resmed sleepyhead or ?? somebody is wrong
-
Guest
Re: resmed sleepyhead or ?? somebody is wrong
that's quite easy: max pressure is the maximum of your measured mask-pressure. that is not the same as the set max ipap and epap.
as I already pointed out a couple of times: there are some instances where this normal and in some and can be a faulty machine - the difference can only be made by actually looking at the data.
as I already pointed out a couple of times: there are some instances where this normal and in some and can be a faulty machine - the difference can only be made by actually looking at the data.
-
Guest
Re: resmed sleepyhead or ?? somebody is wrong
where exactly is now the IPAP-Max coming from?xxyzx wrote:sleepyhead just reported that my max pressure for the day i ran the data was 34
does not make any sense what your are saying.
Pics or get on^^ - there is no other way to be truly talking about the very same!
-
Guest
Re: resmed sleepyhead or ?? somebody is wrong
there simply is nothing more to add or to explain.Guest wrote:that's quite easy: max pressure is the maximum of your measured mask-pressure. that is not the same as the set max ipap and epap.
as I already pointed out a couple of times: there are some instances where this normal and in some and can be a faulty machine - the difference can only be made by actually looking at the data.
you keep confusing 2 very different things constantly - there simply is no way knowing what you want, what you see and what you are talking about.
REALLY - Pics or it didn't happen!
Re: resmed sleepyhead or ?? somebody is wrong
xxyzx wrote:
I am not confusing anything
at the top of sleepyhead it says IPmax was 34
in plain english
pictures wont change the text and
not worth the effort to convince you that i can read
Based on many of your posts, it would be very hard to convince me./
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
| Additional Comments: Back up is a new AS10. |
-
Guest
Re: resmed sleepyhead or ?? somebody is wrong
what is it now?
pressure max (as you said in the beginning)
or IPAP max (as you say now)
????
how can a machine which - as you said - can only achieve 25cm max out to 34? like in: how?
the ipap recorded is not something that is live - the mask-pressure however is - that is something that is measured over time and set by the machine.
what you are saying simply does not make any sense at all.
what would make sense: max-pressure is 34.
that could be a problem or not - depends!
there simply is no way you describe the curves - some (random) value has no meaning at all! ... there also is no way we can tell YOU how to decide if it is a problem or not (way too much to take in for YOU!)
so post some charts or go troll someone else!
pressure max (as you said in the beginning)
or IPAP max (as you say now)
????
how can a machine which - as you said - can only achieve 25cm max out to 34? like in: how?
the ipap recorded is not something that is live - the mask-pressure however is - that is something that is measured over time and set by the machine.
what you are saying simply does not make any sense at all.
what would make sense: max-pressure is 34.
that could be a problem or not - depends!
there simply is no way you describe the curves - some (random) value has no meaning at all! ... there also is no way we can tell YOU how to decide if it is a problem or not (way too much to take in for YOU!)
so post some charts or go troll someone else!
-
user232
Re: resmed sleepyhead or ?? somebody is wrong
I was a able to toy with the data a little bit, while I had my 3 day rental.
If I was concerned about the sleepyhead's interpretation of the pressure, I would try to directly read BRP(dot)edf file of time series
off the machine's data card. If you look under S9 Data fromat SleepyHead Wiki, it shows
BRP High resolution graph data gives "Mask Pres" High-res Mask Pressure cmH2O.
There are a few EDF file readers out there, and you could at least try to correlate those raw numbers to what you expect, before
any sleepyhead processing.
For example, looking under my resmed folder I found
9_18_17\SDHC\DATALOG\20160617143148BRP(dot)edf
If I was concerned about the sleepyhead's interpretation of the pressure, I would try to directly read BRP(dot)edf file of time series
off the machine's data card. If you look under S9 Data fromat SleepyHead Wiki, it shows
BRP High resolution graph data gives "Mask Pres" High-res Mask Pressure cmH2O.
There are a few EDF file readers out there, and you could at least try to correlate those raw numbers to what you expect, before
any sleepyhead processing.
For example, looking under my resmed folder I found
9_18_17\SDHC\DATALOG\20160617143148BRP(dot)edf
-
user232
Re: resmed sleepyhead or ?? somebody is wrong
Maybe someone can clean up and post the image link for me (I don't have permission).
http://i67(dot)tinypic(dot)com/15q6y6q(dot)jpg
But what you can see is a comparison (admittedly formatting is a bit sloppy as I slapped the two images together with mspaint).
On the left, is the sleepyhead data.
On the right is the raw resmed card BRP(dot)edf data for the same evening, BEFORE sleepyhead even sees it.
Notice the measurements and scales are very close, telling me sleepyhead does not do much data transformation, other than pass the data directly.
This can help isolate your issue.
http://i67(dot)tinypic(dot)com/15q6y6q(dot)jpg
But what you can see is a comparison (admittedly formatting is a bit sloppy as I slapped the two images together with mspaint).
