Looking for OSCAR interpretation assistance and Aerophagia

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
kewlpewl
Posts: 4
Joined: Tue Aug 06, 2024 6:27 am

Looking for OSCAR interpretation assistance and Aerophagia

Post by kewlpewl » Tue Aug 06, 2024 7:12 am

Hi, I have been using CPAP for almost 4 years now after an at-home sleep study diagnosis.
It's been going alright I guess, I still feel tired in the mornings but over the past few weeks I have developed significant Aerophagia culminating in an overnight sprint to the bathroom last night thinking I was either going to vomit or defecate after waking up with painful cramping, and learning on arrival it was just air (noted in 3rd attachment).
I would really like to fix this if possible as it's really disrupting my sleep waking up repeatedly to expel gas (sorry I know it's not classy but it's the truth).
Potentially relevant info:
34yo, overweight bmi, using soft cervical collar, medicated for depression (escitalopram) and ADHD (lisdexamfetamine)

Past 3 nights OSCAR data attached
03082024.png
03082024.png (93.75 KiB) Viewed 85258 times
04082024.png
04082024.png (93.03 KiB) Viewed 85258 times
05082024.png
05082024.png (101.12 KiB) Viewed 85258 times
Any other insights not related to the Aerophagia also welcomed, would love to know if my current settings are optimal or not. Ultimately would love to wake up feeling more rested but struggle with cycles of defeatism.

Thankyou in advance,
Lee

User avatar
Pugsy
Posts: 65444
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Looking for OSCAR interpretation assistance and Aerophagia

Post by Pugsy » Tue Aug 06, 2024 7:23 am

Are you experiencing any nasal congestion at night?

How did you end up with the current settings?

Doesn't surprise me at all that the bottom image report that you complain of increased aerophagia issues. You spent a LOT of time with pressures up around 20 and this makes me wonder why ...... sleeping position maybe where the OSA worsens and you need more pressure???? Do you spend much time on your back???? That would be my first thought and it's common for OSA to worsen, in some people, causing higher pressure needs.
I may have to RISE but I refuse to SHINE.

kewlpewl
Posts: 4
Joined: Tue Aug 06, 2024 6:27 am

Re: Looking for OSCAR interpretation assistance and Aerophagia

Post by kewlpewl » Tue Aug 06, 2024 7:44 am

Pugsy wrote: ↑
Tue Aug 06, 2024 7:23 am
Are you experiencing any nasal congestion at night?

How did you end up with the current settings?

Doesn't surprise me at all that the bottom image report that you complain of increased aerophagia issues. You spent a LOT of time with pressures up around 20 and this makes me wonder why ...... sleeping position maybe where the OSA worsens and you need more pressure???? Do you spend much time on your back???? That would be my first thought and it's common for OSA to worsen, in some people, causing higher pressure needs.
Typically no nasal congestion, although I did wake up this morning (which is the day after the 3rd screenshot) a bit under the weather which has progressed to some nasal congestion this evening.

My current settings were advised on a thread I made on apneaboard.com in 2020 after my at home sleep study where I was diagnosed with sleep apnea with 46AHI. My sleep physician at the time wasn't much help and just sent me on my way.
I felt like I was getting ignored in my follow up questions there, and quite possibly rightfully so and also coupled with the aforementioned cycles of defeatism, and thought to try a different community for assistance. My original thread can be found at https://www.apneaboard.com/forums/Threa ... OSCAR-data (I'm sorry if there is a rule i've missed on cross-posting from other forums).

Attached is the original settings given to my by the sleep physician.
Capture.PNG
Capture.PNG (120.39 KiB) Viewed 85244 times
I'm all over the place when it comes to sleeping position, I try to go to sleep on my side (hoping that it will lead to more open airways) but often wake up either on my other side/back/stomach. I also wake up infrequently with numbness/paraesthesia (I hope that's the correct term) when sleeping on my side.

kewlpewl
Posts: 4
Joined: Tue Aug 06, 2024 6:27 am

Re: Looking for OSCAR interpretation assistance and Aerophagia

Post by kewlpewl » Sat Nov 30, 2024 4:28 am

I'm again pleading for help with the aerophagia i'm experiencing. It is waking me up in pain with stomach cramps many times every night. I am really sick of it :'(
27-11-2024.png
27-11-2024.png (172.92 KiB) Viewed 82824 times
28-11-2024.png
28-11-2024.png (174.23 KiB) Viewed 82824 times
29-11-2024.png
29-11-2024.png (172.59 KiB) Viewed 82824 times

