Dry mouth can be an indication of mouth breathing. Do you have a bed witness you can ask? I wonder if Mouth breathing might show up as Central apneas as there would be no "airway resistance" when your mouth was open. Does anyone else see Central apneas in the reports during periods of mouth leaks using the new Autos?What was your leak rate during the Central Apneas? If it was high I would address your mouth leak first. A full face mask is the surest way, a chinstrap might work, taping works but has it's risks (and followers).Dolpp wrote:Nevermind, I found it all. I was trying to scroll down and didn't realize you had to go page by page. So, pretty much all of my apneas were centrals. In the two weeks i've been using this I am at an AHI average of 15.0/hr. 1.4/hr obstructive and 13.6/hr CA. Recommendations? I am on straight BiPap. 12/8.
Don't overlook humidifier adjustments, too much or not enough can make pap therapy less comfortable and effective. Too dry can make your nose a little stuffy causing a need to mouth breathe. Too much humidity can cause nasal moisture/irritation or even "rain out" in the tubing, Tubing rain out can cause problems with the Auto's ability to recognize and titrate for respiratory events.
If you are not mouth breathing, Central Apneas do seem to be a problem. A lower pressure might be better. You could set your machine to wide open settings and let it titrate you. Most BiPAP autos have three basic settings...
Pressure support (the difference between the high and low pressure) in the case of 12/8 the pressure support is 4. Do you feel comfortable with the transition from inhaling to exhaling on this pressure? If you feel comfortable leave this alone for now. You could try more or less pressure support, adjusted to comfort.
Then you have the higher number called inspiratory max or I max (name varies) this is the highest pressure setting you allow. The main reason for limiting this number is poor tolerance of high pressure or to prevent run away titration due to leaks or to prevent overtitration due to central apneas (the new machines like the System One check for central apneas to prevent this).
Then you have the lower number called the Expiratory minimum or E min (name varies), this is the lowest pressure setting you will allow. The main reason for increasing this setting is to overcome obvious obstructive apneas especially if you can tolerate pressure. Being on too low of a starting pressure tends to waste good sleep time waiting for your auto to treat your obvious apneas. In your case central apneas might be the problem so the lowest pressure might be best.
You could try a pressure support of 4, An Imax of 20 and E min of 4?
This means you would start at 8/4. Try 8/4 first. Are you comfortable or do you feel air starved? If you feel air starved you can try a higher E min or a different pressure support setting. Your System One Auto Bipap should titrate up the pressure in response to hypopneas, obstructive apneas and snoring/flow limitation as needed while ignoring central apneas (in theory). It will maintain the same pressure support while adjusting between a minimum of 8/4 to maxiumum of 20/16.
Here are some of the risks....you could have severe apneas on these lower pressures...The auto treats your respiratory events but you have poor oxygen due to an associated health problem (COPD, Heart Disease), It "over titratrates" and causes severe central apneas or over ventilation. You have to take responsibility for these potential risks.
In the sleep lab we usually increase the expiratory pressure in response to obstructive apneas, we increase the inspiratory pressure in response to hypopneas. So if you are having primarily obstructives you could increase the E min, if you have primarily Hypopneas on your reports you could increase the pressure support. We try in have a minimum of 3 for pressure support however we try not to make the difference between the I and E (pressure support) too great as this tends to create a kind of "jolt" of air in transition between inhaling and exhaling, Bi-flex or A-flex on the System One can help with this.
Good Luck
Any one else have any ideas for Dolp's first at home self BiPAP tiration?