On the left, is the sleepyhead data.
On the right is the raw resmed card BRP(dot)edf data for the same evening, BEFORE sleepyhead even sees it.
Notice the measurements and scales are very close, telling me sleepyhead does not do much data transformation, other than pass the data directly.
This can help isolate your issue.
-
user232
Re: resmed sleepyhead or ?? somebody is wrong
You can also read and view raw IPAP and EPAP, if they exist, using PLD(dot)edf files.
One possibility, is that most of the manuals show that there is a "maximum single fault steady state pressure."
In the S9 manual, is is on the order of 30 to 40 (depending on duration of event).
This tells me that the machine is capable of putting out higher than the intended value range(s) set by the user (and your measurement of 34 is conceivable in this range).
However, being that is a "fault" it was not necessarily intended. It also does not give the exact reason for that breech, it is up to
someone with the data to troubleshoot it.
Again, the way to isolate it is to confirm by looking at raw time series files, before sleepyhead. There are several free edf browsers, and as others mentioned, there is a more specific resmed reader software.
One possibility, is that most of the manuals show that there is a "maximum single fault steady state pressure."
In the S9 manual, is is on the order of 30 to 40 (depending on duration of event).
This tells me that the machine is capable of putting out higher than the intended value range(s) set by the user (and your measurement of 34 is conceivable in this range).
However, being that is a "fault" it was not necessarily intended. It also does not give the exact reason for that breech, it is up to
someone with the data to troubleshoot it.
Again, the way to isolate it is to confirm by looking at raw time series files, before sleepyhead. There are several free edf browsers, and as others mentioned, there is a more specific resmed reader software.
-
user232
Re: resmed sleepyhead or ?? somebody is wrong
I don't have enough experience to really say for sure, but it I can say that my sleepyhead and resmed files did not even report ipap data, only epap.
It could be that my machine has settings for epap = ipap or something (again, don't know enough to elaborate yet).
And in most cases I can sample (including 2 years of older user data), I can always see both the printed (statistics) epap range, and charts both show epap,
being below(dynamic Mask) Pressure. If the your chart boundaries are much less than the max reported value on the statisics range entry, then I would also think something is amiss.
But again, you can look at raw data if you want to take sleepyhead out of the equation (or isolate the calculation). Beyond that requires looking inside the software itself, and as I pointed out,
the raw pressure data is no different than sleepyhead's reported data.
It could be that my machine has settings for epap = ipap or something (again, don't know enough to elaborate yet).
And in most cases I can sample (including 2 years of older user data), I can always see both the printed (statistics) epap range, and charts both show epap,
being below(dynamic Mask) Pressure. If the your chart boundaries are much less than the max reported value on the statisics range entry, then I would also think something is amiss.
But again, you can look at raw data if you want to take sleepyhead out of the equation (or isolate the calculation). Beyond that requires looking inside the software itself, and as I pointed out,
the raw pressure data is no different than sleepyhead's reported data.
Re: resmed sleepyhead or ?? somebody is wrong
With EPR, or on a bilevel machine, both epap and ipap are reported.user232 wrote:I don't have enough experience to really say for sure, but it I can say that my sleepyhead and resmed files did not even report ipap data, only epap.
It could be that my machine has settings for epap = ipap or something (again, don't know enough to elaborate yet).
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: resmed sleepyhead or ?? somebody is wrong
I just reviewed one months worth of data in SleepyHead from my Resmed Adapt which is of course the ASV model.
Not once did I see an IPAP max over the 25 cm the machine is supposed to have as a maximum. I saw some 22, 23 and 24 but never anything over 25.
I have no idea if your 34 max IPAP is a reflection of SH just being buggy or if your machine is weirding out.
If you don't see IPAP peaks over 25 on the pressure graphs with either ResScan or SleepyHead...then we don't know where SH got the 34.
If ResScan doesn't show the same IPAP max number...then obviously SH bug is likely the culprit.
I don't remember if ResScan even shows IPAP max...and no time to go fire it up and check it out.
Your DME supplied you with a used machine instead of a new zero hour machine.
Makes a person wonder why it was returned.
Not once did I see an IPAP max over the 25 cm the machine is supposed to have as a maximum. I saw some 22, 23 and 24 but never anything over 25.
I have no idea if your 34 max IPAP is a reflection of SH just being buggy or if your machine is weirding out.
If you don't see IPAP peaks over 25 on the pressure graphs with either ResScan or SleepyHead...then we don't know where SH got the 34.
If ResScan doesn't show the same IPAP max number...then obviously SH bug is likely the culprit.
I don't remember if ResScan even shows IPAP max...and no time to go fire it up and check it out.
Your DME supplied you with a used machine instead of a new zero hour machine.
Makes a person wonder why it was returned.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.