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: Looking for OSCAR interpretation assistance and Aerophagia

Post by Miss Emerita » Sat Nov 30, 2024 12:52 pm

I'd say you need to experiment with lower pressures to find the sweet spot between controlling aerophagia and keeping your AHI low

Another forum member has had some experience doing this; I'll alert him that you could use some advice.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
zonker
Posts: 11514
Joined: Fri Jun 19, 2015 4:36 pm

Re: Looking for OSCAR interpretation assistance and Aerophagia

Post by zonker » Sat Nov 30, 2024 1:26 pm

kewlpewl wrote: ↑
Sat Nov 30, 2024 4:28 am
I'm again pleading for help with the aerophagia i'm experiencing. It is waking me up in pain with stomach cramps many times every night. I am really sick of it :'(
sorry you're going through that pain. had a bit of it myself a few years ago. not quite as bad as what you're suffering. but it felt like a large inflated football in my stomach. and not able to relieve it.

here's what I found worked for me. going back thru my Oscar data, I found a time when I wasn't suffering from aerophagia. now, at that particular time, I found that my ahi was very high. and mostly it was obstructive events. and I had thought the only way to get those events down was to increase pressure. and yeah, increasing pressure brought my ahi under control, but what good was that if I couldn't sleep?

so I went back to those settings, whatever they were. sure enough my ahi went up but my aerophagia was well under control. I still had gass, but nothing I couldn't get rid of.

what I decided to do was increase my maximum ahi by .2. I see you have a resmed and this is possible with the machine. Phillips has a minimum increase of a whole point! glad I have a resmed.

so I let that settle in. after a few nights, if all was well, I'd increase by another .2 and leave it for a few more nights. and so on and so forth.

every now and then, I'd increase the maximum and wham! aerophagia reared it's ugly head. so back down I'd go then start all over again.

the major drawback to this is that it takes a LOT of patience. I don't remember how long this took but i'm thinking that it was at least a month. now many years later I have my machine set at 11 min and 25 max. no areophagia. and my ahi last night was 0.4.

I can only recommend it. I can't guarantee it.

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
robysue1
Posts: 1311
Joined: Sun Sep 18, 2022 3:39 pm
Location: Buffalo, NY

Re: Looking for OSCAR interpretation assistance and Aerophagia

Post by robysue1 » Sat Nov 30, 2024 1:57 pm

kewlpewl,

First, I sympathize with you. Severe aerophagia (at relatively low pressures) was my nemesis during my first year or so of CPAPing, and even now, 15 years later, I have occasional problems with it. But for the most part, I can keep it under control.

Next, a bit about what's going on: Aerophagia is caused by air getting into your stomach. You already knew that, of course. But it's important to have an idea of how that air gets into your stomach. One culprit can be a weak esophageal valve that opens and allows air into the stomach. That same weak valve is also the culprit in GERD---when it opens, acid from the stomach can reflux back into the esophagus. Another culprit is excess air making its way into the oral cavity and then swallowing in an effort to get rid of the excess air. That excess air gets into the oral cavity typically gets into the oral cavity when your tip of tongue drops off of its usual (normal) place touching the roof of the mouth just behind the upper incisors while you are breathing through your nose. When that happens, air can enter the oral cavity from the nasal cavity through the back of the mouth. This can lead to "chipmunk cheeks" as the air builds up in the oral cavity. Then you swallow and that air goes into the tummy, where the only way to get rid of it is through burping and farting.

The other way to attempt to get rid of that excess air in the oral cavity is just opening your mouth. However, when you are using a CPAP, "just open your mouth and exhale" isn't as simple as it seems: If you are using a nasal mask, opening your mouth when the tongue is not in the correct position leads to a significant leak, which, in turn causes the machine to blow air at an increased rate in an effort to compensate for the leak and maintain the current pressure setting. And that can make it difficult (impossible?) to just get rid of the excess air by opening your mouth. If you are using a FFM, opening your mouth doesn't automatically trigger a leak, but since the whole area under the mask cushion in pressurized, you still have to force the air out of the mouth---it won't just "rush out" on it's own.

