There is a good reason why there is a clinical menu included in CPAP / BIPAP units. It is for QUALIFIED clinicians who have been formally trained in the delivery of pressure. Those of you who are not clinicians and feel you know what you are doing are flirting with trouble. When you had you first formal sleep study, ( Some did not and bought their machine on the "black Market") a specific pressure was prescribed to eliminate your apneas / hypopneas. If the pressure is changed without being examined by the M.D., the result could be fatal. Let me explain why. If you have COPD / Emphysema with Blebs and you increase your pressure yourself, you could easily rupture the bleb and die. If you feel you are getting too much pressure and decide to decrease it then you are not being treated for your apnea. It is important to call your doctor with pressure issues instead of self treating. I wear CPAP myself and understand the challenges it brings but decreasing, or even increasing the set pressures are not the answer. I only posted this because I see alot of posts claiming someone has discovered the clinical menu and made changes. The only changes you should make are in the patient menu and relate to ramp time, C-Flex, A-Flex, etc... which doesn't compromise therapy.

User avatar
GumbyCT
Posts: 5780
Joined: Fri Sep 14, 2007 6:22 pm
Location: CT
Contact:

Re: Assessing Clinical Menu on CPAP / BIPAP Units

Post by GumbyCT » Sat Dec 04, 2010 12:53 pm

resp1962 wrote:There is a good reason why there is a clinical menu included in CPAP / BIPAP units. It is for QUALIFIED clinicians who have been formally trained in the delivery of pressure. Those of you who are not clinicians and feel you know what you are doing are flirting with trouble. When you had you first formal sleep study, ( Some did not and bought their machine on the "black Market") a specific pressure was prescribed to eliminate your apneas / hypopneas. If the pressure is changed without being examined by the M.D., the result could be fatal. Let me explain why. If you have COPD / Emphysema with Blebs and you increase your pressure yourself, you could easily rupture the bleb and die. If you feel you are getting too much pressure and decide to decrease it then you are not being treated for your apnea. It is important to call your doctor with pressure issues instead of self treating. I wear CPAP myself and understand the challenges it brings but decreasing, or even increasing the set pressures are not the answer. I only posted this because I see alot of posts claiming someone has discovered the clinical menu and made changes. The only changes you should make are in the patient menu and relate to ramp time, C-Flex, A-Flex, etc... which doesn't compromise therapy.
You can deliver my pressure. I need it today, please. All my trained pressure people were OFF yesterday, I was told to wait until Mon.

Now I will wait for CM to deliver the popcorn. I was wondering what I was gonna do today.

_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: New users can't remember they can't remember YET!
BeganCPAP31Jan2007;AHI<0.5
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember
;)
If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!

User avatar
drj130
Posts: 565
Joined: Mon Jun 15, 2009 1:14 pm
Location: Florida

Re: Assessing Clinical Menu on CPAP / BIPAP Units

Post by drj130 » Sat Dec 04, 2010 1:13 pm

There is a good reason why there is a clinical menu included in CPAP / BIPAP units. It is for QUALIFIED clinicians who have been formally trained in the delivery of pressure.
Yeah, let me run down to the corner and see the local DME for pressure changes.

As a Soldier stationed in Germany, I have to go through hell to get anything for my machine. I see everyone posting about getting too much in supplies, being able to call up their insurance company and complain about problems and then getting them fixed, about being able to switch DMEs. Because I am a Soldier, I have to go through the people that Tri-Care WANTS me to go through. I have so many freaking channels that I have to deal with just to get supplies it isn't funny. I have had my machine for over a year and a half, deployed with it to Afghanistan, and everything that I was using until just the other day was either from when I first got the machine or I paid for out of my own pocket with NO hope of getting paid back. The ONLY reason that I was able to last so long on what I had was that there were people out there who sent me masks and filters at no charge. The only reason that I am doing well with my treatment is that I have taken charge of my care. Is my doctor aware of this, YES. Is he upset that I've made changes, NO. He is happy that I have what I need to ensure that I'm getting the best treatment that I can get and that I have taken charge. I've seen this person once in the whole time that I've been on it.

