Complex sleep apnea and EPR

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
igdoc
Posts: 54
Joined: Sun Dec 22, 2013 1:07 pm
Location: England

Complex sleep apnea and EPR

Post by igdoc » Thu Feb 13, 2014 6:43 pm

Pugsy wrote :
Also, this idea of exhale relief and the drop in pressure during causing centrals...again very small number of people. It happens to a few people on bilevel machines and using EPR on the S9 sort of mimics bilevel but from what I have seen in reports and read...much more likely to occur when the difference between EPAP and IPAP is much larger than the 3 we can do with EPR. Like 8 to 10 difference.
So yes, in theory maybe it is possible for EPR to maybe cause a little imbalance in C020 but it would be extremely rare for it to create a problem. If it were then I would think that EPR would come with warnings and not be considered a comfort feature.


Hi Pugsy,
All respect to you as your knowledge is second to none. However with regard to the above quote I would ask you not to underestimate the potential effect of exhalation relief on excess central events. For example I have complex sleep apnea and, with a pressure of 8 and an EPR of 1, had an AHI averaging 29, all due to central events. Switching off the EPR dropped my AHI to between 5 and 6 and, more importantly, enabled me to start sleeping again. I have avoided needing an ASV machine.

Even an exhalation pressure drop of 1 cm H2O can have a significant effect on the volume of respiration (tidal volume). The blood CO2 level has an apnea threshold, say for example 40mmHg. This means that at 41mmHg - no central apneas/hypopneas occur but at 39mmHg central apneas develop and may then form a recurrent loop. In this way a small change in the volume of ventilation can have a large effect on the number of centrals.

Fully agree that the numbers who are susceptible to Complex sleep apnea are small in percentage terms perhaps 15% initially and only around 3% on long term therapy. However, because of the huge number of CPAP users, this is still a very large number of people who could potentially benefit. I only wish that this information about changing EPR had been available to me when I started CPAP and had all those central events.

Best wishes
Ian

Thought I would put this forward for discussion. Ian.
_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Software: Sandman Series Therapy v1.6 Software
Additional Comments: Sefam APAP machine similar to Sandman

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: Sleepyhead v0.9.6. Encore Pro 2.12. Complex sleep apnea.
IG

User avatar
DoriC
Posts: 5214
Joined: Sat Sep 13, 2008 9:28 pm
Location: NJ

Re: Complex sleep apnea and EPR

Post by DoriC » Thu Feb 13, 2014 10:26 pm

Does this also apply to Pressure Support with Bipaps? This is an important question.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: 14/8.4,PS=4, UMFF, 02@2L,
"Do or Do Not-There Is No Try"-"Yoda"
"We are what we repeatedly do,so excellence
is not an act but a habit"-"Aristotle"
DEAR HUBBY BEGAN CPAP 9/2/08

User avatar
Pugsy
Posts: 65131
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Complex sleep apnea and EPR

Post by Pugsy » Fri Feb 14, 2014 12:11 am

People already panic when they see a central or two...I am not going to tell them to turn off EPR because EPR might be causing them. OMG I got Complex Sleep Apnea because of my exhale relief...I had 5 centrals last night....You have no idea how many panic message I get about centrals.

Not every central is caused by the unstable breathing....sleep onset centrals are normal and not related to pressure one way or the other.

Dori...yes, this would in theory be the same with Pressure support on the bilevel machines..because that is what EPR really does...it's just pressure support up to 3.
If this was such a common problem..then why would the first machine that gets tried often for CompSA be the regular bilevel machines (not ASV)?? Think about it.

Ian...you are the only person that I have ever known have documented evidence that turning off EPR was important in reducing centrals...only person in 5 years of helping people here on the forum. I have read about a study that was done with bilevel machines and it did document the fact that for some people regular bilevel caused central problems in people that had no problems with centrals and regular cpap.
There will always be the exceptions to any norms when it comes to our bodies and health. I don't doubt you at all.
I will put you in the exception category..

