Sleepyhead question

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
JimP
Posts: 181
Joined: Thu Oct 02, 2014 2:49 pm

Sleepyhead question

Post by JimP » Sat Oct 11, 2014 5:51 am

I'm interested in using the sleepyhead software but am concerned about something I think I read in the instructions.

I'm on a Phillips Respironics System One bipap machine. If I understood the instructions correctly, if you're using a PR System One (I presume this is the same machine) then you have to be sure to lock the memory card as sleepytime will write something on the card otherwise. Then if the card is written on, the bipap machine won't read it.

Given the likelyhood that I'll forget to lock the memory card, I'm concerned that I'll do something that'll cause problems in my bipap machine's recording/reporting routine.

Is the locking of the memory card for Phillips Respironic System One bipap machines a requirement?

If I were to forget to lock the card, would it be a simple matter of deleting some files?

With the machine, I can use sleepmapper software which is very limited. I'd like to know more about what's going on.

User avatar
grayghost4
Posts: 1554
Joined: Wed Aug 20, 2014 6:52 pm
Location: Norther Illinois
Contact:

Re: Sleepyhead question

Post by grayghost4 » Sat Oct 11, 2014 7:17 am

I have a pr760 bipap and use sleepyhead .... locking the card is not necessary.
That warning applies mostly to Resmed machines. It will not hurt to lock the card .... just remember to unlock it when returning it to the machine.
If you're not part of the solution you're just scumming up the bottom of the beaker!

Get the Clinicians manual here : http://apneaboard.com/adjust-cpap-press ... tup-manual

User avatar
raftergirl
Posts: 83
Joined: Fri Aug 29, 2014 6:54 pm
Location: Salt Lake City, Utah

Re: Sleepyhead question

Post by raftergirl » Sat Oct 11, 2014 7:24 am

I use a PR machine and SleepyHead. I've never locked my card and never had a problem. I download my data every morning, take a look, then pop the SD card back in my machine. I recently turned the SD card in to my DME/sleep doc for my 2 month check-up, and they had no problems getting the info from it using a different program.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Mask: AirFit™ N10 For Her Nasal CPAP Mask with Headgear
Additional Comments: SleepyHead software

User avatar
Bill44133
Posts: 1087
Joined: Tue Jul 09, 2013 8:34 pm
Location: North Royalton, OH

Re: Sleepyhead question

Post by Bill44133 » Sat Oct 11, 2014 7:28 am

You are getting it confused with the Resmed products. PR System One does not care if you put extra files on the disk..


I wish you well

_________________
Machine: DreamStation BiPAP® Auto Machine
Humidifier: DreamStation Heated Humidifier
Additional Comments: Settings are IPap 23 EPap 19

JimP
Posts: 181
Joined: Thu Oct 02, 2014 2:49 pm

Re: Sleepyhead question

Post by JimP » Sat Oct 11, 2014 7:36 am

Bill44133 wrote:You are getting it confused with the Resmed products. PR System One does not care if you put extra files on the disk..


I wish you well

Thanks

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Sleepyhead question

Post by palerider » Sat Oct 11, 2014 9:25 am

JimP wrote:I'm interested in using the sleepyhead software but am concerned about something I think I read in the instructions.

I'm on a Phillips Respironics System One bipap machine. If I understood the instructions correctly, if you're using a PR System One (I presume this is the same machine) then you have to be sure to lock the memory card as sleepytime will write something on the card otherwise. Then if the card is written on, the bipap machine won't read it.
*NO*!

Sleepyhead *NEVER* *NEVER* *NEVER* EVER writes to your sd card. NEVER.

some *OPERATING SYSTEMS*, like windows 8 and macos WILL write to it, and that annoys RESMED machines, not yours.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
jedimark
Posts: 1300
Joined: Fri Feb 18, 2011 12:45 am
Location: Bundaberg, Australia
Contact:

Re: Sleepyhead question

Post by jedimark » Sat Oct 11, 2014 10:04 am

palerider wrote:
JimP wrote:I'm interested in using the sleepyhead software but am concerned about something I think I read in the instructions.

I'm on a Phillips Respironics System One bipap machine. If I understood the instructions correctly, if you're using a PR System One (I presume this is the same machine) then you have to be sure to lock the memory card as sleepytime will write something on the card otherwise. Then if the card is written on, the bipap machine won't read it.
*NO*!

Sleepyhead *NEVER* *NEVER* *NEVER* EVER writes to your sd card. NEVER.

some *OPERATING SYSTEMS*, like windows 8 and macos WILL write to it, and that annoys RESMED machines, not yours.

Just to reiterate this and be super clear, I have absolutely NO reason whatsoever to write to CPAP DataCards in SleepyHead. Everything I need to do in SleepyHead to import data is read only.

But certain things can screw up on various operating systems, with *any* CPAP datacard, and are important to be aware of, especially if you need to routinely send the datacard in for compliance reporting purposes.

Not unmounting the SDCard properly before ejecting in *any* operating system can corrupt the File Allocation Table (FAT) on the SD card.
CPAP Machines often take shortcuts in their firmware implementation of the FAT/FAT32/etc filesystem driver, and miss little features like maintaining a second FAT copy, which while not strictly required, doesn't help out much when things go wrong.

Pulling the card out in the middle of rewriting a section of the File Allocation Table pretty much guarantees filesystem corruption..
Sometimes you don't know it's in the middle of rewriting this.. because of hidden operating system cache/thumbnail files being written to the card.

