Without CPAP For A Week - What Position To Sleep?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
dangerousdave
Posts: 21
Joined: Sat Jul 07, 2012 6:09 pm

Without CPAP For A Week - What Position To Sleep?

Post by dangerousdave » Sat Oct 11, 2014 5:06 am

Following my cleaning of my Resmed S9 heated tube with white distilled vinegar & water mixture two days ago, I'm getting water on my face nightly. So I guess I mishandled the tube or perhaps damaged the 3 copper electrical connections on the end. Also, I don't know if I could have somehow changed a setting on the Resmed but I didn't go into Clinician Setup Mode. I'm hoping to get an appointment on Monday to request to be sent a new tube & mask but I'm going out of town for a week on Tuesday. I've replaced nothing previously in the over 2 years of ownership of my cpap.

While I'm gone, what is the best sleeping position in general, that I might try to possibly keep my airway from closing AND maybe lessen the chance of snoring? I usually sleep alone however, I may be with someone else in bed. I've never really slept on my stomach and if I sleep on my side I might need a pillow to keep me from rolling over onto my back. If it matters I'm in my 50's.
Thanks

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Without CPAP For A Week - What Position To Sleep?

Post by Julie » Sat Oct 11, 2014 5:09 am

The best position would be in a recliner, semi, if not fully upright, but if it's not practical, then a side or stomach would be better than back sleeping. I'd definitely pick up a soft cervical collar as it helps to keep your head up and your airway open, but all that being said, have you considered trying to rent a machine, even a plain cpap, for the time you'll be away? You may not realize how lousy you could end up feeling by not doing it.

User avatar
englandsf
Posts: 1016
Joined: Thu Jul 10, 2014 6:17 pm
Location: Austin TX

Re: Without CPAP For A Week - What Position To Sleep?

Post by englandsf » Sat Oct 11, 2014 5:26 am

Have you tried it without any water in the humidifier? Better than no CPAP perhaps?

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Sleepyhead SW. NeilMed and Alkalol Nasal rinses. Veramyst. AutoPAP 11-20 cms. Started June '14, untreated AHI 31-38, with PAP around 1.

Lukie
Posts: 162
Joined: Tue Oct 29, 2013 11:06 am

Re: Without CPAP For A Week - What Position To Sleep?

Post by Lukie » Sat Oct 11, 2014 5:36 am

I don't know how severe your sleep apnea is, but I, for one, would not be without my cpap even for a nap. It would be wise for you in the future to keep at least a back up hose. I would take your equipment with you and run it without a humidifier. This may give you a dry nose and mouth but at least you are keeping your airway open thus lessening your chance of dying from an apnea that you just don't wake up from.

User avatar
PST
Posts: 987
Joined: Sat May 30, 2009 9:56 pm

Re: Without CPAP For A Week - What Position To Sleep?

Post by PST » Sat Oct 11, 2014 7:50 am

There is a study that claims right side beats left side. Either side beats back. I've slept on my side in an emergency and lived to tell the tale, but I don't like it.

Many of us find humidifiers to be more trouble than they're worth. It is a very subjective issue.

User avatar
LSAT
Posts: 13412
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Without CPAP For A Week - What Position To Sleep?

Post by LSAT » Sat Oct 11, 2014 8:13 am

If you have any problems with acid/Gerd you are better off sleeping on your left side all the time. Otherwise it doesn't matter.
Anytime I have slept without my CPAP(surgery), I used a recliner.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Without CPAP For A Week - What Position To Sleep?

Post by palerider » Sat Oct 11, 2014 9:21 am

put the cpap down low so there's no dips in the hose between your head and the machine, lower the humidity level some, then if any water condenses in the tube, it'll run back down into the humidifier, and not get blown up into your face.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Drowsy Dancer
Posts: 1271
Joined: Mon Feb 28, 2011 9:35 am
Location: here

Re: Without CPAP For A Week - What Position To Sleep?

Post by Drowsy Dancer » Sat Oct 11, 2014 9:51 am

I don't understand why you are even considering not using your CPAP for a week. Turning off the humidifier would be a far better option.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Mask: Swift™ FX Bella Nasal Pillow CPAP Mask with Headgears
Additional Comments: Software: SleepyHead. Pressure: APAP 9.5 min/11 max, A-Flex x2
How we squander our hours of pain. -- Rilke

User avatar
Wulfman...
Posts: 6689
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Without CPAP For A Week - What Position To Sleep?

Post by Wulfman... » Sat Oct 11, 2014 10:40 am

Drowsy Dancer wrote:I don't understand why you are even considering not using your CPAP for a week. Turning off the humidifier would be a far better option.
RIGHT! ....and +1
The condition you describe indicates there is too much warm humid air cooling in the hose and resulting in "rainout" (water vapor condensed back into liquid by the cooling).


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Without CPAP For A Week - What Position To Sleep?

Post by chunkyfrog » Sat Oct 11, 2014 11:12 am

When you replaced the hose onto the machine, did the usual Climateline recognition message appear?
I'm wondering about the possibility that the cuff may not be turned all the way and the contacts are not touching.
If the problem is the hose, do you have a non-heated hose you can use?--with a quick homemade hose cozy,
you would not need to run without water. Dry running is still better than no blower, though.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Pugsy
Posts: 65399
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Without CPAP For A Week - What Position To Sleep?

