There are two more settings you can consider changing in the future.
Caution: placing these two sensitivity settings too high can lead you to feel as if the CPAP is driving your breathing pattern rather than you being in control of the CPAP machine.
Bigcatdaddy wrote: ↑Fri Aug 04, 2023 8:02 pm
—feels like I’m pulling too hard to get air. I want a slightly larger volume of air to come into my lungs, and I want it to flow in much more easily.
Changing the Trigger setting to High or Very High will speed up when the CPAP transitions from EPAP to IPAP. This change might help with your perception of needing to pull hard to get air.
Bigcatdaddy wrote: ↑Fri Aug 04, 2023 8:02 pm
—feels like I’m pushing too hard to exhale against the pressure. Also when I’m done exhaling, I feel there’s still air in my lungs that needs to come out—and to get it out means I have to consciously force more out.
Similarly setting the Cycle to High or Very High will speed up the CPAP in transitioning from IPAP to EPAP.
However the feeling that you have air left over to breathe out could mean that the Trigger setting is too high.
So these two settings can both work for/against what you feel.
As PR explains:
palerider wrote: ↑Tue Nov 06, 2018 8:42 pm
Trigger is how sensitive the machine is to your starting to take a breath... the lower the trigger, the more of a breath you have to take before the machine goes "AHA! time to switch to IPAP".
TiMin is the minimum amount of time that it will stay at IPAP.
TiMax is the maximum amount of time it'll stay at IPAP.
Cycle is how sensitive the machine is to your starting to exhale.... and when it'll drop the pressure back to EPAP, (if TiMax hasn't already expired, if TiMax expires, then the machine goes back to EPAP then.
For *most* people, just leave those settings at default.
Page 5 of the VAuto Clinical Guide has a diagram showing the timing.
Here's a link to to the guide if you don't have it yet.
Code: Select all
https://www.respshop.com/manuals/37212_clinician_menu.pdf
Good luck.