But I just don't see the need to go shout it from the rooftops and scare the hell out of every newbie who had a handful of centrals last night.
They get all worked up over the numbers (and usually better numbers than I get) when there's no need.

I think the post I comment about what you said..that thread was started by someone with an AHI of 1.0....and all centrals.
That's one per hour average and it doesn't need anything to be done about it. and it doesn't need to cause extra concern and worry.

People accept obstructive apneas but get all freaked out about Centrals and worry too much about stuff they should set on the back burner for a little while.

When I do work with people who do have a number of centrals that I can't poo poo off....we look at all aspects and try all different combinations of settings to see if we can get any relief in an effort to avoid the ASV machine.
One of the setting we will play with will be EPR...because do understand the theory and possibility and want to cover all bases...but I have yet to see one where the centrals reduced when EPR was reduced or turned off. Not a one.
Not saying it never happens but I haven't seen it and in 5 years I have been through a lot of scenarios with people. I do a lot work with people via private messages...they don't want their reports out for the whole world to see.
Never seen it happen with EPRs little PS.

I did see it once on a BiPap report...where someone decided to use big IPAP and little EPAP....6 EPAP PS of 10 and max IPAP 20...truck load of centrals popped up. He was hyperventilating at those settings

If you want to share you experience with people when talking about centrals...by all means do so.
We share our own personal experience all the....heck that's what we do best.
Just remember newbies are easily scared and don't understand 90 % of the technical stuff we throw at them until alter.
I don't want to cause panic needlessly...they get it enough on their own with out my help.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Sludge
Posts: 953
Joined: Wed Dec 04, 2013 4:36 am

Re: Complex sleep apnea and EPR

Post by Sludge » Fri Feb 14, 2014 4:12 am

igdoc wrote:...I would ask you not to underestimate the potential effect of exhalation relief on excess central events. For example I have complex sleep apnea and, with a pressure of 8 and an EPR of 1, had an AHI averaging 29, all due to central events. Switching off the EPR dropped my AHI to between 5 and 6 and, more importantly, enabled me to start sleeping again. I have avoided needing an ASV machine.
And I would ask you not to underestimate the potential effects of "time" and "coincidence". Meta:
CPAP devices produce steady pressure that does not change during the respiratory cycle. It is the first-line noninvasive treatment for OSAS, and can almost completely suppress obstructive respiratory events. But CPAP typically does not acutely resolve central apnea activity. Patients with CompSAS may have a poor initial experience with CPAP and thus may be nonadherent with therapy. Using retrospective data and OSAS patients as a comparison group, Pusalavidyasagar et al39 documented that although patients in both groups had a similar diagnostic AHI, CompSAS patients had a significantly higher AHI during the CPAP titration (CompSAS AHI on CPAP 24.6 ± 21.6 vs OSAS AHI on CPAP 2.1 ± 2.7), and most of the residual respiratory disturbance was from central apneas that emerged on CPAP. Mean time to the first follow-up was shorter in CompSAS patients (46.2 ± 47.3 days vs 53.8 ± 36.8 days), and interface problems were more common in CompSAS patients, especially air hunger/dyspnea (8.8% vs 0.8%) and inadvertent mask removal (17.7% vs 2.6%). Some researchers suggest that most CSA events can disappear after continued CPAP use. It is still not clear what proportion of patients with CompSAS may improve over time with continued CPAP treatment. Dernaika et al,30 in their prospective case-control study, documented the disappearance of central apnea activity with CPAP in 12 (86%) of 14 patients with CompSAS over 2–3 months. The investigators stated that CSA events occurring during CPAP titration were transient and self-limited, whereas only one patient with diastolic dysfunction on baseline transthoracic echocardiography showed continued CSA events. Unfortunately, the applicability of their data is limited because they excluded patients with any significant comorbidities, including CHF by history or previous echocardiography, chronic obstructive pulmonary disease or other significant lung disease, daytime hypercapnia or hypoxemia, cerebrovascular disease, seizure disorder, or a history of use of benzodiazepines, narcotics, or illicit drugs. In a retrospective case series, Kuzniar et al34 evaluated 13 patients who had CompSAS during the first overnight CPAP titration. Repeat therapeutic PSGs were performed after 195 (49–562) days. On the follow-up study, seven of 13 patients reached AHI < 10. The low number of patients and retrospective character of data collection are the main limitations of this study. Cassel et al37 investigated 675 patients with OSA receiving CPAP treatment. During the baseline CPAP night, 82 patients had CompSAS. The follow-up evaluations took place after a median of 14 weeks, and 30 patients fulfilled the criteria for CompSAS. The prevalence of CompSAS was significantly reduced from 12.2% to 6.9%. In a large retrospective study of 1286 patients with a diagnosis of OSA who underwent titration with CPAP, CompSAS occurred in 84 patients (6.5%). However, after 8 weeks of CPAP therapy, they estimate that 1.5% of the patients with OSA continued to have CompSAS with long-term use of CPAP.35 From these very limited data, it seems likely that most CompSAS patients will improve completely with long-term use of CPAP. This variability in CPAP response is probably due to heterogeneity in the CompSAS group or to variable factors governing adaptation that are not yet well understood. However, how to distinguish patients who might stabilize over time from those who will suffer ongoing complex apnea activity is unknown, and the time required for adaptation of ventilation is also unknown. Patients who have a bad initial experience with CPAP have a worse adherence to therapy, and those who fail with CPAP compliance over longer periods will not enjoy benefits of treatment while awaiting adaptation.
http://www.ncbi.nlm.nih.gov/pmc/articles/PMC3704546/
You Kids Have Fun!!