Post by Pugsy » Sat Oct 11, 2014 12:31 pm

Assuming the ClimateLine hose is attached properly...go into the Clinical menu and see if the S9 even recognizes it.

If the Climateline is not working and not recognized by the machine it essentially becomes a non heated hose so make sure the setting choice on the S9 is for the smaller diameter hose...and remember that you can still use a hose cozy on the Climateline. No no need to switch to a different hose even if the ClimateLine isn't working.

Condensation in the hose or mask when the Climateline doesn't want to work...treated the same way we have treated it for years before the heated hoses spoiled us.

Options..you can use one or a combination of choices.

1...reduce the humidity out put choice if your nose permits
2...put the machine lower than the bed mattress to allow gravity to help any condensation in the hose to run back into the humidifier if at all possible. Note that depending on the mask used..if there is condensation in the mask sometimes the water sort of pools in the lowest area and sometimes that area is not where the hose enters the mask and thus the gravity thing doesn't always help for condensation in the mask.
3..if nose permits you can try passover mode where water is in the chamber but no heat is added and there's less chance of condensation anywhere because the humidity added is minimal
4...if the bedroom air temp is warmer then there's less chance of the air in the hose and/or mask cooling and condensation forming...so increasing the bed room temp might help. Back before my heated hose I put a little space heater turned on very low on the floor near my machine when I didn't want to heat the whole house by turning up the thermostat. Often just a couple degrees warmer will prevent condensation.
5...the addition of a homemade hose cozy to the hose will help insulate the air in the hose and maybe prevent it from cooling to the point to where it is annoying. If the condensation is in the mask itself insulation any short hose might also help. Doesn't have to be a fancy hose cover. Old socks with the toes cut out so the sock slides over the hose works well and secured by bread ties or scotch tape...anything to keep the socks in place.

Option 6....Not using the cpap machine...isn't an option. There are too many ways to reduce or eliminate condensation in either the mask or hose. People have battled rain out successfully for many years long before we got spoiled with the heated hose. Sometimes takes a bit of detective work to sort it out but it can be done and not using the cpap machine was never an option.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

dangerousdave
Posts: 21
Joined: Sat Jul 07, 2012 6:09 pm

Re: Without CPAP For A Week - What Position To Sleep?

Post by dangerousdave » Mon Oct 13, 2014 5:54 pm

chunkyfrog wrote:When you replaced the hose onto the machine, did the usual Climateline recognition message appear?
Yes I just checked that it said Climate Control Enabled.
englandsf wrote:Have you tried it without any water in the humidifier? Better than no CPAP perhaps?
Drowsy Dancer wrote:I don't understand why you are even considering not using your CPAP for a week. Turning off the humidifier would be a far better option.
Tried it last night and it was not too bad. I'm going with this option.

Thanks everyone for your sleeping sides suggestions, other machine options and to not go without my CPAP. I'm taking it on the plane with me Tuesday and use it without water but with the pan in place I guess.

User avatar
RogerSC
Posts: 1919
Joined: Sat Jul 21, 2012 1:11 pm

Re: Without CPAP For A Week - What Position To Sleep?

Post by RogerSC » Mon Oct 13, 2014 6:43 pm

I'd add that if you had a copy of your sleep study results (do you?), you could see what your AHI was sleeping on your side or your back. For a lot of people, sleeping on one's side is much better. For example, in my sleep study, on my back my AHI was about 42. On my side, it was about 6. That's quite a difference for me. However, it may be different for you. If and when I have to sleep without a cpap machine, I obviously sleep on my side *smile*. Not hard to figure that one out.

None the less, I agree with making an effort to use your cpap, whether that means putting it on something on the floor (like a piece of wood or cardboard) and trying that to see if cuts down on the rainout, or trying a hose cozy, or just playing with your cpap to see if maybe things aren't quite assembled properly, or turning down the humidifier. Or a combination of a couple of these alternatives. Or whatever else you can figure out. Much better for you.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Additional Comments: ResMed AirFit N30 mask

User avatar
mike1953
Posts: 154
Joined: Sun Aug 31, 2014 7:15 pm
Location: Kansas

Re: Without CPAP For A Week - What Position To Sleep?

Post by mike1953 » Mon Oct 13, 2014 7:22 pm

You say you have a S9 AutoSet with the HI5 Humidifier so just disconnect your humidifier and connect a regular hose to the blower unit.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Using a Buckwheat pillow its great

dangerousdave
Posts: 21
Joined: Sat Jul 07, 2012 6:09 pm

Re: Without CPAP For A Week - What Position To Sleep?

Post by dangerousdave » Mon Oct 13, 2014 7:59 pm

mike1953 wrote:You say you have a S9 AutoSet with the HI5 Humidifier so just disconnect your humidifier and connect a regular hose to the blower unit.
Thanks for the suggestion however, I've only had one hose since I got this unit a little over two years ago. I'm just going to run my CPAP without any water in the reservoir.