So the first piece of advice is to pay attention (in the daytime) to where the tip of your tongue is when you are breathing through your nose. If you find that it's dropping off the roof of the mouth during the day, then it's most likely dropping down at night and that may be the trigger of the aerophagia. By consciously working on training your tongue to stay in the correct position during the day, it will help encourage the tongue to stay put during the night.

Next: Standard advice for dealing with GERD problems tend to help with aerophagia: Don't eat right before going to bed. Try (if possible) to have your biggest meal far away from bedtime. Don't overeat in general. Avoid sleeping on your right side. (Sleeping on your left side is fine.) If you want, you can try elevating the head of the bed a bit.

Now to look at a few things you specifically mention:
kewlpewl wrote: ↑
Tue Aug 06, 2024 7:44 am
My current settings were advised on a thread I made on apneaboard.com in 2020 after my at home sleep study where I was diagnosed with sleep apnea with 46AHI. My sleep physician at the time wasn't much help and just sent me on my way.
I felt like I was getting ignored in my follow up questions there, and quite possibly rightfully so and also coupled with the aforementioned cycles of defeatism, and thought to try a different community for assistance. My original thread can be found at https://www.apneaboard.com/forums/Threa ... OSCAR-data (I'm sorry if there is a rule i've missed on cross-posting from other forums).
Your current settings are Auto mode with a pressure range of 13-20 and EPR = 3. Your original machine appears to be a PR DreamStation 1 and it looks like your original script was for using Auto mode with a pressure range of 4-20 and you had A-Flex set to either 2 or 3. (I would guess 3.)

My guess is that you tweaked the pressures because when the pressure was set to 4-20, your treated AHI was higher than you liked, there was still a lot of snoring going on, and the PR machine was also scoring some flow limitations and RERAs when the pressure was 4-20; and people were correctly pointing out that a higher minimum pressure would help the machine do a better job of preventing your apneas, suppressing the snoring, flow limitations, and RERAs.

But higher pressures tend to aggravate aerophagia. And on August 5, the night you specify that you woke up with the aerophagia was really painful, you woke up after spending quite a bit of time at or near 20 cm H2O (on inhalation). The most recent data (Nov. 27-29), you didn't spend large amounts of time at 20 cm, but you did have peaks of pressure where you were snoring and/or flow limitations were being scored. And it looks like the snoring and flow limitations are what's driving the increase in pressure from your minimum setting of 13 all the way up to the 17-20 cm range.

So one thing you might want to do is experiment with reducing the pressure even though you may have more flow limitations and/or snoring scored. But you may sleep better with a bit more snoring and flow limitations, but less aerophagia. I would suggest using a pressure range of 13-15cm for a few days and see what happens, particularly with the aerophagia. In other words, you may need to do some experimenting to find what the maximum pressure your stomach can tolerate before the aerophagia kicks in and becomes painful. In the end, you want to find a pressure range that keeps your aerophagia under control, but also is high enough to keep your AHI reasonably decent. (You don't need to aim for an AHI < 1.0 every night.)

You also write:
Potentially relevant info:
34yo, overweight bmi, using soft cervical collar, medicated for depression (escitalopram) and ADHD (lisdexamfetamine)
The overweight bmi may be aggravating things, but your weight is what it is.

Using the soft cervical collar to keep your head from falling forward and/or your jaw from opening may be counter-effective: If the collar increases your discomfort when you are trying to get to sleep, that can lead to (unconscious) swallowing of air and an increase in aerophagia.

You also write:
I'm all over the place when it comes to sleeping position, I try to go to sleep on my side (hoping that it will lead to more open airways) but often wake up either on my other side/back/stomach. I also wake up infrequently with numbness/paraesthesia (I hope that's the correct term) when sleeping on my side.
So you're an active sleeper.

If you want to sleep on your side, try to stay (as much as possible) on your left side. As for back or stomach sleeping: Both are likely to cause less aerophagia than sleeping on your right side.

It's also extremely important when fighting aerophagia to sleep in the position(s) that seem most natural to you (with the exception of sleeping on your right side). Here's why: Sleeping in a position that your body doesn't really want to sleep in can lead to restless sleep and micro-arousals. And it's not uncommon to swallow air when you arouse. Hence minimizing arousals also helps with battling aerophagia.

So here's my overall recommendations:

1) Try limiting the max pressure setting to something like 15 for the time being. Yes, you will have more events, and more snoring. But it's important to find a pressure that doesn't trigger the painful aerophagia. Once you know how high you can set the pressure without triggering the aerophagia, you may be able to slowly start increasing the pressure up in order to get the AHI nice and low and eliminate much of the snoring.