Delta4
Posts: 54
Joined: Sun Jan 05, 2014 12:03 pm

Re: Complex sleep apnea and EPR

Post by Delta4 » Fri Feb 14, 2014 7:11 am

Pugsy wrote:...But I just don't see the need to go shout it from the rooftops and scare the hell out of every newbie who had a handful of centrals last night.
They get all worked up over the numbers (and usually better numbers than I get) when there's no need.

I think the post I comment about what you said..that thread was started by someone with an AHI of 1.0....and all centrals.
That's one per hour average and it doesn't need anything to be done about it. and it doesn't need to cause extra concern and worry.

People accept obstructive apneas but get all freaked out about Centrals and worry too much about stuff they should set on the back burner for a little while. ....
Pugsy, I'd really appreciate your thoughts on my situation. Please understand that I'm not trying to be combative.

I assume you're referring to my recent post where I said:
I have 4-5 bad nights each week (50-80 central events with a AHI of 4.5 - 8.) On my better nights, my AHI is less than 1. That leaves me with a monthly AHI of around 3.

On a later post in that thread, I gave more details:
On my bad nights, I have clusters of centrals 1-5 hours after I go to bed. By clusters, I mean I might have 20 events in a 30 minute period. Or 45 events in an hour. Unfortunately my averages get diluted because I sleep so long. I also wake up gasping for air during many of my bad nights. I have a headache after waking up most of the time which was rare before. All this screams to me that we need to try a different machine!

Here's some other information that isn't in that thread:

I'm not ignoring obstructive events because I rarely have them.  Over 90% of my events in a month are central and only 3% are obstructive. ButI think that only looking at the monthly AHI doesn't make sense in my case because I have such wide swings in my nightly AHI. In other words, how can my doctor expect me to ever feel rested when I have 50-80 central events a night (usually in clusters) for 4-5 nights every week? I don't see how I can ever pay down my sleep debt continuing like this.

Sometimes I get up to use the bathroom after sleeping 10 hours and notice my AHI is around 11-12.  I go back to sleep for 4 more hours and my AHI for the night ends up around 8 just because I had zero or only a few events in the last 4 hours of sleeping 14 hours. This situation happens frequently.

I waited until I had 3 months of these consistent results before posting about it here so I wouldn't appear to be overreacting to a few bad nights of data.

