Post by Jerry69 » Sat Dec 24, 2005 7:55 am

I was titrated at 6 cm. I increased this to 8 cm and saw some reduction in AHI's. Last night, I set my Remstar Pro II with C-flex on 4 cm, the lowest it will go. My AHI was 1.5, one of the lowest I have recorded?

The following charts are from Derek's My Encore. Thanks, again, Derek.
Image

I will use 4 cm for a few more nights and track the above chart. The fact that average AHI's were higher at 6 cm may be misleading as I had some bad/high nights upon starting treatment (at 6 cm). Based on one night at 4 cm, I'd say there is little correlation between AHI and the pressure range I'm using.

The SI shows a strong correlation with pressure, however.

Image

I asked the wife if she heard me snoring last night. She said she didn't? I think I was aware of snoring at least one awakening.

I welcome your thoughts on these results. If I continue to get low AHI's at 4 cm, maybe I will go to 10 cm. If it turns out there is no correlation between AHI and pressure, well, what does that tell you?

Jerry Image


_________________
MachineMask
Un-treated AHI = 9.5
Titrated prssure: 6 cm
Ave. AHI after therapy = 0.5
Ave. Snore Index = <10
Current pressure = 9 cm

User avatar
Jerry69
Posts: 334
Joined: Fri Nov 11, 2005 7:10 am
Location: The Villages, Florida

PS

Post by Jerry69 » Sat Dec 24, 2005 7:58 am

The Swift is not noisy at 4 cm.


_________________
MachineMask
Un-treated AHI = 9.5
Titrated prssure: 6 cm
Ave. AHI after therapy = 0.5
Ave. Snore Index = <10
Current pressure = 9 cm

Perry
Posts: 35
Joined: Tue Dec 20, 2005 6:26 pm

Post by Perry » Sat Dec 24, 2005 9:51 am

Jerry:

There are many things that the results you speak of are telling you.

First (and I hope this is the case): That you only need 4 Cm H2O pressure to get adequate treatment. Increasing pressures do not always improve things.

Second: you may just have had a good night. It is not uncommon for peoples pressure needs to change somewhat from night to night (or week to week) due to a variety of reasons. While some people are rock steady on their needs, others range somewhat in needed peak pressures (a range of 4 over a month is not uncommon). So tonight - or another night you may need more pressure.

I suggest that you do not draw any real conclusions on any one setting without a multi-day trial.

Third (and beyond): One of the things people have to be careful on APAPs and there ability to collect data is believing that the chart is really telling you the real story. They look so good - and can be so wrong.

The real test of if PAP is working is that you feel adequately rested - and have energy and mental alertness all day - for many days in a row. If you are well rested you may not notice a not so great night of sleep - it may take 3 or 4 days before you start to notice how tired you are.

The problem with APAP Charts is that the machine may be looking for something different than what you personally experience. What is a Hypopnea, an Apnea, a Flow Limitation? How has the Mfr of your machine defined it? How do they detect it - and what are the limitations in that detection technology?

People can have real obstructive events that cause an arousal (clearly seen in the sleep lab). One machine may see that event - but be programed to ignore it. Another machine may not even detect that event. Both of those machines would produce "great" output data. To them - no event happened. To you - you were prevented from getting stage 4 sleep. On the flip side. A third machine may read all kinds of things that do not cause you an arousal - over respond and run you up to maximum pressure - produce graphs that look "horrible" - when in actuality you personally did not have a problem (as an example: many of the early APAPs would record rolling over - or being jabbed in the back by your spouse - as a major event and respond).

I am a great proponent of APAP and the data that they can collect - but their are limitations in the machines and the technology. Only if the machine is actually detecting your personal events, appropriately responding to them, and recording them does it really work. While it is usually clear when a machine actually does not work right (totally under responds - or totally over responds), the toughest cases are when the machine partially correctly reads you and partially correctly responds to you.

So the real question is how do you feel. Only if you feel really good for a series of days can you start to correlate to what the graph is telling you.

Which brings up the question of "What is feeling good?" Here are how I define the recovery and treatment process as I experienced it.