2) Try to avoid sleeping on your right side. Allow yourself to go to sleep in whatever position is most comfortable, except for the right side.

3) Don't eat close to bedtime. Also try to limit or eliminate any alcohol close to bedtime.

4) When you wake up with aerophagia, turn the machine off. Sit up in bed and take the mask off and just breathe for a few minutes. See if moving around in bed helps you burp or fart enough for the stomach to feel better. If not, get out of bed and move around. Sit quietly in a chair in another room while allowing the stomach to deal with the air and feel less bloated. Go back to bed only when the stomach feels better, and put the mask back on and restart the machine. By restarting the machine, you force it to go back to your lowest pressure setting, which will help the stomach.

5) You might also want to consider turning the Auto Ramp on with a beginning ramp pressure of something like 10cm. This will allow the stomach to not deal with your full beginning pressure of 13cm when you are first trying to get back to sleep after every wake. But the Auto Ramp will increase the pressure up to 13cm once you have established a nice, regular, sleep breathing pattern of respiration.
Joined as robysue on 9/18/10. Forgot my password & the email I used was on a machine that has long since died & gone to computer heaven.

Correct number of posts is 7250 as robysue + what I have as robysue1

Profile pic: Frozen Niagara Falls

kewlpewl
Posts: 4
Joined: Tue Aug 06, 2024 6:27 am

Re: Looking for OSCAR interpretation assistance and Aerophagia

Post by kewlpewl » Sat Nov 30, 2024 10:16 pm

Thankyou everyone for your such kind responses :') I will take the day to digest (no pun intended) all the helpful information!

BigNortherner
Posts: 268
Joined: Mon Jun 12, 2006 8:20 pm

Re: Looking for OSCAR interpretation assistance and Aerophagia

Post by BigNortherner » Sun May 11, 2025 6:26 pm

robysue1;

Why is sleeping on your left side much better for reducing burping than on your right side?

User avatar
robysue1
Posts: 1311
Joined: Sun Sep 18, 2022 3:39 pm
Location: Buffalo, NY

Re: Looking for OSCAR interpretation assistance and Aerophagia

Post by robysue1 » Mon May 12, 2025 10:26 am

BigNortherner wrote: ↑
Sun May 11, 2025 6:26 pm
robysue1;

Why is sleeping on your left side much better for reducing burping than on your right side?
Unlike our outsides, our insides are not symmetric. It turns out that the location and the orientation of the valve that closes off the top of the stomach from the esophagus makes it sensitive to which side you are lying on: That valve is much more prone to opening if you are on your right side. And that can lead to both problems with GERD (acid reflux) as well as aerophagia---i.e. air getting into the stomach.

I first learned about this fact way back when I was pregnant with my oldest child and suffering from severe indigestion problems that kicked mainly when I was trying to get to sleep. At the time, I slept on my right side almost all the time. After complaining about the problem to my OB, he explained that location and orientation of that valve made the valve's job of staying shut harder when lying on one's right side. Sleeping on my left side did a lot to help.

Fast forward a bit more than 2 decades and when I started CPAP? Lots and lots of aerophagia. And, at the time I was sleeping about 50% of the time on my right side. And when I woke up in the night with the aerophagia at its worst? I was usually on my right side. Now, conquering the aerophagia wasn't as simple (in my case) as retraining myself to sleep almost exclusively on my left side. I had a really hard, prolonged adjustment period to CPAP which included a really nasty bit of severe insomnia. And the more time I spent in bed not sleeping or trapped in long periods of SWJ, the more severe the aerophagia was. In other words, the more I aroused at night, the more air managed to get into my mouth, and the more I wound up unconsciously swallowing that air instead of opening my mouth to let it out that way. And the more air I swallowed, the worse the aerophagia became until I would fully wake up in the middle of the night with a rock-hard stomach full of air. Lying on my left side did help (some) with preventing air from sneaking into the stomach, but even today, after 15 years of PAPing, I still find aerophagia can be an issue if I lie on my right side while trying to get to sleep or trying to get back to sleep after a middle of the night wake.
Joined as robysue on 9/18/10. Forgot my password & the email I used was on a machine that has long since died & gone to computer heaven.

Correct number of posts is 7250 as robysue + what I have as robysue1

Profile pic: Frozen Niagara Falls