can too high of pressure prevent treatment?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Pugsy
Posts: 65121
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: can too high of pressure prevent treatment?

Post by Pugsy » Sun Jul 22, 2018 4:01 am

Ariseal wrote:
Sat Jul 21, 2018 10:36 pm
The centrals aren’t caused by raising the pressure to high. They are caused by raising the pressure to fast coupled with too high.
And then there are the people who get too many centrals with the least little pressure...like 5 or 6 cm. You start them out at 5 cm and they get too many centrals right from the get and go. It doesn't necessarily have to be higher or too fast...it can be simply any pressure.

But then most people won't have too many centrals even at 20 cm. It's not nearly as common as people tend to think.

The OP in this thread...long story there but there is no money for an in lab sleep study or even a home sleep study (of the Type II variety)....out of work and no insurance.
His centrals are primarily SWJ centrals IMHO..post arousal..he is awake and not sleeping well and his not sleeping well predates cpap use and is why he is going down the DIY road.
The machine was given to him..we are making the best we can of a less than ideal situation here in an effort to help him feel better by sleeping better.

As to why he isn't sleeping well...we don't know. The cpap/apap trial was a shot in the dark to see if maybe it was related to SDB of some sort that the machine might help with. Worth trying. It's not classic OSA I don't think but it might be a UARS kind of thing. As we both know it's really hard to diagnosis UARS without a sleep study done with a Pes device.

Zack doesn't have a lot of patience and he expects the machine to provide data that is clear cut and help him figure out why his sleep is crappy but unfortunately the machine's data doesn't always do that. Can't say as I blame him as I would very likely be the same way if I was in his shoes. :lol:
He needs to do a crash course in figuring out of the apnea event flags he sees are real or not. I don't know if he has ever done this but I know I have suggested it several times. And I know how difficult that is...I can't always tell for sure myself by flow rate examination but it's all we have from the machine to maybe help us figure out asleep vs awake breathing.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
zackds
Posts: 130
Joined: Fri Mar 30, 2018 10:13 am

Re: can too high of pressure prevent treatment?

Post by zackds » Sun Jul 22, 2018 11:03 am

im trying to be diligent about turning the machine off as soon as i wake up
i also want to get an oximeter like you suggested pugsy
i just hope im on the right track at this point
if not, i want to look else where obviously

i have felt better for about 5 days since being on cpap
most days i still dont sleep well
as ive stated before, im not sure if its the combination of cpap
and something else or what
i know i have most symptoms of some type of breathing difficulty
when sleeping
morning headache, bruxism, heart flutter upon awakening,
nocturia, waking very often in early morning hours, feeling like death when morning came
and all throughout the day
and both of my parents have diagnosed sleep apnea
those are all i can think of now

some of these have subsided since being on cpap
like the headache, bruxism
but the heart flutter is still there sometimes
i dont feel like death when it comes time to wake up
but i still feel very much less than optimal upon awakening
and throughout most days

based on the trend that im seeing when i look at all of my sleepyhead data
i see waxing and waning
and in that waxing and waning i see flat lines that are usually just under 10 seconds long
i cant tell you for sure if im awake or not during this time
but ive been trying to cut my machine off as soon as i wake up
sometimes i see what look like violent lines at the end of these waxing and waning spells
i dont know if its normal to stop breathing for 9 seconds even when awake though
also, pressure changes dont seem to do anything for the waxing/waning

_________________
Machine: DreamStation Auto CPAP Machine
Mask: DreamWear Full Face CPAP Mask with Headgear (Small and Medium Frame Included)
Additional Comments: machine is phillips respironics system one 50

User avatar
Pugsy
Posts: 65121
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: can too high of pressure prevent treatment?

Post by Pugsy » Sun Jul 22, 2018 11:40 am

zackds wrote:
Sun Jul 22, 2018 11:03 am
i dont know if its normal to stop breathing for 9 seconds even when awake though
It's normal to do a lot of weird breathing stuff while awake and we simply don't realize we are doing it.
For our sleep problems to be related to breathing and sleep...we gotta be asleep and if we are awake any of that weird breathing stuff simply doesn't count.

Now figuring out the why for being awake...not always so easy to do. There's lot of stuff that can cause us to wake up and it doesn't necessarily have to be related to our breathing. The machine can only fix problems related to the breathing problems though. It does nothing for fixing a problem causing arousals that aren't related to breathing issues.

I will repeat what I have said before...pick a pressure number that seems to work and just use the machine for a period of time with no willy nilly changing of anything. Pick a pressure that you used one of those nights where you didn't have to pee during the night and you felt decent during the day with a minimal number of wake ups during the night.

Then be prepare to use it for a month...keep a log as to how you feel or sleep and the pee breaks.
Accept the limitations of the data from your machine...it is what it is. Learn to recognize arousal/awake/post arousal breathing event flags and be prepared to mentally remove them from any evaluation except the fact that the arousal breathing means poor sleep quality for some reason.