"Walking Dead" is where most of us start at. That is how a person with severe OSA problems feels before PAP. You seem to sleep very well - but am always tired, cannot think hardly at all, and can barely function well enough to eat, bathe, do a basic job (with not a good performance), and usually come home to die in front of the TV and sleep (you can sleep at the drop of a hat). Life is no fun at all, and you can be extremely frustrated because you know that something is wrong and you used to be able to do better. Your personal relationships sink to all time lows. You are probably so far out of it that your "other" in life may seek others for intellectual level or intimate companionship so that they can have a life at all (The best time to tell your “other” that you love them or care for them is before someone else tells them – I suggest that you do so daily – starting right now: even if you are not in the “walking dead” zone).

"Basic Treatment" is the next step. In my case this happened first on CPAP, then on AutoCPAP that was set-up totally wrong for me. All the sudden you feel much better. You start to notice that there is more to life than you have been experiencing lately. You now find the energy to do things in the morning you really should have been doing all along like flossing teeth, or some house cleaning. But you still cannot think clearly like you used to, while your job and personal relationships improve a little, you still come home from work exhausted, and retire to the TV or recreational reading because you don't have the mental energy to do anything else.

It is amazing how well you feel compared to before - but you still don't have the energy that you remember (or that you see other people with). Naps are still needed if you are to function at all throughout the day. Often it takes a couple of months – but there comes a point where you know that you are not where you should be.

I will note that most sleep Dr’s are satisfied once a person progresses from Walking Dead to Basic Treatment. They consider you a success, and are not likely to provide much support past this point. My personal experience is that I was better off dealing with a family Dr to move beyond this point as I was dealing with other health issues.

"Advanced Treatment" may follow. Here your feel so tremendously great compared to before. Life starts to be good again; you can see that there is a future (especially in the morning when you get up). Your sleep seems to be under control, you have the opportunity to do things in your job and personal life that you could not even imagine before. Mentally you can start to think of several different things early in a day. But you still feel run down in the afternoon (this must be normal you figure) and look forward to bed each night.

"Mental Zest" is where you want to be, what happens when treatment is really working well for you. All the sudden your mind and life is free again. You typically wake up feeling great, can do multi-tasking of different mental task (which really helps your job and personal life), and have energy all day. Can stay up much of a night if needed - with energy and mental involvement - without major effects on the next day. You can see how to make life great once you solve those problems that accumulated when you weren’t feeling so well (and some of those problems can be major issues to solve). You finally realize that this is what normal life should be. You may recognize at this stage that this is how you felt when you were a teenager. As long as you have a positive attitude about yourself and life - you are ready again to take on the word…

It took me years to get to the Mental Zest stage. I would like to claim that I can stay there. Pollen season knocks me back to the Advanced Treatment, and excessive work hours can easily knock me back to the Basic Treatment Stage (simple exhaustion). Proper diet, nutrition, and exercise are also required to keep you at the Mental Zest stage.

May each of you have a safe and quick journey to “Mental Zest.”

Perry



Perry


Mrs.Smith20
Posts: 1
Joined: Sat Dec 24, 2005 9:43 am

Personal Sleep Study

Post by Mrs.Smith20 » Sat Dec 24, 2005 9:54 am

I was wondering how you are getting your AHI?
Does your CPAP record the number of event you are having during the night?


User avatar
rested gal
Posts: 12881
Joined: Thu Sep 09, 2004 10:14 pm
Location: Tennessee

Post by rested gal » Sat Dec 24, 2005 10:22 am

Hi Mrs.Smith,

Jerry's machine, the Respironics REMstar Pro 2 (a straight CPAP machine) records data which can be downloaded with Encore Pro software. Encore Pro software tells him what his AHI was for that session.

If Jerry had been using a REMstar Pro without the "2" in its name, he would not have gotten all that info. The most basic machines record only "compliance" (hours of use.) Jerry's Pro 2 CPAP machine gives him more data.

Incidentally, the graph Jerry posted was generated by a program called "MyEncore" - a supplemental free program developed by Derek.

P.S. Very interesting post, Perry. Thanks for the info!

