Hello all,

I addressed the issue of my non data capable machine with the PA at the sleep clinic. She says that there is no medical necessity for a data capable machine that costs "thousands of dollars."

I raised the point that the MaskFit, AHI stats are valuable information for a CPAP user to have, especially for fine tuning on a day to day basis, but she says that these values tend to not be accurate and she found them to not be useful in her years of experience. Therefore there is no medical necessity to change out the CPAP machine. I could get a replacement if either I or my insurance company were able to pay for it just based on my wishes.That involves returning my current used machine, then pay the difference.

I'm not sure at this point. I would feel much more strongly if I could get to the point where I keep my mask on the full night and have several hours worth of data per night to review.

On a positive note she thinks that I might be cured of needing a machine by simply losing weight. It's true that I have a lot of weight to lose; by BMI is 32. Unfortunately, my energy level as a result of sleep apnea causes me to look to food for fighting fatigue, and the lack of energy makes it all the more difficult to turn down food. Others have observed this too. Inversely, some people lost significant weight as a "side effect" of CPAP therapy.
Last edited by davep700 on Wed Dec 16, 2015 9:35 pm, edited 2 times in total.

User avatar
LSAT
Posts: 13353
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: No Medical Necessity to replace a Brick

Post by LSAT » Wed Dec 16, 2015 9:31 pm

davep700 wrote:Hello all,

I addressed the issue of my non data capable machine with the PA at the sleep clinic. She says that there is no medical necessity for a data capable machine that costs "thousands of dollars."

I raised the point that the MaskFit, AHI stats are valuable information for a CPAP user to have, especially for fine tuning on a day to day basis, but she says that these values tend to not be accurate and she found them to not be useful in her years of experience. Therefore there is no medical necessity to change out the CPAP machine. I could get a replacement if either I or my insurance company were able to pay for it just based on my wishes.That involves returning my current used machine, then pay the difference.

I'm not sure at this point. I would feel much more strongly if I could get to the point where I keep my mask on the full night and have several hours worth of data per night to review.
She is giving you a line of BS...the difference in cost between a brick and a data capable machine is a few dollars. In fact...the insurance companies consider them the same and pay the DME the same amount for each. The reason for the DME to give you a brick is that the make a few extra dollars by giving you the brick. See if you can get your doctor to write a prescription for a data capable machine. Then the DME must comply with the prescription.

davep700
Posts: 156
Joined: Tue Oct 06, 2015 9:02 am

Re: No Medical Necessity to replace a Brick

Post by davep700 » Wed Dec 16, 2015 9:39 pm

Thanks. Seems that the sleep clinic, DME and insurance company are of the same opinion that a data capable and non data capable machine are equivalent. Maybe manufacturers should stop making two classes of CPAP machines.

I guess I should look to my PCP for assistance with a prescription.

User avatar
Pugsy
Posts: 65127
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: No Medical Necessity to replace a Brick

Post by Pugsy » Wed Dec 16, 2015 9:47 pm

What would your insurance be doing in terms of helping with costs if you didn't already have a machine?
What is your deductible?
What would be your co pay IF you could get a RX for a full data machine and run it through insurance.
Do the math...it might be less costly to just buy out of pocket another blower unit that you could use with your existing humidifier.

Full data machines don't "cost thousands of dollars" even if brand new. The PA at the sleep clinic...well let's just say not fully informed as to actual costs or what these new machines can do....or he/she has been out in the rain too long (as in all wet).

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Pugsy
Posts: 65127
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: No Medical Necessity to replace a Brick

Post by Pugsy » Wed Dec 16, 2015 9:54 pm

viewtopic.php?f=1&t=109450&view=unread#unread

Check out this thread above...that machine is a full data machine and while primarily a fixed cpap machine like you are currently using it does have some limited apap mode capabilities.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Julie
Posts: 20052
Joined: Tue Feb 28, 2006 12:58 pm

Re: No Medical Necessity to replace a Brick

Post by Julie » Wed Dec 16, 2015 10:11 pm

Her ignorance showed up the minute she told you to just lose weight.

Sonnyboy
Posts: 637
Joined: Sun Nov 23, 2014 8:56 am

Re: No Medical Necessity to replace a Brick

Post by Sonnyboy » Wed Dec 16, 2015 10:18 pm

I am sorry they are giving you such a hard time.

I agree with Pugsy, LSAT and Julie.

