Aircurve 10 ASV Pressure bar

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
WesternEagle
Posts: 3
Joined: Tue Jul 28, 2015 10:00 pm

Aircurve 10 ASV Pressure bar

Post by WesternEagle » Thu Feb 11, 2021 4:31 pm

I was 5 years with a Resperonics System One.
I now have a new Aircureve 10 ASV.

IN the Manual it says, "You will know that therapy is on when the SLEEP REPORT is displayer. The 'Pressure bar shows the inspiratory and expiratory pressures in green. The GREEN bar will expand and contract as you breathe in and out."

When I start 'therapy' the screen comes up, but the green bar does NOT move or 'expand and contract as I breath in and out."

I called the supplier teck guy and he told me it was because I had it on "Ramp" I told him I did not have it on Ramp. He then said to call Resmed.
I did and they did not know why it was not working and to take it back to the supplier.

So I thought I would ask here first.
I have attached a picture of the screen (with the bar that does not move.)
Thanks
M Jones
Attachments
IMG_20210210_081057795.jpg
IMG_20210210_081057795.jpg (6.14 KiB) Viewed 377 times

User avatar
Pugsy
Posts: 65327
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Aircurve 10 ASV Pressure bar

Post by Pugsy » Thu Feb 11, 2021 5:16 pm

Lets see what the settings are and what the machine pressure graph looks like.

OSCAR https://www.sleepfiles.com/OSCAR/
http://www.apneaboard.com/wiki/index.ph ... stallation
http://www.apneaboard.com/wiki/index.ph ... rpretation


viewtopic/t158560/How-to-post-images-for-review.html

I can't see your image well enough to tell much of anything about it. Too tiny and I can't make it larger but we really need to see what all the settings are along with the graphs.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

WesternEagle
Posts: 3
Joined: Tue Jul 28, 2015 10:00 pm

Re: Aircurve 10 ASV Pressure bar

Post by WesternEagle » Thu Feb 11, 2021 6:06 pm

The first line says
ASVAUTO and to the right of it is PS 0.0

Underneath the green bar is

8.0 on the left and 8.0 on the right.

User avatar
Pugsy
Posts: 65327
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Aircurve 10 ASV Pressure bar

Post by Pugsy » Thu Feb 11, 2021 6:28 pm

It won't change with inhale and exhale if the PS is 0.0.

You need to confirm the settings in the clinical setup menu area....or use the software.

But you won't see a pressure change while awake with PS set to 0.0.
You need PS to be greater than 0.0 to see what your manual says change between inhale and exhale.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

WesternEagle
Posts: 3
Joined: Tue Jul 28, 2015 10:00 pm

Re: Aircurve 10 ASV Pressure bar

Post by WesternEagle » Thu Feb 11, 2021 7:16 pm

Thank you very much.
Very reasonable answer. I will check it out.

Thanks for your quick reply!!