Hello,

I saw some discussion on another post regarding intentional and unintentional leaks. One user defined the intentional leak as the venting that takes place from my full face mask
and therefore the leak they would expect to take place. My question therefore becomes, when I look at my data, how can I know what part of the leak data is the right amount to
be okay with?

Image

Does it depend upon the set pressure?

Thanks in advance.

Glen

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: Intentional vs. Unintentional Leaks

Post by robysue » Mon Jul 29, 2013 11:07 pm

Glen,

You can't tell what the expected or intentional leak rate is the leak rate built into the design of the mask so that you do not rebreathe the CO2 you exhale. You can feel the leak that is built in to your mask by putting your hand in front of the vents when the mask is on.

You cannot tell the expected leak rate from only the Total Leak Rate. And you gave me only the summary figures for Total Leak Rate.

To best determine the intentional leak rate you need to look in the owners manual for the mask you are using. There should be a chart or graph in it that gives the intentional or expected leak rate at different pressures.

If someone gives me both their unintentional leak rate data and the total leak rate data, then it's easy to calculate an approximate intentional leak rate:

Intentional leak rate = Total leak rate - unintentional leak rate

Note 1: The + is really only an approximation.

Note 2: In SleepyHead the phrase Leak Rate means the unintentional leak rate. The phrase Total Leak Rate is the total leak rate.

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

User avatar
Sheriff Buford
Posts: 4111
Joined: Mon Aug 09, 2010 8:01 am
Location: Kingwood, Texas

Re: Intentional vs. Unintentional Leaks

Post by Sheriff Buford » Tue Jul 30, 2013 5:13 am

Please place your equipment in your profile so we know what stuff you are using. I suspect you are using a Resmed machine because Resmed reports 95 percentile while Philips Respironics reports in the 70's, and has a different method of reporting leaks. Resmed can compensate leak rates up to 24 l/min. and... will not report the mask vent leak rate used to remove CO2.

How's that, Pugs??

Sheriff

User avatar
Pugsy
Posts: 65416
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Intentional vs. Unintentional Leaks

Post by Pugsy » Tue Jul 30, 2013 6:42 am

Sheriff Buford wrote: How's that, Pugs??
Good effort...but wrong conclusion. Kudos for trying though.

He has a Respironics machine and not a ResMed and BTW not all Respironics show 70 L/min 95% numbers.

How do I know it's a Respironics?...The RERA shown on the statistics is a dead give away. Respironics flags RERAs...ResMed doesn't so it's for sure a Respironics machine.
The 95% thing that SleepyHead is showing...that's the default number which can be changed to 90% which is what Respironics reports but most people don't bother and it's such a small difference that I don't care. It's still the same at or below definition. With people tending to forget the "or below" part of the definition.


So in this situation the leaks are totally acceptable and I suspect that excess leaks are most likely minimal to none depending on pressures. Also people need to remember that when auto adjusting pressures are used that the leak line is going to move with the pressure movement...and doesn't necessarily mean excess leak. It might just be pressure increases causing the increase venting.
Average total leaks at about 36 L/min....that's going to be real close to the vent rate at whatever pressure is used.
I bet the leak line doesn't vary very much and if using Encore to report...it will look almost flat.
glen4cindy wrote:Does it depend upon the set pressure?
Yes, intentional leak/vent rate depends on pressures and mask used. More pressure is going to increase the vent rate. Also full face masks at the same pressure as a nasal pillow mask will also typically have a larger vent rate.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Pugsy
Posts: 65416
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Intentional vs. Unintentional Leaks

Post by Pugsy » Tue Jul 30, 2013 6:51 am

Anyone wondering....in EncoreBasic there is a way to change the default Total Leak report to show only Excess Leak like the S9 reports.
It's not going to be perfectly exact because the machine doesn't know which mask you are using but instead takes sort of a generic reading at the beginning of the night when leaks are most likely not an issue and bases calculations on that guesstimate.
If you are using higher pressures it doesn't mean that your baseline is always at zero though...I don't often go into this explanation because of that fact.
People assume that the baseline for excess leak has to be 0.0 L/min and it isn't always the case. Gets them all upset.
I have a friend using 16.5 cm pressure...her baseline when reporting excess leak only is around 20 L/min. It simply won't/can't go below that line and we know the leaks are very well controlled. Her same total leak line is around 55 with her FFM and her pressure.
So people have to base excess leak on their baseline and not necessarily 0.0. It's hard for people to comprehend sometimes and gets them upset thinking they never have leaks well controlled and that's why I don't go into pushing changing that setting.
Instead I tell people to look at the line...if it is pretty stable it doesn't matter what the numbers say.

And remember...if using auto adjusting pressures that leak line is going to move somewhat as the pressures increase.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Sheriff Buford
Posts: 4111
Joined: Mon Aug 09, 2010 8:01 am
Location: Kingwood, Texas

Re: Intentional vs. Unintentional Leaks

Post by Sheriff Buford » Tue Jul 30, 2013 7:28 am

Dang...

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: Intentional vs. Unintentional Leaks

Post by robysue » Tue Jul 30, 2013 8:46 am

Pugsy wrote:Anyone wondering....in EncoreBasic there is a way to change the default Total Leak report to show only Excess Leak like the S9 reports.
It's not going to be perfectly exact because the machine doesn't know which mask you are using but instead takes sort of a generic reading at the beginning of the night when leaks are most likely not an issue and bases calculations on that guesstimate.
Actually Pugsy I think it's doing some kind of statistical analysis to decide what it thinks is the intentional leak.