User avatar
Ric
Posts: 612
Joined: Sat Oct 22, 2005 5:41 pm
Location: Left Coast

what's missing?

Post by Ric » Sat Dec 24, 2005 11:12 am

Jerry,

IMO, you should bump it high enough so you know where the ordinate (AHI) touches zero. How else will you know what pressure it will take to yield ZERO uncompensated apnea/hypopnea events? You may actually be a "NINE", but who knows? Maybe more. I think the pressure should be high enough to at least find the end point, and then adjust the upper limit.

Another point to ponder, those (presumed) events above 8 on your graph, if they exist, have got to be the most worstest. And they escape treatment.

I have had a few excursions into the 15-16 range, but that was before I changed to EP 1.5 (and lost all my previous data). I now keep the upper limit set on 16, and hope I don't actually get there. In the SS I was titrated at 7, which suggests that if I had a CPAP set at a constant 7, I would miss 87% of the hypopneas, and 59% of the apneas. I just happened to have a "good" night at the sleep lab. And that's not good. You will notice most of the action is above 7.


Image


Perry, it looks like Jerry's data is based on just short of a month, not just a night. And he probably IS spending most of the time at 4, as you suggest. The advantage of an APAP is that it can respond immediately to an apneic event above some predetermined threshold or values. I suppose there are different interpretations of what it means to be "treated". Granted that CPAP at the lower pressures will increase ventilation and thereby perfusion and O2 saturation and all that, accompanied by subjective feelings of "improvement". I see "treatment" as an intervention of specific severe events that if untreated will obviate all the other wonderful benefits. In my case there were recorded high pressure apneas lasting over a minute. I question that I would even benefit much from the lower settings suggested by the SS. And I am not patient enough to wait months feeling sort of better knowing that I should really feel a heck of a lot better than that. You obviously have much longer experience at using CPAP, and your remarks are well intended and well taken and appreciated. I'm not sure I understand what you meant about the data "looks so pretty and can be so wrong". I spend most of my time "questioning data", (but not as a sleep study professional). I understand what you say about the limitations of machines and instruments. Some of this data looks pretty solid.

(And likewise, thanks again to Derek for this powerful new tool).

He who dies with the most masks wins.

User avatar
rested gal
Posts: 12881
Joined: Thu Sep 09, 2004 10:14 pm
Location: Tennessee

Re: Doing my own sleep study - surprising results

Post by rested gal » Sat Dec 24, 2005 11:41 am

Jerry69 wrote:I was titrated at 6 cm. I increased this to 8 cm and saw some reduction in AHI's. Last night, I set my Remstar Pro II with C-flex on 4 cm, the lowest it will go. My AHI was 1.5, one of the lowest I have recorded?
I think Jerry is using a straight CPAP machine (Respironics REMstar Pro 2 with C-Flex) - not an autopap.

User avatar
Jerry69
Posts: 334
Joined: Fri Nov 11, 2005 7:10 am
Location: The Villages, Florida

Post by Jerry69 » Sat Dec 24, 2005 2:05 pm

Ric, I'm pondering your chart. You have an APAP, right? Doesn't it eliminate all apneas/hypopneas if the set pressure range is broad enough? To answer my own question: apparently not, as there are AhI's recorded for each pressure 6-11. Does this mean that the machine doesn't respond adequately to eliminate the disturbances? I'm obviously confused.

Perry, thanks, for your evaluation. My symptoms were never "Walking Dead", more like "Basic Treatment". I could function satisfactorily but craved a 'power nap' of 20 minutes at noon. But, if I had commitments that prevented the 'power nap', I could get over the craving and function okay until bed time (11 pm). The symptoms that brought me to the sleep doc were snoring, the desire for a short nap at noon, and sleepiness when driving. I was otherwise alert and functional. I will use 4 cm for a few more nights to see the results. Actually, it seems easier to breathe with 6, or even, 8 cm, than 4 cm. And, I'm considering your point that maybe 4 cm is enough to keep my airway open, in which case 6 or 8 or more would be unnecessary. Of course, I was leaning the other way, i.e, if 4 cm is enough, I don't need any at all. Not as sound as your position, admittedly. I'll probably concede [to myself] that I need some PAP and settle on CPAP of 6 cm, as prescribed from the results of the sleep test. But, I'm not going to have a guilt fit, if I choose not to use CPAP while on travel or vacationing, or even at home occasionally, if I don't feel like it.