I don't know that the gentle waxing and waning that you see means much one way or the other. Even if it did I don't know what you can do about it. The only machine that will smooth that out is one of the ASV machines and it will make the flow rate nice and pretty but no guarantee that it will mean you will sleep any better and you can't afford one of those machines anyway.

If this UARS you have going on...I told you a long time ago that there were no quick fixes and you need months of use to figure out if it helps or not because the improvements are so very gradual and small and take a long time to realize that things are getting any better.

And back to your original question in this thread...can you maybe use a pressure that is making things worse?
Maybe but really only if the more pressure causes comfort issues which in turn affect sleep quality. I don't think more pressure causes centrals in your situation. I think the bulk of your centrals are related to SWJ. An occasional real central is normal and not a worry. So the higher pressures might cause comfort issues (aerophagia or just high enough to not be comfortable exhaling) but otherwise more pressure doesn't really hurt anything.

You have some crappy sleep issues...now why your sleep is crappy we can't really know for sure. About all we can do is try stuff and see if we can maybe get lucky and figure out a cause and a solution.
But at some point you have to quit dial a wingin and pick something that can work and give it an extended trial.

It's hard...I know that we want to try to adjust something based on last night's results but we simply don't sleep the same each night. Get close and then give it time.

Some time back when I got a bilevel machine I had to come up with some sort of settings. My nightly results were sort of all over the place...some really good and some not so good and some even down right horrible.
What I did was pick some settings that could give me decent results and then I prepared myself to use those settings no matter what the nightly report showed and not go changing things willy nilly. My AHI was running in the 3 ish category...some good nights sleep and some not so good but remember I have some sleep maintenance insomnia issues related to pain and/or the medication I take for the pain.

At first the results were all over the place...mostly not so good but with an occasional good night but over time (I did this experiment for 6 weeks where I forced myself to not change anything because I knew that the settings could work) everything improved without my changing anything. Hours of sleep improved...awakenings became less numerous...AHI reduced and how I felt in general during the day improved.
My AHI was cut in half at the end of the 6 weeks. My hours of sleep increased to somewhere around 45 minutes to an hour more sleep and that extra time asleep has always been critical in how I felt during the day. Slowly the "good" nights became the majority of what I saw instead of in the beginning the "good" nights were in the minority.
I am not a very patient person. I am one of those people that wants immediate marked relief...the hardest thing I have ever done was not change anything those 6 weeks but it was one of the best experiments I ever have done. There is some truth to the "give it time" thing because the body and the brain needs to adjust to the settings.

Your AHI is never going to be a good measuring tool because your AHI is primarily SWJ and all it really tells us is that when we have more SWJ we have more crappy sleep. You are going to have to rely on how you feel and that's really hard to do because we want to look at numbers or the flow rate and see a problem so we can fix the problem. Unfortunately the data from the machine can't always give us what we need in that regard.

I would also like to see you try a fixed pressure or a really tight apap range...just to eliminate any pressure changes as being a factor in the arousals. Since your machine won't flag FLs in fixed cpap mode...I suggest keeping it in apap mode and make the minimum equal the maximum. It will work like cpap mode but you still have FL flagging.

That's what I would have you do if you were a family member of mine. It's not perfect but given what we have to work with and your limited budget...about all we can do.

The fact that the nocturia has improved at times...that alone is worth continuing the experiment.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: can too high of pressure prevent treatment?

Post by palerider » Sun Jul 22, 2018 12:08 pm

zackds wrote:
Sun Jul 22, 2018 11:03 am
i dont know if its normal to stop breathing for 9 seconds even when awake though
yes, it is, wake breathing is very irregular.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zackds
Posts: 130
Joined: Fri Mar 30, 2018 10:13 am

Re: can too high of pressure prevent treatment?

Post by zackds » Mon Jul 23, 2018 12:36 pm

i woke up last night on my back with a headache
first time thats happened in a while
last time that happened was when i fell asleep without the machine
a while ago
i dont understand why, if im waking up from events
i wouldnt have a headache, but its still saturation enough to wake me up?

i can try the fixed pressure like you say pugsy
im also getting an oximeter soon
because i think that can pretty much tell me if im having events or not


also, i have been trying to find trends in my data like you mentioned
i do notice that i have less reras at higher pressures
with more hypopneas/obstructives
and the machine changes pressures instead of just pulsing
like it does at lower pressures
although, im not sure why reras at lower pressures
would turn into hypopneas/obstructions at higher pressures
i dont know if im just awake more with higher pressures or what
i also notice i have way more vibratory snores
not sure if thats just from other vibrations/parts moving/air or
if its actual snoring
but thats the trend ive noticed when looking at my data

at the same time, i look back on the first couple of weeks when i
started using the machine and the pressure increases there also
the ahi is higher, up to 4 on one night, with more centrals
i was probably awake then though as you know

i may try to sleep without the machine for a night or two
before trying the fixed pressure experiment
last time i did that i woke up with a headache though
if that happens ill just put the mask back on

i almost wonder if just the lowest pressure i have used has fixed
any apnea problem i may have had and maybe im just dehydrated or
something on top of it and its waking me up rather than events anymore

_________________
Machine: DreamStation Auto CPAP Machine
Mask: DreamWear Full Face CPAP Mask with Headgear (Small and Medium Frame Included)
Additional Comments: machine is phillips respironics system one 50

User avatar
zackds
Posts: 130
Joined: Fri Mar 30, 2018 10:13 am

Re: can too high of pressure prevent treatment?