My family and friends tell me to quit using the cpap. I'm not doing that because it has my OSA under control. And I have other conditions which make me more likely to have a stroke. But I'm positive something needs to change with my cpap therapy because I started feeling significantly worse right after starting on cpap 3 months ago.  I'm talking about debilitating fatigue, daytime sleepiness, and brain fog. I'm desperate for some relief.

Thank you for your time.

_________________
Mask: SleepWeaver Elan™ Soft Cloth Nasal CPAP Mask - Starter Kit
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResScan & Sleepyhead, pulse-oximeter CMS-50D+, pressure=12, SoClean2, Quattro FX FFM

User avatar
Pugsy
Posts: 65131
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Complex sleep apnea and EPR

Post by Pugsy » Fri Feb 14, 2014 8:03 am

Delta4 wrote:I assume you're referring to my recent post where I said:
I have 4-5 bad nights each week (50-80 central events with a AHI of 4.5 - 8.) On my better nights, my AHI is less than 1. That leaves me with a monthly AHI of around 3.
No, I actually wasn't thinking of you or anyone recently in particular. I was thinking of past newbies that I have had to reassure that a few to several centrals don't mean that they have complex sleep apnea. I have had people PM me and want to know about getting an ASV machine because they developed 3 centrals and didn't have any during the sleep study.
Way too much "OMG, I have centrals" panic.

I only recently read your thread and have had other irons on the fire and really didn't have much to offer that would have been different from what was already offered.

When we have more bad nights than good nights and we feel worse than without CPAP therapy...something is changing or not right. Gotta try to figure out what before trying to fix it though..and that's sometimes really difficult.

Rather than muck up this thread...I will go back an read your thread again to see if maybe I missed a question to ask.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
JohnBFisher
Posts: 3821
Joined: Wed Oct 14, 2009 6:33 am

Re: Complex sleep apnea and EPR

Post by JohnBFisher » Fri Feb 14, 2014 11:59 am

Several comments on this thread:
Pugsy wrote:... People already panic when they see a central or two...I am not going to tell them to turn off EPR because EPR might be causing them. OMG I got Complex Sleep Apnea because of my exhale relief...I had 5 centrals last night....You have no idea how many panic message I get about centrals.

Not every central is caused by the unstable breathing....sleep onset centrals are normal and not related to pressure one way or the other. ...
I also get those PMs. And it's true, but it's important to keep reminding people that we awaken all the time during sleep and fall back asleep without any memory of it. And those instances are when we end up with those sleep transition centrals. That's why I found an posted the following article:

Ventilation is unstable during drowsiness before sleep onset
http://jap.physiology.org/cgi/reprint/99/5/2036

Those and many others are available in Rested Gal's Links to Central Apnea:
viewtopic.php?t=3025
Pugsy wrote:... Ian...you are the only person that I have ever known have documented evidence that turning off EPR was important in reducing centrals...only person in 5 years of helping people here on the forum. ...
I've seen about three in about 10 years. Remember because of my focus I get more of the really persistent central apnea cases. Still it's REALLY rare that EPR causes problems.
Pugsy wrote:... But I just don't see the need to go shout it from the rooftops and scare the hell out of every newbie ...
Exactly. Turn it off if there is a problem. Otherwise it helps folks new to xPAP therapy better accept their therapy.
Pugsy wrote:... When I do work with people who do have a number of centrals that I can't poo poo off....we look at all aspects and try all different combinations of settings to see if we can get any relief in an effort to avoid the ASV machine. ...
Exactly. They are both more expensive and require yet another sleep study or two (depending on the insurance company and person) to document the need for the therapy and identify the correct settings for the units. And then we often need to work with the person to properly tweak their therapy. And worse yet, it often takes a few months to fully acclimate to the ASV therapy. The wild swings in pressure are sometimes too much for some people. It's not a step to be taken if there is no real need for it.
Delta4 wrote:... Pugsy, I'd really appreciate your thoughts on my situation. Please understand that I'm not trying to be combative. ...
Hope you don't mind, but I'll also wade into this conversation.
Delta4 wrote:... I assume you're referring to my recent post where I said:
I have 4-5 bad nights each week (50-80 central events with a AHI of 4.5 - 8.) On my better nights, my AHI is less than 1. That leaves me with a monthly AHI of around 3.