I wonder if there is a pressure at which you are assured of eliminating all apneas and hypopneas? Ric, you seem to think so and your chart shows that at 12 cm you achieve that. Correct? I don't understand APAP, however, and I don't know if that means that you have never had an event at 12 cm or higher. Please explain what this chart means for APAP. I'm using CPAP and it is easy. When the machine is set to a constant 4, 6, or 8, as was mine, the results are as defined by the chart. I'm going round and round, so help me: "Hep, hep, hep me, Rhonda." Anyone know where that comes from?

Rested Gal, thanks for your response to Mrs. Smith. And, yes, my machine maintains a constant pressure (CPAP) and is not an Auto (APAP). I suppose it is something new that a CPAP will record all of the sleep disturbing events (to a Smart Card, in this case), as this was only the ability of an APAP in the past. Right?

I'd be bored with CPAP by now if it weren't for ability to see and chart my results...and to continue the quest for the perfect interface. (Aura is on the way. I hope it has a clear box on top, RG. I forgot to specify. Will the picture on the box reveal?)

Jerry Image


_________________
MachineMask
Un-treated AHI = 9.5
Titrated prssure: 6 cm
Ave. AHI after therapy = 0.5
Ave. Snore Index = <10
Current pressure = 9 cm

User avatar
Ric
Posts: 612
Joined: Sat Oct 22, 2005 5:41 pm
Location: Left Coast

Post by Ric » Sat Dec 24, 2005 5:20 pm

Jerry69 wrote:Ric, I'm pondering your chart. You have an APAP, right? Doesn't it eliminate all apneas/hypopneas if the set pressure range is broad enough? To answer my own question: apparently not, as there are AhI's recorded for each pressure 6-11. Does this mean that the machine doesn't respond adequately to eliminate the disturbances? I'm obviously confused.
The surprise answer, YES it does. The machine stays at the lower limit that you have set until it detects a dimished or severely attenuated airflow. Then it bumps the pressure up stepwise until it restores the airflow. The various points on the graph represent the AHI index for which the machine responded and was able to compensate for the event AT A GIVEN PRESSURE. The meaning is the same for a CPAP and an APAP. The difference being that an APAP is able to throw each AO/H event into a pressure bucket, prospectively.
jerry69 wrote:I wonder if there is a pressure at which you are assured of eliminating all apneas and hypopneas? Ric, you seem to think so and your chart shows that at 12 cm you achieve that. Correct?
You are correct, this suggests 12, although I think the number might actually be 11, since there was actually NO cumulative time spent on 12. The highest pressure required to compensate for ALL events was 11. Another informative graph would show the cumulative AHI as a function of pressure. And eventually there is a given pressure (assuming the upper limit is high enough) which is adequate for all detected events. This one displays the cumulative amount of time spent at each pressure level. You can see the lower pressure is at 6, it doesn't "really" trail off to 5 the way it appears.
Image
Jerry69 wrote:I don't understand APAP, however, and I don't know if that means that you have never had an event at 12 cm or higher. Please explain what this chart means for APAP.
Not with this data set. Prior to that I had a couple of events that were in the 15-16 range, and the APAP compensated as needed.
Jerry69 wrote:"Hep, hep, hep me, Rhonda."
Roy Orbison? Beach Boys?
Rested Gal wrote:I think Jerry is using a straight CPAP machine (Respironics REMstar Pro 2 with C-Flex) - not an autopap.
You are so right, I stand corrected. And part of my reply was with a mindset of thinking "upper limit, lower limit". Not so with a CPAP. Sorry, I have only owned an APAP. (Can't imagine NOT owning one.!)
Perry wrote:I suggest that you do not draw any real conclusions on any one setting without a multi-day trial.
Yes, I can see that Jerry only has one data point at 4 cm/H2O pressure. And I TOTALLY AGREE with Perry, a fair assessment would need significantly more data to make any inferences about what is going on there. In fact the graph as it reads is somewhat counter-intuitive, the lower setting on a CPAP I would not expect to have a lower AHI.

