air curve 10 asv settings
air curve 10 asv settings
Hi my problem is after ramp time ends and the pressure increases the pressure is to long of a time, when ready to exhale the pressure is still flowing in.
I use a nose pillow. my settings are:
min epap 4.0
max epap 11.6
mim ps 3.0
max ps 10,0
unable to inhale/ exhale normally. air backs up in mouth.
Hope you can help.
Thank you,
Don
I use a nose pillow. my settings are:
min epap 4.0
max epap 11.6
mim ps 3.0
max ps 10,0
unable to inhale/ exhale normally. air backs up in mouth.
Hope you can help.
Thank you,
Don
- Respirator99
- Posts: 389
- Joined: Mon Jul 05, 2021 12:39 am
- Location: Australia
Re: air curve 10 asv settings
Your pressures are low enough that you really don't need the ramp. I suggest you turn it off. I'm running somewhat higher pressure and found that the transition from ramp to normal mode was too abrupt and would wake me - I got around it by not using ramp at all.aarpdon wrote: ↑Tue Dec 27, 2022 11:34 amHi my problem is after ramp time ends and the pressure increases the pressure is to long of a time, when ready to exhale the pressure is still flowing in.
I use a nose pillow. my settings are:
min epap 4.0
max epap 11.6
mim ps 3.0
max ps 10,0
unable to inhale/ exhale normally. air backs up in mouth.
Hope you can help.
Thank you,
Don
The machine will accommodate itself to your breathing pattern - start out breathing deeply and evenly and the machine will match you. If it does blow too hard, just blow back hard. This will cause the machine to back off and re-accommodate itself to you.
_________________
| Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier |
| Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
* Download Oscar
* Oscar help
* An alternative to Oscar - try SleepHQ
I have no medical training or qualifications. Take my advice for what it's worth.
* Oscar help
* An alternative to Oscar - try SleepHQ
I have no medical training or qualifications. Take my advice for what it's worth.
Re: air curve 10 asv settings
Don,
I believe the minimum time inhale setting (TI Min) on your machine is set too high. The setting needs to be reduced from the existing setting. I would suggest reducing it in small increments until you are able to breath comfortably. If it’s set too low you will feel like you are gasping for air.
Hope this helps,
Sam
I believe the minimum time inhale setting (TI Min) on your machine is set too high. The setting needs to be reduced from the existing setting. I would suggest reducing it in small increments until you are able to breath comfortably. If it’s set too low you will feel like you are gasping for air.
Hope this helps,
Sam
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Been doing this for 29 years |
Re: air curve 10 asv settings
There is no TiMin setting on an ASV.
Please don't guess about things like that.
It misleads people that are looking for help.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: air curve 10 asv settings
Don,
I apologize. Palerider is correct. I jumped in because I had an issue similar to yours, however, I did so without getting a full understanding beforehand.
Bad advice is worse than no advice at all. I hope you are able to resolve your issue quickly.
Sam
I apologize. Palerider is correct. I jumped in because I had an issue similar to yours, however, I did so without getting a full understanding beforehand.
Bad advice is worse than no advice at all. I hope you are able to resolve your issue quickly.
Sam
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Been doing this for 29 years |
Re: air curve 10 asv settings
Just trying to make sure that I understand what you are saying.
When you start to exhale, the pressure is still at IPAP and does not decrease to EPAP? Is that what you are trying to describe?
If that's the case, then we need to see the following settings for your machine:
Ti_min
Ti_max
Trigger
Cycle
My guess is that your Ti_min may be set too long. And you may need to change the Cycle as well.
Tell us those settings and we'll be able to provide some ideas on how to get the machine's changing from IPAP to EPAP and back in better synchronization with your breathing so that air isn't "backing up" into your mouth.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Also use a P10 mask |
Joined as robysue on 9/18/10. Forgot my password & the email I used was on a machine that has long since died & gone to computer heaven.
