To my surprise, although a Medium, the cushion felt a little tighter around the nose, and the leak rate was far lower than with the previous China lot.
These are the readings:
Night#/95% Leak Rate
====================
Week of May 9 (China lot):
1/22
2/25
3/20
4/19
5/19
6/18
Week of May 16 (Costa Rica lot):
1/3.6
2/28
3/8.4
4/11
5/13
The cushions are obviously dissimilar as show in the photo.

I provided these details to Philips Patient Support and they said the issue would be investigated further.
Also considering reporting this to the FDA through their public complaint system, MedWatch, as this appears to be a Medical Device manufacturing quality control problem.