He who dies with the most masks wins.

User avatar
Jerry69
Posts: 334
Joined: Fri Nov 11, 2005 7:10 am
Location: The Villages, Florida

Post by Jerry69 » Sat Dec 24, 2005 5:49 pm

Jerry69 wrote:
Ric, I'm pondering your chart. You have an APAP, right? Doesn't it eliminate all apneas/hypopneas if the set pressure range is broad enough? To answer my own question: apparently not, as there are AhI's recorded for each pressure 6-11. Does this mean that the machine doesn't respond adequately to eliminate the disturbances? I'm obviously confused.
Ric wrote:
The surprise answer, YES it does. The machine stays at the lower limit that you have set until it detects a dimished or severely attenuated airflow. Then it bumps the pressure up stepwise until it restores the airflow. The various points on the graph represent the AHI index for which the machine responded and was able to compensate for the event AT A GIVEN PRESSURE. The meaning is the same for a CPAP and an APAP. The difference being that an APAP is able to throw each AO/H event into a pressure bucket, prospectively.
Whoa! You mean that for APAP, the chart tells what would have been had not the pressure of whatever been reached to prevent the event? But, how does the machine know that the pressure change would have resulted in an apnea or a hypopnea, if it fixes it before it occurs?

Ric said:
The meaning is the same for a CPAP and an APAP. The difference being that an APAP is able to throw each AO/H event into a pressure bucket, prospectively.
But, with CPAP, the event actually occurs, and if I understand what you are saying, with APAP, it didn't occur, but would have [at the pressure indicated on the chart].

This is spooky! I still don't understand. Please try again. BUT, I THINK WE SHOULD ALL HAVE APAP'S IF THEY FIX ALL DISTURBANCES.

Jerry Image

Jerry


_________________
MachineMask
Un-treated AHI = 9.5
Titrated prssure: 6 cm
Ave. AHI after therapy = 0.5
Ave. Snore Index = <10
Current pressure = 9 cm

User avatar
Jerry69
Posts: 334
Joined: Fri Nov 11, 2005 7:10 am
Location: The Villages, Florida

'Help Me Rhonda'

Post by Jerry69 » Sat Dec 24, 2005 5:56 pm

Yep, Beach Boys, Brian Wilson, 1965.

Jerry Image

_________________
MachineMask
Un-treated AHI = 9.5
Titrated prssure: 6 cm
Ave. AHI after therapy = 0.5
Ave. Snore Index = <10
Current pressure = 9 cm

User avatar
Ric
Posts: 612
Joined: Sat Oct 22, 2005 5:41 pm
Location: Left Coast

Depends on the meaning of "IS"

Post by Ric » Sat Dec 24, 2005 9:14 pm

Jerry69 wrote:But, with CPAP, the event actually occurs, and if I understand what you are saying, with APAP, it didn't occur, but would have [at the pressure indicated on the chart]. This is spooky! I still don't understand. Please try again. BUT, I THINK WE SHOULD ALL HAVE APAP'S IF THEY FIX ALL DISTURBANCES.
no no no no no no. NO.
Depends on what the meaning of "eliminates" is.
Not in the sense of "prevents".
Finds Destroys. Kills. Deals with.
But not "prevents".

In both cases (CPAP/APAP) the event "occurs". At least on what is recorded and represented on the AHI vs. PRESSURE graph.

In the event of an OSA, an APAP at 6 would bump itself up to a 7, 8... until the event goes away, or reaches its upper pre-set limit. then drops back to the lower pressures only when it detects normal breathing. but the event still "occurs". and it is recorded as an "event" at the highest pressure required to make it go away.

A CPAP at 6, an OSA event occurs, what do you do? stop breathing? wait it out? Wait for the CO2 levels to wake you up or at least get you jump started again, naturally? That is my understanding of CPAP, not actually having experience using one.

