idamtnboy wrote:yarrow, what mask do you have selected in the clinician's menu of the S9?
Away from the machine right now, so can't check -- but I have looked in the past and believe it to be correct.
idamtnboy wrote:Looking at the graph I think you have the wrong mask selected. Mathematically the area between the curve and the X axis should be equal on both sides, i.e. the volume of air, which is the area under the curve above the zero line, has to be equal to the volume of air exhaled, the area below the zero line. The zero line should be somewhat higher than it is. It must be displaced due the wrong mask being selected in the S9, or there is a problem with Sleepyhead.
I have a vague recollection of reading a post here that the weirdly-center graphs are a known flaw either of SleepyHead or the S9 itself.
idamtnboy wrote:You do know the S9 subtracts the mask vent from the total flow to get the volume breathed, as well as leak rate, right? Respironics, and maybe others, do not.
What kind of leak rates are you having? How about looking at the data with Resscan? What does it show? In a case like this I would be inclined to trust Resscan to give a more accurate picture of what's happening.
Mouth breathing will show up as leak rate. It won't offset the flow rate graph.
Yes, understood about S9 leak reports. For the period of these flow graphs, leak is low (for me, at least!): 5 to 8 L/min. I do have Resscan set up on a virtual machine -- will check tonight.