I'm new to this c-pap thing. I got mine from someone else but they didn't have the manuals. I found the user's manual online, but can't find a provider's manual. Would any of you out there have one for a RemStar pro2 Auto that you'd be willing to share?
Thank-you.
provider's manual
Re: provider's manual
These instructions will get you into the setup menu. If you need the manual, please PM me with your e-mail address and I will send it to you the .pdf file.
Den
REMstar Pro 2 Setup
When in the Setup Menu, the humidifier ^ and ramp v buttons operate as up and down keys to change the settings, the left/right user buttons < > allow you to go to the previous/next question or setting, and the pressure start/stop button is used to exit the Setup Menu. Holding the humidifier or ramp buttons down will cause the values to change more quickly.
To enter the Therapy Setup Menu, hold the two top user buttons < > down while plugging in the power cord.
Continue holding the buttons down until the REMstar Pro 2 with C-Flex beeps twice.
Note: The word "setup" will appear on all of the screens indicating that you are in the Therapy Setup Menu.
If you press the Pressure start/stop button, you will exit the Setup Menu.
a. Compliance hours/nights: (recommend leaving alone, but CAN be cleared at this point)
Select next setting with >
b. Therapy mode: (CPAP or CFLE) select with ^ or v
Select next setting with >
c. CPAP or CFLE pressure setting: Select with ^ or v
Select next setting with >
d. C-Flex mode (if you chose CFLE mode): Select setting 1, 2 or 3 with ^
Select next setting with >
e: Ramp time setting: Select with ^ or v
(ramp time will be turned off with a setting of 0:00)
Select next setting with >
f. Ramp pressure setting: Select with ^ or v
Select next setting with >
g. Patient disconnect setting: 1 = on 0 = off Select with ^ or v
Select next setting with >
h. Buttons lights setting: 1 = on 0 = off Select with ^ or v
(Last setting. Use On/Off button to exit Setup)
Den
REMstar Pro 2 Setup
When in the Setup Menu, the humidifier ^ and ramp v buttons operate as up and down keys to change the settings, the left/right user buttons < > allow you to go to the previous/next question or setting, and the pressure start/stop button is used to exit the Setup Menu. Holding the humidifier or ramp buttons down will cause the values to change more quickly.
To enter the Therapy Setup Menu, hold the two top user buttons < > down while plugging in the power cord.
Continue holding the buttons down until the REMstar Pro 2 with C-Flex beeps twice.
Note: The word "setup" will appear on all of the screens indicating that you are in the Therapy Setup Menu.
If you press the Pressure start/stop button, you will exit the Setup Menu.
a. Compliance hours/nights: (recommend leaving alone, but CAN be cleared at this point)
Select next setting with >
b. Therapy mode: (CPAP or CFLE) select with ^ or v
Select next setting with >
c. CPAP or CFLE pressure setting: Select with ^ or v
Select next setting with >
d. C-Flex mode (if you chose CFLE mode): Select setting 1, 2 or 3 with ^
Select next setting with >
e: Ramp time setting: Select with ^ or v
(ramp time will be turned off with a setting of 0:00)
Select next setting with >
f. Ramp pressure setting: Select with ^ or v
Select next setting with >
g. Patient disconnect setting: 1 = on 0 = off Select with ^ or v
Select next setting with >
h. Buttons lights setting: 1 = on 0 = off Select with ^ or v
(Last setting. Use On/Off button to exit Setup)
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Re: provider's manual
After re-reading your post, I don't know whether you have the Pro 2 or the Auto......they aren't the same machine.
So, here are the instructions for the Auto.
Den
REMstar Auto w/C-Flex Setup
When in the Setup Menu, the humidifier ^ and ramp v buttons operate as up and down keys to change the settings, the left/right user buttons < > allow you to go to the previous/next question or setting, and the pressure start/stop button is used to exit the Setup Menu. Holding the humidifier or ramp buttons down will cause the values to change more quickly.
To enter the Therapy Setup Menu, hold the two top user buttons < > down while plugging in the power cord.
Continue holding the buttons down until the REMstar Auto w/C-Flex beeps twice.
Note: The word "setup" will appear on all of the screens indicating that you are in the Therapy Setup Menu.