Without data, you and your doctor can not assess your treatment and when you think about it this is the medical necessity justifying data capable machines.

Talk to your PCP. This is between you and your doctor and if your doctor is not in agreement with you consider finding a better doctor.

_________________
Mask

Janknitz
Posts: 8512
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: No Medical Necessity to replace a Brick

Post by Janknitz » Wed Dec 16, 2015 11:41 pm

Even if you get an RX from your doctor for a data capable machine it's going to be an uphill battle with the DME who now has to accept a used brick and give you a brand new data capable machine withou making a single penny more. (BTW, the RETAIL difference between a brick and a data capable machine is about $35. You do the math on the wholesale difference. The DME foisted tge brick on you to make an extra $10 or $12.

Anyway, you're going to have to grow a pair and get fully educated to have a snowball's chance of getting your DME to replace the brick, RX or not. So if you're not the assertive type, consider other options like buying your own blower with data to go with the humidifier you already have. Secondwind.com is a place to look.
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

User avatar
kaiasgram
Posts: 3569
Joined: Sat Jun 02, 2012 2:08 pm
Location: Northern California

Re: No Medical Necessity to replace a Brick

Post by kaiasgram » Wed Dec 16, 2015 11:42 pm

davep700 wrote:I raised the point that the MaskFit, AHI stats are valuable information for a CPAP user to have, especially for fine tuning on a day to day basis...
As a matter of strategy -- From the perspective of so many providers, the CPAP user doesn't need any information about what the machine is doing, wouldn't understand it anyway, and certainly has no business fine-tuning on a day to day basis. Remember, we're not even supposed to know how to access those settings. So I would not make me, 'the CPAP user' the focus of my argument if I was trying to convince the PA of the need for a data capable machine. Instead I would point out that the doctor (or whoever the provider is) has no way to evaluate the treatment he or she has prescribed. That's it. There is no way the PA can refute that argument, because it's true. The objective measure of successful CPAP treatment is AHI < 5. AHI is not available on a brick.

Your PA sounds like she won't budge anyway, but for anyone in a similar bind, it's helpful to think about how to approach this strategically. Appeal to the provider in terms of how their hands will be tied when it comes to evaluating your treatment and/or troubleshooting problems, not how your hands will be tied.

_________________
Machine: AirSense 10 AutoSet with Heated Humidifer + Aifit N30i Nasal Mask Bundle
Additional Comments: SleepyHead-now-OSCAR software on Mac OSX Ventura

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: No Medical Necessity to replace a Brick

Post by archangle » Thu Dec 17, 2015 12:48 am

Unfortunately, you're dealing with the medical mafia here.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

Greg Riddle
Posts: 409
Joined: Wed Aug 13, 2014 4:48 pm
Contact:

Re: No Medical Necessity to replace a Brick

Post by Greg Riddle » Thu Dec 17, 2015 8:04 am

Julie wrote:Her ignorance showed up the minute she told you to just lose weight.
My wife just lost 115 pounds abd I've had to raise her issue to keep get ahi below 5. She is going to sleep dr tomorrow

davep700
Posts: 156
Joined: Tue Oct 06, 2015 9:02 am

Re: No Medical Necessity to replace a Brick

Post by davep700 » Thu Dec 17, 2015 11:06 am

kaiasgram wrote:
davep700 wrote:I raised the point that the MaskFit, AHI stats are valuable information for a CPAP user to have, especially for fine tuning on a day to day basis...
As a matter of strategy -- From the perspective of so many providers, the CPAP user doesn't need any information about what the machine is doing, wouldn't understand it anyway, and certainly has no business fine-tuning on a day to day basis. Remember, we're not even supposed to know how to access those settings. So I would not make me, 'the CPAP user' the focus of my argument if I was trying to convince the PA of the need for a data capable machine. Instead I would point out that the doctor (or whoever the provider is) has no way to evaluate the treatment he or she has prescribed. That's it. There is no way the PA can refute that argument, because it's true. The objective measure of successful CPAP treatment is AHI < 5. AHI is not available on a brick.

Your PA sounds like she won't budge anyway, but for anyone in a similar bind, it's helpful to think about how to approach this strategically. Appeal to the provider in terms of how their hands will be tied when it comes to evaluating your treatment and/or troubleshooting problems, not how your hands will be tied.
PA already said that she wouldn't use the data anyway for evaluating treatment because it's inaccurate and unreliable. Apparently many people getting treatment at this clinic are receiving data-less machines and being evaluated for their treatment based on, I suppose, their hours of use and how they feel.