I will keep monitoring MY treatment and making the needed adjustments as I see fit to ensure that I live as long as I can.

David

_________________
Mask: ComfortGel Blue Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Wisp Mask, Pressure is set between 8.5 and 12.5. With a M Series Auto as a backup. Also a CMS-50D+ PulseOX
Look twice, save a life. You never know when you might see a motorcyclist.

resp1962
Posts: 6
Joined: Sat Dec 04, 2010 11:14 am

Re: Assessing Clinical Menu on CPAP / BIPAP Units

Post by resp1962 » Sat Dec 04, 2010 1:34 pm

Dear drj130,

I did not mention anything in my post stating that people should not use different masks, hoses, and filters. I said it is dangerous for non-clinical to change pressures in the clinical menu of their machines. I respect you as a patient, a soldier and for defending our country. I realize that going to the doctor is not easy for you. If you have pressure issues with your machine and can contact me with pressure or supply issues, I will help in getting them to you.
Last edited by resp1962 on Sun Dec 05, 2010 11:14 am, edited 1 time in total.

User avatar
Slinky
Posts: 11372
Joined: Wed Nov 01, 2006 3:43 pm
Location: Mid-Michigan

Re: Assessing Clinical Menu on CPAP / BIPAP Units

Post by Slinky » Sat Dec 04, 2010 3:54 pm

Ah, Resp1962, welcome to the forum and thanks for your input - but - while you can ENCOURAGE us to rely on our doctors and RRTs, "please" don't hand us the horsepuckies about changing our therapy settings ourselves. Most of us in this forum found our way here BECAUSE our sleep professionals either let us down or just plain screwed us.

I have COPD. I have a history of a spontaneous pneumothorax in the past - yet I had a sleep doctor script me a loaner APAP set at 4-20 and w/no humidifier to boot. And, yes, he was and is aware of my SP history.

Shortly after getting a new mask my cat grabbed it while I was making the bed and I dropped it and the elbow broke as well as the cat's claw put a small puncture in the cushion. The DME provider's RRT told me they didn't carry extra cushions or elbows and I would just have to wait until I was due for a new mask!!!

Needless to say, that DME provider is no longer MY local DME provider nor is that sleep doctor MY sleep doctor anymore. But it takes TIME to get an appointment w/a new sleep doctor. FORTUNATELY, I had the members of cpaptalk to bail me out.

So - we appreciate your sharing your advice and support w/us - but please be sure it is support and advice and not propaganda.

And I would suggest that you allow PMs and invite members who would like some assistance to PM you for advice. Posting your e-mail addy in a public forum isn't all the great an idea.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: PR SystemOne BPAP Auto w/Bi-Flex & Humidifier - EncorePro 2.2 Software - Contec CMS-50D+ Oximeter - Respironics EverFlo Q Concentrator
Women are Angels. And when someone breaks our wings, we simply continue to fly.....on a broomstick. We are flexible like that.
My computer says I need to upgrade my brain to be compatible with its new software.

flash 2

Re: Assessing Clinical Menu on CPAP / BIPAP Units

Post by flash 2 » Sat Dec 04, 2010 4:18 pm

Slinky is right. I came here because of bad dme, and sleep, pulmonologist, dentist and allergist that couldn't agree on the right pressure. Sleep study was useful but it wasn't the end all. I found my doc now listens to me and lets me tell him what's going on. Most of the sleep docs I saw were actually poorly versed in dealing with my situation besides taking forever to see. I am more than happy to follow a doc who is accurate, timely and thorough. I'd like the sun, moon and starts too. Besides, I went to school with most of these folks, I know I'm more intelligent than many of them. It's not that hard to understand or learn.