On a later post in that thread, I gave more details:
On my bad nights, I have clusters of centrals 1-5 hours after I go to bed. By clusters, I mean I might have 20 events in a 30 minute period. Or 45 events in an hour. Unfortunately my averages get diluted because I sleep so long. I also wake up gasping for air during many of my bad nights. I have a headache after waking up most of the time which was rare before. All this screams to me that we need to try a different machine!

Here's some other information that isn't in that thread:

I'm not ignoring obstructive events because I rarely have them. Over 90% of my events in a month are central and only 3% are obstructive. ButI think that only looking at the monthly AHI doesn't make sense in my case because I have such wide swings in my nightly AHI. In other words, how can my doctor expect me to ever feel rested when I have 50-80 central events a night (usually in clusters) for 4-5 nights every week? I don't see how I can ever pay down my sleep debt continuing like this.

Sometimes I get up to use the bathroom after sleeping 10 hours and notice my AHI is around 11-12. I go back to sleep for 4 more hours and my AHI for the night ends up around 8 just because I had zero or only a few events in the last 4 hours of sleeping 14 hours. This situation happens frequently.

I waited until I had 3 months of these consistent results before posting about it here so I wouldn't appear to be overreacting to a few bad nights of data.

My family and friends tell me to quit using the cpap. I'm not doing that because it has my OSA under control. And I have other conditions which make me more likely to have a stroke. But I'm positive something needs to change with my cpap therapy because I started feeling significantly worse right after starting on cpap 3 months ago. I'm talking about debilitating fatigue, daytime sleepiness, and brain fog. I'm desperate for some relief. ...
I mentioned in the other thread that you probably need to see another doctor. Of course you might be able to educate your doctor that "average" is a horrible single number to use to determine how you are sleeping. Would a cardiologist be happy to hear that his patient only had 0.25 heart attacks per month? It's both the frequency (which the average partially reveals) and the intensity (which average completely obscures) that are important.

Many, many people think "average" is all you need to understand a group of numbers. In fact, it's only one part of the measure needed. Let's take two different groups of numbers:

Group A
50, 50, 50, 50, 50, 50

Group B
0, 100, 0, 100, 0, 100

The AVERAGE of both groups is a nice "50". But that's meaningless to explain the difference between the two. This is where the 90% value is often used or the Standard Deviation. For example, the Standard Deviation of Group A is 0 (zero). There is no variation. But the standard deviation of Group B is about 55. That indicates there is a HUGE variation in the set of data.

Unfortunately for you, you have large variations in your data. An average only tells part of the story. If your current doctor won't help you address the situation (and remember that some doctors think they know more than everyone), then find a doctor that will help you.

However, please be aware that you still are a bit on the borderline side of needing an ASV unit. If you go to see another doctor they may well see if a BiLevel machine would help. Sometimes by using a BiLevel machine you can back off the pressure enough that those central induced apneas are reduced if not eliminated. It happens often enough that your doctor's next move should be to try that.

Hope that helps.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: User of xPAP therapy for over 20 yrs. Resmed & Respironics ASV units with EEP=9cm-14cm H2O; PSmin=4cm H2O; PSmax=15cm H2O; Max=25cm H2O
"I get up. I walk. I fall down. Meanwhile, I keep dancing” from Rabbi Hillel
"I wish to paint in such a manner as if I were photographing dreams." from Zdzisław Beksiński

igdoc
Posts: 54
Joined: Sun Dec 22, 2013 1:07 pm
Location: England

Re: Complex sleep apnea and EPR

Post by igdoc » Fri Feb 14, 2014 3:47 pm

Hi DoriC, yes the same applies to Bipaps and more so. Bilevel with backup rates can sometimes be successful in treating
Complex sleep apnea but this is by pushing breathing when it would otherwise stop and this can further lower CO2 levels.