If this AHI info is truly medically necessary then why do manufacturers continue to make crippled machines over a $35 difference? Has the data not been embraced and adopted by the medical establishment due to cost, corruption or due to given or other reasons?

User avatar
49er
Posts: 5624
Joined: Mon Jan 16, 2012 8:18 am

Re: No Medical Necessity to replace a Brick

Post by 49er » Thu Dec 17, 2015 11:13 am

davep700 wrote:
kaiasgram wrote:
davep700 wrote:I raised the point that the MaskFit, AHI stats are valuable information for a CPAP user to have, especially for fine tuning on a day to day basis...
As a matter of strategy -- From the perspective of so many providers, the CPAP user doesn't need any information about what the machine is doing, wouldn't understand it anyway, and certainly has no business fine-tuning on a day to day basis. Remember, we're not even supposed to know how to access those settings. So I would not make me, 'the CPAP user' the focus of my argument if I was trying to convince the PA of the need for a data capable machine. Instead I would point out that the doctor (or whoever the provider is) has no way to evaluate the treatment he or she has prescribed. That's it. There is no way the PA can refute that argument, because it's true. The objective measure of successful CPAP treatment is AHI < 5. AHI is not available on a brick.

Your PA sounds like she won't budge anyway, but for anyone in a similar bind, it's helpful to think about how to approach this strategically. Appeal to the provider in terms of how their hands will be tied when it comes to evaluating your treatment and/or troubleshooting problems, not how your hands will be tied.
PA already said that she wouldn't use the data anyway for evaluating treatment because it's inaccurate and unreliable. Apparently many people getting treatment at this clinic are receiving data-less machines and being evaluated for their treatment based on, I suppose, their hours of use and how they feel.

If this AHI info is truly medically necessary then why do manufacturers continue to make crippled machines over a $35 difference? Has the data not been embraced and adopted by the medical establishment due to cost, corruption or due to given or other reasons?
Dave,

Ask the PA what the evidence is for her position? Not sure how to ask that question in a non confrontational way but maybe someone has a suggestion. Her position is also absurd because that would be like my saying to my doctor that my blood pressure is fine due to my feeling great even though I didn't measure it.

I have seen several sleep doctors and none of them had the attitude that I shouldn't have a data capable machine. Now a few of them may have been threatened that I knew how to change the pressure but that is another issue.

49er

User avatar
Pugsy
Posts: 65127
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: No Medical Necessity to replace a Brick

Post by Pugsy » Thu Dec 17, 2015 11:23 am

davep700 wrote:If this AHI info is truly medically necessary then why do manufacturers continue to make crippled machines over a $35 difference? Has the data not been embraced and adopted by the medical establishment due to cost, corruption or due to given or other reasons?
Because DMEs and Sleep Clinics continue to buy them.

If you owned a sleep clinic...which would you buy...when the bulk of your income is from sleep studies and the sale of a machine?
If they sell you a brick and you aren't doing well then they get to say "let's do another sleep study to see if we can figure out why you aren't doing well"...
or they could sell you a full data machine and look and find out the problem is maybe a leak issue and they don't get to do another sleep study.
It's all about the dollars.

While the difference in cost at a wholesale level is all that great...and insurance pays the same no matter whether the machine is a brick or full data...multiply that slight difference by 1,000 machines and there is a substantial potential income difference. It's all about the dollars and not what truly might benefit the patient.

CPAP machine manufacturers don't count us the end user as their customer...their customer, that they cater to, is the DME and sleep clinics.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

davep700
Posts: 156
Joined: Tue Oct 06, 2015 9:02 am

Re: No Medical Necessity to replace a Brick

Post by davep700 » Thu Dec 17, 2015 11:28 am

49er, I'm not sure yet what evidence for her position is that the data tends to be inaccurate.

On this forum I heard that machines might miscategorize apneas, hypopneas and centrals, but that's about it.

I could certainly use your argument if and only if I knew the measurement data I can get is reliable.
I might look for some studies or articles on the use of the data in CPAP therapy (??).

Fortunately, I am going to my visit my PCP tomorrow. I assume I need to inqurie about a prescription worded as, "data capable CPAP machine @ pressure 12cmh2o".