I can make the APAP actually BE a cpap by setting the upper and lower limits the same. in which case all the AHI data would appear on that one pressure setting. I think I agree with your analysis, my titration would be a 12, the pressure at which all OA/H events would not occur at all, if that was the intention. That is to say, "prevents".

I'm not sure that total prevention is necessarily the therapeutic goal, so much as "intervention". In my data the average DURATION of OA/H events has dropped DRAMATICALLY to under 15 secs, as opposed to about a minute prior to APAP. (I won't bore you with another graph). That's gotta be good. I can hold my breath 15 seconds anytime, not a problem. Holding my breath for over a minute, hundreds of times a night, that gets annoying.

My APAP with lower limit at 6, and your CPAP set at 6 will "prevent" all OA/H events which would otherwise not occur at all pressures up to and including 6. In the event of an "event" that appears even when the pressure is at 6, the APAP will become a 7, 8... until the "event" goes away. oh quit, i know you know what i mean...
Jerry69 wrote:BUT, I THINK WE SHOULD ALL HAVE APAP'S IF THEY FIX ALL DISTURBANCES.
YES !!!

They "fix", "deal with", 'CRUSH", "demolish", "make go away"...

But not "prevent". Not all. Just the little ones.


(oh, and MERRY CHRISTMAS EVE to y'allz,
Tom, Jerry, Perry, RG, MrsSmith, Guest, et. al.,
we really know how to PARTY do we?)

He who dies with the most masks wins.

User avatar
Jerry69
Posts: 334
Joined: Fri Nov 11, 2005 7:10 am
Location: The Villages, Florida

APAP and AHi's

Post by Jerry69 » Sat Dec 24, 2005 10:00 pm

Ric wrote:
Finds Destroys. Kills. Deals with. But not "prevents".
And,
I'm not sure that total prevention is necessarily the therapeutic goal, so much as "intervention".
I like 'intervention' best.

I understand. The APAP 'intervenes' by increasing pressure until the event goes away, i.e., breathing returns to normal. I suppose that as the event persists, the pressure continues to increase until the event finally subsides. And, it is at this point that the pressure is recorded for that event, be it apnea or hypopnea.

My apneas are usually one per night, sometimes two, sometimes none, and their average duration is 12 sec. Not bad.

Okay, time to turn on the CPAP, at 4 cm, and see what is recorded. After a few nights, I'm going back to at least 6 cm, as it seems easier to breathe.

'Merry Christmas to all and to all a good night.'

Thanks for participating.

Jerry


_________________
MachineMask
Un-treated AHI = 9.5
Titrated prssure: 6 cm
Ave. AHI after therapy = 0.5
Ave. Snore Index = <10
Current pressure = 9 cm

Perry
Posts: 35
Joined: Tue Dec 20, 2005 6:26 pm

Post by Perry » Sat Dec 24, 2005 10:30 pm

RIC Wrote:

I'm not sure I understand what you meant about the data "looks so pretty and can be so wrong".
Ric also had a long discussion on how APAPs prevent events - which is close to what happens.


Here is the explanation of why the data can be wrong - and what happens in the modern machines.

I am sure that many of you have seen graphs of what a normal breathing cycle looks like, and probably some that show how the top of the waveform is clipped flat (flow limitaton) and other things.

The problem is that those are idealized "average" graphs. Real people vary from those ideals. Otherwise, building an APAP would have been ridiculously easy (and everyone's machine would work the same way).

So the Mfr takes their "Ideal" breathing pattern - as seen by the technology they chose to use to detect it (and there can be large variations in that).

Then they define what an "Ideal" snore, flow limitation, hypopnea, apnea, etc are and write a program to look for those "Ideal" patterns (depending on which of those they chose to use - no machine that I am aware of uses all "theoretically" detectable events).

Then experience has taught the Mfr's that you cannot respond to just one "event" - and that in most cases the response needs to be modest. So the Mfr determines how many of what kinds of "ideal" events have to occur before the machine will respond (at all) - and how fast (folks - there once were machines that would literally increase pressure so fast as to blow the mask off your face - waking you in the process - from the first indication of an event - and by false events indications such as rolling over or being jabbed by your spouse - or that "pregnant" pause when the dog licks your face in the middle of the night).