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: Assessing Clinical Menu on CPAP / BIPAP Units

Post by robysue » Sat Dec 04, 2010 5:45 pm

resp1962 wrote: The only changes you should make are in the patient menu and relate to ramp time, C-Flex, A-Flex, etc... which doesn't compromise therapy.
And then of course there are plenty of folks whose machines were set up in such a way that they, the patient had no access to C-Flex, A-Flex, ramp, etc. that don't affect treatment. To set the time correctly on my machine you need the clinicians menu. The way mine was set up originally, the EPR (Resmed's A-Flex) was set to 3 and unchangable by me.

Just cause we know about the clinician's menu doesn't necessary we muck around with the pressure settings willy-nilly.

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

User avatar
GumbyCT
Posts: 5780
Joined: Fri Sep 14, 2007 6:22 pm
Location: CT
Contact:

Re: Assessing Clinical Menu on CPAP / BIPAP Units

Post by GumbyCT » Sat Dec 04, 2010 6:03 pm

robysue wrote:
resp1962 wrote: The only changes you should make are in the patient menu and relate to ramp time, C-Flex, A-Flex, etc... which doesn't compromise therapy.
Just cause we know about the clinician's menu doesn't necessary we muck around with the pressure settings willy-nilly.
Speak for yourself....IF I can't go dial wingin what will I do during the time period I am NOT sleeping? Seems like a waste of time not to be changing something, eh?

_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: New users can't remember they can't remember YET!
BeganCPAP31Jan2007;AHI<0.5
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember
;)
If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!

User avatar
toddlorensinclair
Posts: 5
Joined: Sat Dec 04, 2010 10:21 am
Location: Westernport, Md
Contact:

Re: Assessing Clinical Menu on CPAP / BIPAP Units

Post by toddlorensinclair » Sat Dec 04, 2010 7:12 pm

robysue wrote:
resp1962 wrote: The only changes you should make are in the patient menu and relate to ramp time, C-Flex, A-Flex, etc... which doesn't compromise therapy.
And then of course there are plenty of folks whose machines were set up in such a way that they, the patient had no access to C-Flex, A-Flex, ramp, etc. that don't affect treatment. To set the time correctly on my machine you need the clinicians menu. The way mine was set up originally, the EPR (Resmed's A-Flex) was set to 3 and unchangable by me.

Just cause we know about the clinician's menu doesn't necessary we muck around with the pressure settings willy-nilly.
I agree with Robysue ... there are several items on the clinicians menu that are useful ... including detailed information. I can't see changing a pressure setting ... but I can see turning off c-flex which isn't in the user menu on my unit. If you're having a problem adjusting to the unit turning off the c-flex and trying it that way is one way to eliminate a cause.

_________________
Mask
Additional Comments: Backup is a respironics aria lx straight cpap with resmed heated humidifier setting is 13

brazospearl
Posts: 704
Joined: Sun Nov 15, 2009 5:51 pm

Re: Assessing Clinical Menu on CPAP / BIPAP Units

Post by brazospearl » Sat Dec 04, 2010 8:13 pm

resp1962, honey, bless your heart! While we completely appreciate that you have our best interests in mind, your credibility is a bit tarnished since, despite your formal training and expertise, you don't seem to know that your statements were not accurate. Since most of us don't live in Utopia, where all the sleep medical professionals are helpful, accessible, affordable, and know all the nuances of every aspect of successful therapy, we've decided to take matters into our own hands. You're so sweet to be concerned, but if I accidently die by increasing my pressure nobody on this board is responsible. My health, my therapy, my decisions, my responsibility.