And the Resmed machine does it's own guessing. You choose the type of mask and the S9 uses a generic figure for that type of mask. But different masks of the same type do not have exactly the same expected leak curve. However, the differences between the individual mask's expected leaks is (properly) assumed to be statistically insignificant.
If you are using higher pressures it doesn't mean that your baseline is always at zero though...I don't often go into this explanation because of that fact.
People assume that the baseline for excess leak has to be 0.0 L/min and it isn't always the case. Gets them all upset.
I have a friend using 16.5 cm pressure...her baseline when reporting excess leak only is around 20 L/min. It simply won't/can't go below that line and we know the leaks are very well controlled. Her same total leak line is around 55 with her FFM and her pressure.
Asking for clarification: What do you mean when you say her baseline for reporting excess leak is only around 20 L/min? And her total leak line is around 55 L/min?

My guess is that you're saying that her total leak line hovers right around 55 L/min, but that SH shows a constant (excess) leak rate of around 20 L/min that parallels the stable total leak line? Most FFM user guides that I've seen show an expected leak rate of around 50 L/min for a pressure of 16.5: So a stable leak line at 55 L/min would indeed indicate no problems with leaks. But if SH's stats were correctly figuring the (unintentional or excess) leak rate, then she ought to have a leak graph that is hovering right around 0.0 instead of hovering around 20 L/min. Under the CPAP tab of the SH preferences panel, does she have "statistical model" selected for leak calcs? If she's got that option selected and her leak line is hovering around 20 L/min, she ought to submit a bug report through the SleepyHead webpages at sourceforge.net.

And it's worth pointing out here for others: The user manual for your mask will contain a chart or a graph of the expected (intentional) leak rates for the mask at different presssures. But even here, the numbers are not absolute. The reason? Those published intentional (expected) leak rates are statistical data. They should have +/- margins of error published with them; some masks' user guides do include the +/- margins of error, but others don't include that info. Because of this, the published expected leak rates should be interpreted as "roughly where you should see a mostly flat total leak line". But the total leak line might be a bit above or the published expected leak rate:
  • Pugsy's friend's mask's user guide says she should expect a total leak line somewhere near 50 L/min and her leak line winds up at 55 L/min, which is a bit above the published expected leak rate.
  • The Swift FX user manual indicates that at my pressures the expected leak rates should range from 20 L/min (at min EPAP=4) to 29 L/min (at max IPAP=8) and in practice my leaks are usually stable around 15-16 L/min at pressures of 6/4 and 18-20 L/min at pressures of 8/6, which is a bit below the expected leak rate.
But both Pusgy's friend and myself have stable total leak lines that are near the published expected (intentional) leak rate for our masks and our pressures. So the data indicates for both of us that leaks are well controlled.
So people have to base excess leak on their baseline and not necessarily 0.0. It's hard for people to comprehend sometimes and gets them upset thinking they never have leaks well controlled and that's why I don't go into pushing changing that setting.
Instead I tell people to look at the line...if it is pretty stable it doesn't matter what the numbers say.
Agreed. But typically both Encore and SH should do a pretty good job statistically speaking of picking a decent estimated leak rate for the Excessive Leak (Encore) and Leak (SH) graphs.

For example, on a night where my baseline total is pretty close to 15, SH will start showing small bumps in the Leak graph if my total leak gets up near 18 L/min; but on a night where my baseline tota is pretty close to 19, places where the total leak drops to 18 don't show up as bumps in the Leak graph.

For me I find that subtracting the median total Leak from the median Total Leak gives a very good estimate of my expected leak for the night. But my leaks are well controlled.
And remember...if using auto adjusting pressures that leak line is going to move somewhat as the pressures increase.
Total leak line will DEFINITELY go up as pressure goes up. The Leak line (SH) and Excess Leak line (Encore) may or may not go up; it all depends on the statistics are being done AND how fast and how far the pressure increases. For my small 2cm increases in pressure I often see a rise in the Total Leak line in SH and no rise in the Leak line. In looking at other's data here, sometimes I see rises in both graphs and sometimes the rise is only in the Total Leak graph. It all depends on how fast and how much the pressure increased.

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

glen4cindy
Posts: 19
Joined: Thu Mar 08, 2007 10:36 pm

Re: Intentional vs. Unintentional Leaks

Post by glen4cindy » Tue Jul 30, 2013 4:22 pm

Here is a report from Encore. I think it is the section you are asking about.

I'm using a F & P FlexiFit 432 Full Face Mask.

Image

glen4cindy
Posts: 19
Joined: Thu Mar 08, 2007 10:36 pm

Re: Intentional vs. Unintentional Leaks

Post by glen4cindy » Tue Jul 30, 2013 4:34 pm

I also found this information about my mask from the manufacture website.

At my pressure of 14, the exhaust flow from the mask should be about 40 if I read that correctly.

Image

User avatar
Pugsy
Posts: 65416
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Intentional vs. Unintentional Leaks

Post by Pugsy » Tue Jul 30, 2013 5:04 pm

Yep, about 40 L/min give or take a little. There's always a little plus or minus variation.
Your leak line graph...as I suspected pretty much a straight line with only a few little blips. I can guarantee you that yours is a whole lot better looking than mine was for last night.
Good job. Whatever you are doing...don't change a thing.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: Intentional vs. Unintentional Leaks

Post by robysue » Tue Jul 30, 2013 5:10 pm

As pugsy put it: That's a beautiful leak line. Mostly flat (except for the small bump near the end) and close to the expected leak rate of 40 L/min.