Hi Sludge, fully agree that many people with complex sleep apnea will improve over time as the body can often reset the CO2 apnea threshold. Unfortunately I was not one of them and my AHI was always over 20 for 2.5 months and was not dropping. The AHI dropped on the night I switched off the EPR and only returned to over 20 when I tried switching it back on one month later.

Hi Pugsy, fully agree about the sleep onset centrals which occur due to the body changing the CO2 apnea threshold when falling asleep and wakening. I talk about these regularly in my posts and I have never suggested than people with low AHIs should panic about their small number of normal central events. I posted on Delta4's thread because she had some nights of clustered central events with a raised AHI which could well be related to hyperventilation induced central apnea. I am sure she would welcome any other suggestions though.

Although rare I am not the only one out there, for example there have been 2 people on this forum in the last 3 months who have shown the same result by reducing their EPR/CFLEX/AFLEX, Kcellwood springs to mind. I have also found 4 other people on other forums in 3 months with the same result. Not many I grant you but most of the Complex sleep apnea folks are already on ASV machines.

Hi JohnBFisher, as mentioned above I am regularly posting about these normal physiological central events. I am not suggesting that EPR is the cause of complex sleep apnea but just suggesting a potential and easy therapeutic measure which is generally harmless. Hyperventilation may be at the root of many non-cardiac/non neurological cases of central sleep apnea and be similar to altitude induced hyperventilation central sleep apnea (apart from not being driven by low O2) but this is theory and remains to be proven.

Sorry If I upset anyone. I am not knocking EPR which is of great benefit to a large number. Just ask that reducing EPR/CFLEX/AFLEX be considered if large numbers of centrals appear. After all there is nothing to lose.

Regards to all
Ian

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: Sleepyhead v0.9.6. Encore Pro 2.12. Complex sleep apnea.

User avatar
Pugsy
Posts: 65131
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Complex sleep apnea and EPR

Post by Pugsy » Fri Feb 14, 2014 4:53 pm

Let me look at this from a newbie point of view. Putting newbie hat on......

I read....
"I turned EPR off and my centrals went away"...cause and effect ...that means EPR must cause centrals if turning it off eliminates them.....so I had 2 centrals last night....I guess I can't use EPR....it causes centrals.

I guarantee that is what newbies will see. They already think that they have one foot in the grave if a central gets flagged anyway.

I am beginning to think Dr K is right...give everyone an ASV and cover all bases.

I think the polite grown up thing to do here is politely agree to disagree and leave it at that.
At least that is what I am going to attempt to do.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Sludge
Posts: 953
Joined: Wed Dec 04, 2013 4:36 am

Re: Complex sleep apnea and EPR

Post by Sludge » Fri Feb 14, 2014 7:13 pm

igdoc wrote:The AHI dropped on the night I switched off the EPR and only returned to over 20 when I tried switching it back on one month later.
Well, I am quite aware of the argument that EPR is bilevel (having invented it).

However, seeing summary data (was 20, then 5, then 20 again) can sure be misleading.

Can you upload the raw data files for review?

Tnx.
You Kids Have Fun!!

User avatar
Sludge
Posts: 953
Joined: Wed Dec 04, 2013 4:36 am

Re: Complex sleep apnea and EPR

Post by Sludge » Fri Feb 14, 2014 7:15 pm

Because when I see somebody say
igdoc wrote:Switching off the EPR dropped my AHI to between 5 and 6 and, more importantly, enabled me to start sleeping again.
what I'm thinking is that AHI 20 is just a pile of SWJ.
You Kids Have Fun!!