As you can see - even in the ideal case - the machine allows a certain number of events, and then in almost all cases slowly ramps up (allowing more). The idea is to get to a point where the "event rate" stops increasing and then to stabilize at a low event rate (and not wake you from too fast of a pressure increase). Then - after a predetermined amount of time determined by the Mfr the machine will reduce pressure. Should more significant events occur - increasing the event rate - the pressure returns to where it was - or even increases depending on the type of events and the rates.

Just to complicate things: In almost all cases - each Mfr has defined different "Ideal" breathing curves and different "Ideal" events - as they try to target what they think works best (if you would consider just a simple case - How would you define "snore" and how would you detect it...., then how would you respond to it - is some snore OK).

Of course, each person is different - and the key to how well any APAP works in auto mode (and the data collection in any mode) is how well the individuals personal breathing curve and personal event curves match what the machine is looking for. This can be great, good, fair, or bad.

The machine only records and responds to events that match what it has been told to record and respond to. That can be much different than what the person is actually experiencing. Hence the data output may or may not have a good correlation to the patient. In my example: Snore just right for the machine - and it's got a good precursor for you. Snore outside of the machines definition or detection capability - and the machine will report no snore.


This is why I challenge data with the simple test of is the person getting really good sleep treatment. An APAP that matches you well - and is setup properly (a separate issue) - should get a person to - or real close - to "mental zest" in a couple of weeks (or at worst a couple months)assuming that there are no other medical problems.

I can dig up Plenty of "good charts" out of my and other's personal experience alone where the APAP machine did not work for that person - did not allow them to get a good nights sleep. In my case: testing APAP machines from various Mfr's was really instructive - It really takes a masicist to do some of the things I have done - like repeating a 4 day multi-night test 3 times (after full recovery between test) with a machine that does not properly read me and respond to me - pushing me into exhaustion by day 3).

Now, truth be told - most people will do well on most any of the current APAPs. The problems are the people who's breathing and event curves are sufficiently far from "ideal" for that specific machine - as that machine is substantially missing the picture. The graphs look great - and the person is tired and miserable (and having real events that the machine does not see).

One of the advantages to having different Mfr's is that most likely a person can be adequately treated by one of the machines (since they all are actually looking for something different, even if the event shares a common name, - and respond at different rates).

Note however, there are people who's breathing and event patterns are sufficiently removed from the ideals that APAP will not work for them. There are also a number of other health issues that can prevent APAP from working. Of course, there are also a number of other health problems that a properly functioning APAP helps with (compared to CPAP).

Hope this helps (and yes, I know that I have probably created more questions than I answered. How APAPS work - under what conditions - etc is a long subject with a number of possible variables - I can't cover all of them in a few days).

I will be out of town tomorrow as I travel to visit a brother. Have a great Xmas.


Edited to add: If it seems that an APAP is not working that well for you - the first things to look for are proper set-up, mask leaks, and other health problems. Those charts often have other really interesting things in them for people who know how the machines sees, for example, asthma.

Perry

_________________

CPAPopedia Keywords Contained In This Post (Click For Definition): CPAP, Hypopnea, auto, APAP, Travel


Guest

Post by Guest » Sun Dec 25, 2005 12:47 am

the Remstar ProII has the ablity to record data that it sees, but I don't know for sure if that includes a pneumotach sensor like the autopap version, that would greatly increase it's sensitivity to read events. I suspect it does as it would seem cheaper for Respironics in the long run to make all the hardware the identical and then just change the firmware that gets loaded and silkscreen on the model on the outside.

But I wonder if by lowering the pressure to 4cm if that does not somehow impede the ability of the machine to record events accurately. if the machine is going to read events it first has to determine the volume of air in your breath. It has to also compensate for mask & exhaust leak.

your data could be accurate but 4cm is low as you can go, I'm surprised you can tolerate that low a pressure, I'd be starving for air to breath at that pressure. Then if you enabled the c-flex feature I bet it gets pretty stuffy.