User avatar
Willtry
Posts: 40
Joined: Fri Nov 19, 2010 5:48 pm
Location: South Australia

Re: Assessing Clinical Menu on CPAP / BIPAP Units

Post by Willtry » Sat Dec 04, 2010 8:40 pm

I too am new to this forum.
The one thing I can say with out contradiction is that the people on this forum support you.
The basic and most important form of support supplied here is MORAL support.
It is not easy to start therapy and learn to sleep with a piece of plastic strapped to your head it is very daunting.
The people here encourage you to keep at it, they encourage your to follow guide lines, they educate you about how what when and where with out pre-justice on what colour race or creed.
They give you more details about what is happening and why, they do not ask for anything just give and give which is more than a lot of money hungry GP's and other medical support do.

If you where to add up the total years of Knowledge accumulated from experience on this forum it will exceed the total number of hours most GP’s receive on the condition.

That is all I wish to say.

Life is like a box of chocolates you never know what you will get.

_________________
Mask
Additional Comments: F&P Infosmart software

Janknitz
Posts: 8517
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: Assessing Clinical Menu on CPAP / BIPAP Units

Post by Janknitz » Sat Dec 04, 2010 10:33 pm

There's been an especially lively discussion about this on Sleep Guide.com and I agree with the conclusion--if you don't have something like Emphysema with blebs or other such risk factors, about the only way a regular CPAP or Auto machine can hurt you is if it's thrown at your head.

And maybe you should learn to spell A-C-C-E-S-S-I-N-G before you go preaching to people about whether or not it's OK to do it. I don't think even people with very serious lung disease can be hurt by "assessing" anything.
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

User avatar
linagee
Posts: 120
Joined: Sun Oct 03, 2010 2:32 am
Location: USA

Re: Assessing Clinical Menu on CPAP / BIPAP Units

Post by linagee » Sat Dec 04, 2010 11:24 pm

I could easily create an algorithm to do what a sleep doctor does to titrate to the right pressure for the minimum amount of AHIs. (In fact, they already have that. It's called an APAP. APAPs fail at analyzing multi-day data though.) As for the risks of killing yourself with too much pressure, yes there is that risk. A doctor is covered by insurance to protect from the risks he takes. But everything in life has risk and the doctor does not know *that* much more than a well educated CPAP patient using an APAP. BTW, when they titrate you, they put you at unknown pressures without knowing if it will kill you or not. They don't have some magical device in sleep labs that brings people back to life if they give you too much pressure. There is risk there.

May you all sleep well with low AHI,
linagee

DaveMunson
Posts: 266
Joined: Thu Jun 23, 2005 8:36 pm

Re: Assessing Clinical Menu on CPAP / BIPAP Units

Post by DaveMunson » Sun Dec 05, 2010 12:19 am

resp1962 wrote:QUALIFIED
In order to have people trust what you say, you need to come across more helpful. Your message was not put out well. I hope you stick around enough to learn why we do what we do. It will make you a better professional. As you look at the different stories of why anyone would struggle and change their pressure, well, it may surprise you. It's educational. I would not preach, if I were you, here, without learning more.

_________________
MaskHumidifier
Additional Comments: SleepyHead, ResScan 3.16, ClimatelLne Tube, Hosehuggie in Plaid (it's so cool)

User avatar
GumbyCT
Posts: 5780
Joined: Fri Sep 14, 2007 6:22 pm
Location: CT
Contact:

Re: Assessing Clinical Menu on CPAP / BIPAP Units

Post by GumbyCT » Sun Dec 05, 2010 9:41 am

DaveMunson wrote:
resp1962 wrote:QUALIFIED
In order to have people trust what you say, you need to come across more helpful. Your message was not put out well.
I could be out in the weeds again (or still) but I think instead of kicking the door in and telling a group of people (who have been extremely successful) they are doing things all wrong and you are here to save the day - that an introduction of yourself would fare better.

How have we managed for so long with out you?

See also the quotes Slinky "Flies and Honey" lectures -
search.php?keywords=flies+honey&terms=a ... mit=Search

JMMO - Just MY Meaningless Opinion

_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: New users can't remember they can't remember YET!
BeganCPAP31Jan2007;AHI<0.5
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember
;)
If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!