User avatar
Sludge
Posts: 953
Joined: Wed Dec 04, 2013 4:36 am

Re: Complex sleep apnea and EPR

Post by Sludge » Fri Feb 14, 2014 7:32 pm

igdoc wrote:Just ask that reducing EPR/CFLEX/AFLEX be considered if large numbers of centrals appear.
Actually, I do not believe that the mechanism that EPR may be able to generate centrals is able to occur with Respironics expiratory relief modalities.
You Kids Have Fun!!

igdoc
Posts: 54
Joined: Sun Dec 22, 2013 1:07 pm
Location: England

Re: Complex sleep apnea and EPR

Post by igdoc » Fri Feb 14, 2014 9:30 pm

Hi Sludge
Not going to enter into a contest. Good point about the potentially different effects of EPR and CFLEX/AFLEX.
Ian

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: Sleepyhead v0.9.6. Encore Pro 2.12. Complex sleep apnea.
IG

User avatar
Sludge
Posts: 953
Joined: Wed Dec 04, 2013 4:36 am

Re: Complex sleep apnea and EPR

Post by Sludge » Sat Feb 15, 2014 2:13 am

igdoc wrote:Not going to enter into a contest.
Well then next time don't put up 6 posts about EPR causing CompSAS on a discussion forum if you don't expect "discussion" to ensue.
...different effects of EPR and CFLEX/AFLEX.
This was looked at nearly 9 years ago when EPR first appeared:

viewtopic.php?f=1&t=4514&p=38022&hilit= ... ave#p38022

http://www.apneasupport.org/topic894.html
Here's the waveform I generated at 20 cmH2O:

Image

The first two arrows represent the termination of EPR (the first tiny waveform), and you can see how close they are to inspiration. This suggests that the trigger might be inspiratory flow, which could create a problem if you're trying to inspire from a sub-therapeutic pressure level. Or, it could be a very low expiratory flow rate. Try as I might, I could not verify this either way, so we'll have to wait for the manufacturer's response. But you really need to be at therapeutic CPAP level prior to inspiration, so we'll have to look at this carefully.
The second two arrows show where the EPR did not terminate, and inspiration occurred at well under therapeutic levels. I was able to generate this situation at all pressure levels. The obvious problem here would be if the EPR level was below the apnea threshold, therapy would not be effective. Again, I don't know why this occurred, and passed this information along to the rep.
This situation might be overcome by raising the baseline CPAP slightly, but the point is that using this EPR might change your CPAP requirements slightly.
I should also point out that there is a time function for EPR, if no breath is detected after a period of time, EPR is immediately suspended, so it seems that only a single breath would miss detection.
sleepydave wrote:When the coming of Expiratory Pressure Relief (EPR) was first announced, I had some questions as to whether this modality would offer relief on active expiration only during the CPAP mode, and perhaps address the issues that other expiratory adjuncts were having, or if the drop in expiratory pressure were carried out all the way to the next inspiration, relying on inspiration as the trigger to terminate EPR, and thus essentially operate in a BiPAP mode.

The following waveform analyses were performed on the EPR mode at 10 cmH2O with an EPR setting of 3 cmH2O. The breath rate is approximately 12.

The first graph shows the breathing waveform on top, inspiration being an upward deflection and expiration downward, while the bottom graph is measuring pressure. The pressure settings are seen as faint numbers at the left of the pressure waveform. You can see that the EPR, reflected as a drop in the therapeutic pressure on the pressure waveform down to about 7 cmH2O, is carried out all the way to the point of inspiration, and the inspiratory effort therefore takes place at a sub-therapeutic pressure. The baseline pressure returns to 10 cmH2O, but not until after inspiration has begun. In other words, inspiration is the trigger to terminate EPR, and instead of a CPAP pressure of 10 cmH2O with an expiratory adjunct, we are effectively left with BiPAP of 10/7:

Image

This might not make a clinical difference if the patient ends up with the same results on BiPAP 10/7 that he would have on CPAP 10 cmH2O (which could be the case if there were only flow limitations, snores, or hypopneas). But if the new EPR-defined EPAP is below the apnea threshold, then there could be a problem.

