S9 VPAP Auto Question

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Sam_01
Posts: 7
Joined: Thu Jul 04, 2013 9:33 am
Location: CT

S9 VPAP Auto Question

Post by Sam_01 » Thu Jul 04, 2013 1:19 pm

Just replaced the S8 VPAP Auto with the S9.On the first night of use I checked the results screen and realized the exhale and inhale pressures were running the same (no difference in the pressures).

When I asked the Respiratory Tech about this I was told that everything was fine and that they were only the same now because I was awake but if I were asleep and needed treatment for an apnea they would be different.

I told her that with my S8, there was a setting called Pressure Support (PS) and this was set at 4.0 and as a result, the exhale pressure always ran 4.0 cm less than the inhale pressures. Also, the PS setting on the new machine was set at 0.0 and asked her if this was the reason there was no differential between the inhale and exhale pressures on the new machine. She responded that the PS was only required if you needed additional pressure support.

Personally I think she is incorrect and that the PS setting determines the differential but I'm not the expert.

Is there anyone out there that can tell me the function of the PS setting on the S9 unit and whether it needs to be set in order to take advantage of the BiPap functionality?

Thanks,

Sam_01

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Been doing this for 29 years

User avatar
kaiasgram
Posts: 3569
Joined: Sat Jun 02, 2012 2:08 pm
Location: Northern California

Re: S9 VPAP Auto Question

Post by kaiasgram » Thu Jul 04, 2013 1:37 pm

Sam you are correct. Your PS is set at 0.0 so your EPAPand IPAP will always be the same even if they increase or decrease. And yes on our machine The PS sets pressure support at a fixed number which will be the same even as EPAP and IPAP auto adjust during the night. If PS is set at 3, the EPAP/IPAP difference will always be 3 between EPAP and IPAP.

Do you have the clinical manual for your S9 VPAP auto? Do you need one?

_________________
Machine: AirSense 10 AutoSet with Heated Humidifer + Aifit N30i Nasal Mask Bundle
Additional Comments: SleepyHead-now-OSCAR software on Mac OSX Ventura

User avatar
Sam_01
Posts: 7
Joined: Thu Jul 04, 2013 9:33 am
Location: CT

Re: S9 VPAP Auto Question

Post by Sam_01 » Thu Jul 04, 2013 1:43 pm

Kaiasgram,

Thaks for the very concise response.

No, I do not have the clinical manual and yes, I would love to have one.


Sam_01

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Been doing this for 29 years

User avatar
kaiasgram
Posts: 3569
Joined: Sat Jun 02, 2012 2:08 pm
Location: Northern California

Re: S9 VPAP Auto Question

Post by kaiasgram » Thu Jul 04, 2013 1:47 pm

Sam_01 wrote:Kaiasgram,

Thaks for the very concise response.

No, I do not have the clinical manual and yes, I would love to have one.


Sam_01
I'll send you a PM when I get home in a little while. Check your Inbox later today.

_________________
Machine: AirSense 10 AutoSet with Heated Humidifer + Aifit N30i Nasal Mask Bundle
Additional Comments: SleepyHead-now-OSCAR software on Mac OSX Ventura

User avatar
RandyJ
Posts: 1673
Joined: Thu Apr 21, 2011 7:22 pm
Location: Connecticut, USA

Re: S9 VPAP Auto Question

Post by RandyJ » Thu Jul 04, 2013 1:51 pm

How can an RT be so stupid? They don't even understand the machines their companies are selling.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Backup & Travel Machines: PR System One Bipap Auto, S9 VPAP Auto, S9 Autoset, Oximeter CMS-50E
Diagnosed March 2011, using APAP 14 - 16.5 cm, AFlex+ 2
Alt masks Swift FX pillows, Mirage FX nasal mask, Mirage Quattro full face mask

User avatar
jdm2857
Posts: 2982
Joined: Wed Aug 20, 2008 8:29 pm
Location: South Jersey

Re: S9 VPAP Auto Question

Post by jdm2857 » Thu Jul 04, 2013 2:47 pm

You should think about having a discussion with whomever employs this RT.

Doctors are prescribing auto bilevel machines and her lack of knowledge is turning them into APAPs.

She needs either education or a new career.
jeff

User avatar
kaiasgram
Posts: 3569
Joined: Sat Jun 02, 2012 2:08 pm
Location: Northern California

Re: S9 VPAP Auto Question

Post by kaiasgram » Thu Jul 04, 2013 3:18 pm

Sam, you can check your private message inbox now.

Kaiasgram

_________________
Machine: AirSense 10 AutoSet with Heated Humidifer + Aifit N30i Nasal Mask Bundle
Additional Comments: SleepyHead-now-OSCAR software on Mac OSX Ventura

User avatar
DoriC
Posts: 5214
Joined: Sat Sep 13, 2008 9:28 pm
Location: NJ

Re: S9 VPAP Auto Question

Post by DoriC » Thu Jul 04, 2013 3:29 pm

Sam_01 wrote:Just replaced the S8 VPAP Auto with the S9.On the first night of use I checked the results screen and realized the exhale and inhale pressures were running the same (no difference in the pressures).