Correct number of posts is 7250 as robysue + what I have as robysue1
Profile pic: Frozen Niagara Falls
Correct number of posts is 7250 as robysue + what I have as robysue1
Profile pic: Frozen Niagara Falls
Re: air curve 10 asv settings
Ti_min should never leave you feeling like you are gasping for air: Ti_min is the minimum amount of time the machine will stay at IPAP before it is allowed to cycle to EPAP. As long as your inhalation is at least as long as Ti_min, your exhalation should cause the machine to cycle to EPAP.Sam_01 wrote: ↑Wed Dec 28, 2022 9:18 amDon,
I believe the minimum time inhale setting (TI Min) on your machine is set too high. The setting needs to be reduced from the existing setting. I would suggest reducing it in small increments until you are able to breath comfortably. If it’s set too low you will feel like you are gasping for air.
Now, if Ti_max is set too short, then that can cause the machine to cycle to EPAP before you start to exhale. And that could lead to feeling like you're gasping for air at the end of your inhalation.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Also use a P10 mask |
Joined as robysue on 9/18/10. Forgot my password & the email I used was on a machine that has long since died & gone to computer heaven.
Correct number of posts is 7250 as robysue + what I have as robysue1
Profile pic: Frozen Niagara Falls
Correct number of posts is 7250 as robysue + what I have as robysue1
Profile pic: Frozen Niagara Falls
- Respirator99
- Posts: 389
- Joined: Mon Jul 05, 2021 12:39 am
- Location: Australia
Re: air curve 10 asv settings
The Resmed Aircurve 10 ASV DOES NOT HAVE THOSE SETTINGS. All those parameters are taken care of in software - there is no way to adjust them.robysue1 wrote: ↑Thu Dec 29, 2022 6:17 pmIf that's the case, then we need to see the following settings for your machine:
Ti_min
Ti_max
Trigger
Cycle
My guess is that your Ti_min may be set too long. And you may need to change the Cycle as well.
Tell us those settings and we'll be able to provide some ideas on how to get the machine's changing from IPAP to EPAP and back in better synchronization with your breathing so that air isn't "backing up" into your mouth.
_________________
| Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier |
| Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
* Download Oscar
* Oscar help
* An alternative to Oscar - try SleepHQ
I have no medical training or qualifications. Take my advice for what it's worth.
* Oscar help
* An alternative to Oscar - try SleepHQ
I have no medical training or qualifications. Take my advice for what it's worth.
Re: air curve 10 asv settings
Thanks everyone for the FYI. You guy's and girls are great Happy New Year.....
I'm going to switch to full face mask (so I can blow back air and see if it helps) and see how it goes .
Can you explain the Min EPAP and Max EPAP and the Min PS and Max PS, which setting would increase air flow and decrease air flow.
Don
I'm going to switch to full face mask (so I can blow back air and see if it helps) and see how it goes .
Can you explain the Min EPAP and Max EPAP and the Min PS and Max PS, which setting would increase air flow and decrease air flow.
Don
- Respirator99
- Posts: 389
- Joined: Mon Jul 05, 2021 12:39 am
- Location: Australia
Re: air curve 10 asv settings
EPAP, PS and IPAP are all linked.arpdon wrote:Can you explain the Min EPAP and Max EPAP and the Min PS and Max PS, which setting would increase air flow and decrease air flow.
EPAP is the exhale pressure
IPAP is the inhale pressure
PS is pressure support - the pressure increase needed to get from EPAP to IPAP.
EPAP + PS = IPAP.
EPAP is your exhale pressure and basically controls obstructive events such as obstructive apneas and snores. EPAPmin should be set to a comfortable low pressure for normal breathing. If an obstructive event is encountered, the machine will increase EPAP to overcome it. EPAPmax needs to be high enough that the machine can eliminate all (or most) of these obstructive events. If you're getting a lot of apneas (which are reported as UA) then increase the EPAPmax.
PSmin needs to be set at a comfortable level - I have mine on 3, which gives a comfortable inhale / exhale transition. A high PSmin will increase ventilation in normal breathing, but isn't really needed as the machine can adjust as required. Resmed recommend PSmax be >15 - this will allow the machine to send big puffs of air to get you breathing in the event of a central apnea.
_________________
| Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier |
| Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
* Download Oscar
* Oscar help
* An alternative to Oscar - try SleepHQ
I have no medical training or qualifications. Take my advice for what it's worth.
* Oscar help
* An alternative to Oscar - try SleepHQ
I have no medical training or qualifications. Take my advice for what it's worth.