Post by zackds » Tue Jul 24, 2018 7:47 am

last night
not sure what the large chunk of breathing not detected is from
i dont remember getting up around then
its like 5 minutes long
when i was waking up in the early morning hours
i had a slight headache

and the second ss is
something that seems to happen a lot
the first little piece of activity
with the 'fl' flag....
does that just look like a spontaneous wake up?
that is a common trend i see in my sleep reports

_________________
Machine: DreamStation Auto CPAP Machine
Mask: DreamWear Full Face CPAP Mask with Headgear (Small and Medium Frame Included)
Additional Comments: machine is phillips respironics system one 50
Attachments
screenshot-20180724-094221.png
screenshot-20180724-094221.png (147.92 KiB) Viewed 6919 times
screenshot-20180724-093605.png
screenshot-20180724-093605.png (147.45 KiB) Viewed 6919 times

User avatar
Pugsy
Posts: 65121
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: can too high of pressure prevent treatment?

Post by Pugsy » Tue Jul 24, 2018 8:24 am

Here's the deal with the arousal breathing...with the machine provided data we have zero way to know for sure if the arousal was spontaneous or related to a breathing issue. The subtle differences that might be a factor with a breathing related arousal won't necessarily be evident on the flow rate chart.
This is where the in lab sleep study is needed...first to verify sleep status and secondly to evaluate the flow rate prior to the whatever caused arousal.
We can make some educated guesses given the data that we have but it's not 100%.
Heck, I have seen hyponea flags that I sit here scratching my head wondering where the restricted flow is.

We sometimes need or want a level of data from these machines that simply is not available. We want something the machine was never designed to provide.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
zackds
Posts: 130
Joined: Fri Mar 30, 2018 10:13 am

Re: can too high of pressure prevent treatment?

Post by zackds » Tue Jul 24, 2018 9:08 am

yea, despite my ahi being higher than it has been in a while
i feel better today than i have in a bit
i made sure i was hydrated yesterday before going to bed
i think thats been part of my trouble for sure

_________________
Machine: DreamStation Auto CPAP Machine
Mask: DreamWear Full Face CPAP Mask with Headgear (Small and Medium Frame Included)
Additional Comments: machine is phillips respironics system one 50

User avatar
zackds
Posts: 130
Joined: Fri Mar 30, 2018 10:13 am

Re: can too high of pressure prevent treatment?

Post by zackds » Tue Jul 24, 2018 3:33 pm

also, what i meant pretty much was
is something like that
could that not be an arousal at all?
spikes that are like 3 seconds in length
with nothing before or after
does that happen to everyone?

_________________
Machine: DreamStation Auto CPAP Machine
Mask: DreamWear Full Face CPAP Mask with Headgear (Small and Medium Frame Included)
Additional Comments: machine is phillips respironics system one 50

User avatar
Pugsy
Posts: 65121
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: can too high of pressure prevent treatment?

Post by Pugsy » Tue Jul 24, 2018 4:30 pm

It could be a lot of things.
Maybe there was a sudden noise of some sort outside...
And yes...it could also happen to anyone and everyone.

Arousals...you may or may not remember them and they may or may not be related to breathing at all.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
djams
Posts: 927
Joined: Mon Jun 18, 2018 4:37 pm
Location: DFW area

Re: can too high of pressure prevent treatment?

Post by djams » Tue Jul 24, 2018 5:08 pm

zackds wrote:
Sun Jul 22, 2018 11:03 am
i dont know if its normal to stop breathing for 9 seconds even when awake though
Pugsy wrote:
Sun Jul 22, 2018 11:40 am
It's normal to do a lot of weird breathing stuff while awake and we simply don't realize we are doing it.
palerider wrote:
Sun Jul 22, 2018 12:08 pm
yes, it is, wake breathing is very irregular.
When I was in the first couple of weeks and getting adjusted, I'd tell myself "just breathe normally stupid". Quickly found out that I have no idea what normal breathing is. Just having this thought would put me into a very poor breathing pattern. Even causing light-headedness. This very thing happened again when I tried to jump from 8cm-12cm min pressure. Just isn't something we pay any attention to.

Three months in, respiration-related charts before I fall asleep are vastly different from when I'm asleep. No idea if my awake breathing with the mask on is anywhere near my "normal" awake breathing. I highly doubt it.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Backup machine: AirSense 10 AutoSet with all the fixins