ResMed adds a simple "Journal" layer on top of the filesystem, which is designed to make sure write operations are completed, but it's got some serious weaknesses when files are copied in without it's knowledge. (This is mostly Microsoft and Apples fault, but ResMed should know better!)

Anyway, to be safe, it's a good idea to be extra careful with CPAP SDCards, and always write-lock the SDcard, and unmount and eject properly even if write protected.. some card readers don't honor the write-lock switch properly.

This same caution applies to users of ResScan, Encore Pro, etc...

F&P Icon users should be extra careful to unmount properly with the smartstick, as there is no way of write protecting this device.

Hope that makes sense: Just remember to write protect on inserting the datacard (if possible) every time you import, and always unmount and eject safely...

You might get away with yanking it out with a Camera card, but CPAP machines are an awful lot less forgiving when it comes to filesystem corruption...

/Mark

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS50D+/F Oximeter, S9 VPAP Auto/Adapt, PRS1 Auto, Intellipap Auto, SleepyHead :)
Author of the free, cross platform, open-source sleep tracking software SleepyHead.
Download http://sleepyhead.jedimark.net
Source Code http://gitlab.com/sleepyhead/sleepyhead-code

js12278
Posts: 41
Joined: Thu Sep 25, 2014 8:28 pm

Re: Sleepyhead question

Post by js12278 » Sat Oct 11, 2014 10:21 am

Is there something we can do to duplicate a copy if we need to replace the SDcard ever?

_________________
MaskHumidifier
Additional Comments: Sleepyhead 9.6 test build

User avatar
bwexler
Posts: 1586
Joined: Mon Nov 08, 2010 11:52 pm
Location: San Marcos, Ca. USA

Re: Sleepyhead question

Post by bwexler » Sat Oct 11, 2014 10:41 am

Sleepyhead 0.9.8-1 automatically makes a backup copy of your data each time you import it.
Various amounts of data are stored in your machine depending on the brand and model. That data will be automatically written to a new or blank card when you insert it in your machine. That data usually includes compliance data, since that is how everyone gets paid.

_________________
Mask: SleepWeaver 3D Soft Cloth Nasal CPAP Mask with Headgear
Additional Comments: AurCurve 10 ASV Also using Sleaplyhead 1.1, ResScan 6 and CMS50i

calusa

Re: Sleepyhead question

Post by calusa » Sat Oct 11, 2014 2:21 pm

JimP wrote: always write-lock the SDcard, and unmount and eject properly even if write protected.. some card readers don't honor the write-lock switch properly.
/Mark
Can you make this clearer?

Are you saying that we should "write-lock" the SDcard before removing it form the CPAP machine?

Or before removing it from our computer after looking at the data in Sleepyhead?

User avatar
Pugsy
Posts: 65207
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Sleepyhead question

Post by Pugsy » Sat Oct 11, 2014 2:46 pm

calusa wrote: Can you make this clearer?

Are you saying that we should "write-lock" the SDcard before removing it form the CPAP machine?

Or before removing it from our computer after looking at the data in Sleepyhead?
Remove SD card from the cpap machine. You can't do anything to the slide thing to lock it while it is any the machine anyway...there is no way to get to it.
Then slide the write protect/lock on the SD card before putting it in the computer.
Do your usual download of reports.
Remove SD card and then unlock the write protect slide thingy.
Then put the SD card back into the cpap machine....if you forget to unlock it then the cpap machine can't write to the SD card when in write protect mode and will tell you ..."hey dude, this SD card is in read only mode and I can't put nothing on it this way...please unlock the SD Card."

I never bother myself. I am long past having to prove compliance or show my reports to anyone and if I have to erase a SD card it isn't the end of the world for me because I have any data I might want in my software reports anyway.
My computer is Windows 7 and it's the Macs or Windows 8.xxx that some to be the main culprit for putting extra files on the SD card and causing the ResMed machines to refuse to work with the SD card. Respironics machines don't care.
But I suppose I should say...don't do as I do unless you are like me and it isn't the end of the world to lose the data stored on the SD card.
I have enough trouble remembering to put the SD card back in the cpap machine as it is.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
RogerSC
Posts: 1914
Joined: Sat Jul 21, 2012 1:11 pm

Re: Sleepyhead question

Post by RogerSC » Sat Oct 11, 2014 6:48 pm

Having used computers quite a lot over the years (decades, actually), and having made my share of sleepy or just unthinking mistakes, I've learned to write-protect any media that has data that I care about, and keep backups *smile*. So I have my sleepyheaddata folder backed up. And always always always write-protect the SD card. It's a habit, if I take the SD card out of my cpap, I flip the write-lock tab. While I'm not worried about SleepyHead writing to my SD card, once it's mounted in my computer I could accidentally do something that I might regret *smile*. Maybe because I'm not much afraid of trying things or something *smile*. Even your computer crashing or a power glitch could corrupt your SD card.

Anyway, it's always good to have protection, you never know when the statistics will include you *smile*.

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Sleepyhead question

Post by archangle » Sat Oct 11, 2014 7:24 pm

js12278 wrote:Is there something we can do to duplicate a copy if we need to replace the SDcard ever?
Copy the entire SD card to a directory on your computer. Name it something convenient. I have a folder where I put all my SD card images, and I name the directories with today's data, for instance 141011 if it's 10/11/2014. That way, the directories sort correctly by name.

I think you can copy this back to an SD card for PRS1 machines, but I haven't tried it. I know it won't work for S9 machines, although I've heard that ResScan will accept a "bogus" SD card like that even though the S9 machine will require you to reformat.

In theory, you don't need to "eject safely" on SD cards if you haven't written anything to it. I usually don't bother on my win7 machine.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.