In the second graph, the waveforms are superimposed to show more clearly that inspiration is occurring at a sub-therapeutic level:

Image

There is a time limitation associated with the termination of EPR. In the next graph, you can see how the EPR eventually terminates and returns to baseline. In this instance, the breath rate was approximately 6, so the time to EPR termination was appoximately 5 seconds. The first arrow represents EPR termination, while the second signifies patient breath:

Image

And here again, the graphs are superimposed to show the return to baseline relative to inspiration:

Image

This means that eventually, there will be a return to baseline CPAP if in fact, an apnea occurs, and at the most, only one breath would be missed. Is the net result clinically relevant? I'm not sure either way. But if you're generating negative intrathoracic pressure or creating arousals, then there could be an issue.

In re: putting EPR in the AutoSet mode, that could be an effective way to overcome this supposed shortcoming of EPR. If events were to start occurring at, in this case, "10 cmH2O of CPAP". then baseline pressure could be raised, theoretically, to "13 cmH2O of CPAP", or effectively BiPAP 13/10. Course now we're right back where we started. The only outstanding question would be if the 13/10 format was better tolerated than the straight 10.

But how would the algorithm work? If the apnea identification in AutoCPAP is 10 seconds, and the EPR terminates at (in this case) 6 seconds, how would it know to increase the CPAP (really the "EPAP" segment of EPR) to address apneas? I would assume that flow limitations would be properly addressed with CPAP increase (because you're really raising the "IPAP" segment of EPR).

As an aside, therein lies the problem once you start talking about Auto-BiPAP. Do you increase the IPAP, and keep EPAP fixed, or do you vary the EPAP as well, looking to address apneas. You're gonna need two totally separate algorithms, and they can't interfere with each other.

There are a couple of options available with EPR. It can be used during the ramp period only, which would offer significant patient comfort during a time where the perhaps the greatest period of patient difficulty occurs. After the ramp period is over, it returns to the set pressure.

Before you select full-time EPR, though, and carry EPR throughout the night, you should consider how your particular situation might respond to this modality. And I think the key to EPR, AutoCPAP with EPR (should that ever come about) and AutoBiPAP will be how apneas are addressed. You might be OK dealing with hypopneas, snores and RERAs if you're of the belief that BiPAP can properly address these issues.
Consequently, it would seem that as long as an Expiratory Relief Adjunct (ERA) stays in its mileau, then there would be no issues, but if the EPA changes therapy by dropping therapeutic CPAP and creating Pressure Support, then that potential may reveal itself.
You Kids Have Fun!!

JV1967
Posts: 178
Joined: Wed Mar 11, 2015 7:36 am
Location: Northeastern USA

Re: Complex sleep apnea and EPR

Post by JV1967 » Sat Mar 21, 2015 2:09 pm

Pugsy wrote:
Delta4 wrote:I assume you're referring to my recent post where I said:
I have 4-5 bad nights each week (50-80 central events with a AHI of 4.5 - 8.) On my better nights, my AHI is less than 1. That leaves me with a monthly AHI of around 3.
No, I actually wasn't thinking of you or anyone recently in particular. I was thinking of past newbies that I have had to reassure that a few to several centrals don't mean that they have complex sleep apnea. I have had people PM me and want to know about getting an ASV machine because they developed 3 centrals and didn't have any during the sleep study.
Way too much "OMG, I have centrals" panic.

I only recently read your thread and have had other irons on the fire and really didn't have much to offer that would have been different from what was already offered.

When we have more bad nights than good nights and we feel worse than without CPAP therapy...something is changing or not right. Gotta try to figure out what before trying to fix it though..and that's sometimes really difficult.

Rather than muck up this thread...I will go back an read your thread again to see if maybe I missed a question to ask.
My sleep doctor told me she's going to put me on an ASV, because I have "slightly worse onset and waking centrals than normal." She scared the p**p out of me while showing me my report and pointing to it; she is mailing the report to me. My central score was 5.1, is that low? I know very little about all of this stuff, as I am a total noob. She told me I have complex apnea. The overnight tech freaked me out a bit too, during the study, when he said "One central is too many centrals. You have complex apnea."
AirCurve10 ASV Bilevel, with Humidifier. F&P Simplus FFM. Using Sleepyhead software.