings, but to try to do it in a controlled manner and give yourself some time (maybe 2-3 nights) at each setting, so you don't make a decision based on a fluke night. Just in case you didn't know, there is also a Classic mode where heat & humidity can be adjusted independently (the default is to adjust temperature only). It runs more humid in Classic mode, which for some folks causes rain out or even for the tank to run dry during the night (I've had neither problem, but I have a Resmed not a PR machine so its not really applicable).

As for me, the heat & humidity is never too high. I actually wish the settings went higher! If I don't max out the temperature I feel the airflow much more and it feels cold against my upper lip. If I don't max out the humidity my dry mouth upon waking is even worse. I find high heat & humidity help with congestion, although I suspect I mouth breath 90%+ of the time once asleep. I never seem to get enough air when breathing nasally while lying down. Best wishes.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CPAP pressure of 5; diagnosed AHI=9; also have PLMD & insomnia

User avatar
Pugsy
Posts: 65414
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Can Heat/Humidity Be Too High?

Post by Pugsy » Mon Oct 13, 2014 9:32 pm

tortoisegirl wrote:Just in case you didn't know, there is also a Classic mode where heat & humidity can be adjusted independently (the default is to adjust temperature only). It runs more humid in Classic mode,

Actually Classic mode in Respironics machines doesn't offer any temperature settings at all. It's strictly heat to the water chamber only...higher setting is higher heat to the water chamber and doesn't utilize any humidity sensors and more heat means more moisture output.
It's simply the way we used to do humidity for years before all the sensors and heated hoses.
It does put out more humidity though and chance of rain out is indeed greatly increased because the hose doesn't warm the air.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

dduttonnc
Posts: 4
Joined: Mon Oct 13, 2014 10:22 pm
Location: Charlotte, nc, USA

Re: Can Heat/Humidity Be Too High?

Post by dduttonnc » Mon Oct 13, 2014 11:03 pm

Paralel wrote:Is it possible to set the heat/humidity so high on a heated hose/heated humidifier machine that it can actually increase your AHI?
I'll say what others said. Set it what's best for YOU

Get a heated hose if u lack one. Without one air temp is not regulated and then condensation happens and you can breathe in large amounts of water.

The resmed s9 series with climate max hose has auto and manual

In auto, you set the air temp, and it sets a preset relative humidity to that temp. That doesn't work for me

I need humidity on 6.0 and air temp 80-86. With a humidity setting of 6 , any temp under 80 will cause rain out (condensation).

My ins won't pay for the heated tube , but if u care for it, clean it regularly, it lasts 6-9months
Pugsy wrote:
tortoisegirl wrote:Just in case you didn't know, there is also a Classic mode where heat & humidity can be adjusted independently (the default is to adjust temperature only). It runs more humid in Classic mode,

Actually Classic mode in Respironics machines doesn't offer any temperature settings at all. It's strictly heat to the water chamber only...higher setting is higher heat to the water chamber and doesn't utilize any humidity sensors and more heat means more moisture output.
It's simply the way we used to do humidity for years before all the sensors and heated hoses.
It does put out more humidity though and chance of rain out is indeed greatly increased because the hose doesn't warm the air.

_________________
MaskHumidifier
Additional Comments: I don't use software. My machine has an sd card and an ATT modem so I don't send the sd card in. My dr gets realtime updates
I have both CSA & OSA. OSA 75%, CSA 25%. My sleep studies showed I stoppbreathing over ***900*** times in an 8 hrs for seconds and up to 3-4min. Now it's 0-3Xs/night. I feel like 25 again and my libido restored to what it was 20 YEARS ago