Auto CPAP vs BiPAP with pressure support 0

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
void
Posts: 7
Joined: Thu Sep 10, 2020 10:58 am

Auto CPAP vs BiPAP with pressure support 0

Post by void » Tue Apr 05, 2022 9:24 am

I got Resmed Aircurve 10VAuto.
1. If I set Vauto mode and pressure support = 0, EPAP=4, IPAP=15, Does it operate exactly like the Auto CPAP (Airsense 10) with pressure range from 4 to 15?
2. Do you think that the 2 machines should operate the same with same pressures range, if centrals occur?

User avatar
Pugsy
Posts: 65266
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Auto CPAP vs BiPAP with pressure support 0

Post by Pugsy » Tue Apr 05, 2022 10:20 am

1. Yes.....as long as you aren't using EPR on the AutoSet.
2. Don't really understand your question but if you mean "will the VAuto do anything to address any centrals" then the answer is no.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Auto CPAP vs BiPAP with pressure support 0

Post by palerider » Tue Apr 05, 2022 10:40 am

void wrote:
Tue Apr 05, 2022 9:24 am
I got Resmed Aircurve 10VAuto.
1. If I set Vauto mode and pressure support = 0, EPAP=4, IPAP=15, Does it operate exactly like the Auto CPAP (Airsense 10) with pressure range from 4 to 15?
2. Do you think that the 2 machines should operate the same with same pressures range, if centrals occur?
The VAuto will do everything an autoset will do, and more, you can program it to act like four different machines.

so yes, MinIPAP 4, MaxIPAP 15, PS 0 is the same as an autoset of min 4, max 15, epr off.

Those are likely inadequate settings though.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.