(If you press the Pressure start/stop button, you will exit the Setup Menu.)
A. Compliance hours/nights: (recommend leaving alone, but CAN be cleared at this point)
Select next setting with >
B. Therapy mode: (CPAP/CFLE/APAP/AFLE) select with ^ or v
Select next setting with >
B1. If CPAP or CFLE select pressure setting: Select with ^ or v
Select next setting with >
B1a. C-Flex mode (if you chose CFLE mode): Select setting 1, 2 or 3 with ^
Select next setting with >
B2. If APAP or AFLE, select minimum pressure setting: Select with ^ or v
Select next setting with >
B2a. If APAP or AFLE, select maximum pressure setting: Select with ^ or v
Select next setting with >
B2b. C-Flex mode (if you chose AFLE mode): Select setting 1, 2 or 3 with ^
Select next setting with >
C: Ramp time setting: Select with ^ or v
(ramp time will be turned off with a setting of 0:00)
(if using APAP or AFLE mode, Ramp setting MUST be set to 0:00)
Select next setting with >
D. Ramp pressure setting: Select with ^ or v (use only in CPAP or CFLE mode)
Select next setting with >
E. Patient disconnect setting: 1 = on 0 = off Select with ^ or v
Select next setting with >
F. Buttons lights setting: 1 = on 0 = off Select with ^ or v
(Last setting. Use On/Off button to exit Setup)
So, here are the instructions for the Auto.
Den
REMstar Auto w/C-Flex Setup
When in the Setup Menu, the humidifier ^ and ramp v buttons operate as up and down keys to change the settings, the left/right user buttons < > allow you to go to the previous/next question or setting, and the pressure start/stop button is used to exit the Setup Menu. Holding the humidifier or ramp buttons down will cause the values to change more quickly.
To enter the Therapy Setup Menu, hold the two top user buttons < > down while plugging in the power cord.
Continue holding the buttons down until the REMstar Auto w/C-Flex beeps twice.
Note: The word "setup" will appear on all of the screens indicating that you are in the Therapy Setup Menu.
(If you press the Pressure start/stop button, you will exit the Setup Menu.)
A. Compliance hours/nights: (recommend leaving alone, but CAN be cleared at this point)
Select next setting with >
B. Therapy mode: (CPAP/CFLE/APAP/AFLE) select with ^ or v
Select next setting with >
B1. If CPAP or CFLE select pressure setting: Select with ^ or v
Select next setting with >
B1a. C-Flex mode (if you chose CFLE mode): Select setting 1, 2 or 3 with ^
Select next setting with >
B2. If APAP or AFLE, select minimum pressure setting: Select with ^ or v
Select next setting with >
B2a. If APAP or AFLE, select maximum pressure setting: Select with ^ or v
Select next setting with >
B2b. C-Flex mode (if you chose AFLE mode): Select setting 1, 2 or 3 with ^
Select next setting with >
C: Ramp time setting: Select with ^ or v
(ramp time will be turned off with a setting of 0:00)
(if using APAP or AFLE mode, Ramp setting MUST be set to 0:00)
Select next setting with >
D. Ramp pressure setting: Select with ^ or v (use only in CPAP or CFLE mode)
Select next setting with >
E. Patient disconnect setting: 1 = on 0 = off Select with ^ or v
Select next setting with >
F. Buttons lights setting: 1 = on 0 = off Select with ^ or v
(Last setting. Use On/Off button to exit Setup)
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Re: provider's manual
Most of Respironics manuals can be downloaded on line. Go respironics.com and try to get it. Good luck. rjonz
Re: provider's manual
Thanx for all the suggestions.
I still have one question: How do I save the results after I have changed the pressure setting?
I still have one question: How do I save the results after I have changed the pressure setting?
Re: provider's manual
'srjonz wrote:Most of Respironics manuals can be downloaded on line. Go respironics.com and try to get it. Good luck. rjonz
I have never managed to find a provider's manual in respironics.com. Only users' manuals.
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Re: provider's manual
When you make a change to the pressure setting (or any of the others), it IS saved when you press the On/Off button to exit the setup.230jhp wrote:Thanx for all the suggestions.
I still have one question: How do I save the results after I have changed the pressure setting?
Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05