When I asked the Respiratory Tech about this I was told that everything was fine and that they were only the same now because I was awake but if I were asleep and needed treatment for an apnea they would be different.

I told her that with my S8, there was a setting called Pressure Support (PS) and this was set at 4.0 and as a result, the exhale pressure always ran 4.0 cm less than the inhale pressures. Also, the PS setting on the new machine was set at 0.0 and asked her if this was the reason there was no differential between the inhale and exhale pressures on the new machine. She responded that the PS was only required if you needed additional pressure support.

Personally I think she is incorrect and that the PS setting determines the differential but I'm not the expert.
Is there anyone out there that can tell me the function of the PS setting on the S9 unit and whether it needs to be set in order to take advantage of the BiPap functionality?
Thanks,
Sam_01
It amazes me that this RT had all the quick answers even though all wrong! I wonder if there is any training involved at these DME's? I wish she had said she didn't know the answers but would find someone who could help. Wishful thinking!

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: 14/8.4,PS=4, UMFF, 02@2L,
"Do or Do Not-There Is No Try"-"Yoda"
"We are what we repeatedly do,so excellence
is not an act but a habit"-"Aristotle"
DEAR HUBBY BEGAN CPAP 9/2/08

User avatar
Sir NoddinOff
Posts: 4189
Joined: Mon May 14, 2012 5:30 pm
Location: California

Re: S9 VPAP Auto Question

Post by Sir NoddinOff » Thu Jul 04, 2013 4:03 pm

Whew, this is appalling. For once, I am speechless (text-less?).

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2.
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.

User avatar
DoriC
Posts: 5214
Joined: Sat Sep 13, 2008 9:28 pm
Location: NJ

Re: S9 VPAP Auto Question

Post by DoriC » Thu Jul 04, 2013 8:30 pm

Sorry, I forgot to say that you're correct about the PS. Let's say your Epap is 10 with PS=4,then your Ipap will be 14. If your Ipap is 16 and your Epap increases to 11 then your Ipap increases to 15, 12 increases Ipap to 16, always difference of 4. Good luck and keep us posted.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: 14/8.4,PS=4, UMFF, 02@2L,
"Do or Do Not-There Is No Try"-"Yoda"
"We are what we repeatedly do,so excellence
is not an act but a habit"-"Aristotle"
DEAR HUBBY BEGAN CPAP 9/2/08

alancalan
Posts: 132
Joined: Wed Dec 27, 2006 4:27 am

Re: S9 VPAP Auto Question

Post by alancalan » Mon Jul 08, 2013 1:58 pm

Kaiasgram,

I need the Clinical Manual too and the software also, if you have a link.

Alan

_________________
MachineMask
Additional Comments: also have mirage activa 2nasal masks but using the p10

User avatar
kaiasgram
Posts: 3569
Joined: Sat Jun 02, 2012 2:08 pm
Location: Northern California

Re: S9 VPAP Auto Question

Post by kaiasgram » Mon Jul 08, 2013 2:08 pm

alancalan wrote:Kaiasgram,

I need the Clinical Manual too and the software also, if you have a link.

Alan
Hi Alan -- Can you clarify: Your icon shows a ResMed humidifier but you identified you have a Respironics bipap machine. Which clinician manual do you need?

_________________
Machine: AirSense 10 AutoSet with Heated Humidifer + Aifit N30i Nasal Mask Bundle
Additional Comments: SleepyHead-now-OSCAR software on Mac OSX Ventura

User avatar
jdm2857
Posts: 2982
Joined: Wed Aug 20, 2008 8:29 pm
Location: South Jersey

Re: S9 VPAP Auto Question

Post by jdm2857 » Mon Jul 08, 2013 2:54 pm

In this thread Alan indicates that he is looking for the S9 Clinical Manual.

viewtopic/t90667/posting.php?mode=quote&f=1&p=833858

I sent him the link for SleepyHead (I think it's what he wants), but do not have the S9 info.
jeff

User avatar
kaiasgram
Posts: 3569
Joined: Sat Jun 02, 2012 2:08 pm
Location: Northern California

Re: S9 VPAP Auto Question

Post by kaiasgram » Mon Jul 08, 2013 4:42 pm

jdm2857 wrote:In this thread Alan indicates that he is looking for the S9 Clinical Manual.

viewtopic/t90667/posting.php?mode=quote&f=1&p=833858

I sent him the link for SleepyHead (I think it's what he wants), but do not have the S9 info.
Thanks jdm

Alan, I sent you a private message with the link you need to access the clinician manual. You need to do a bit of scrolling down that page but you'll find what you need.

_________________
Machine: AirSense 10 AutoSet with Heated Humidifer + Aifit N30i Nasal Mask Bundle
Additional Comments: SleepyHead-now-OSCAR software on Mac OSX Ventura